Tektronix 2246 Oscilloscope CRT Sparkle/Noise - Normal?

  • Thread starter jordankersten
  • Start date
  • Tags
    Crt Noise
In summary, the CRT "sparkle" or "noise" is present on the traces and readout displays and cursors of the Tektronix 2246 Mod A Oscilloscope. It is slight and visible both on the traces and the readout. It could be the result of an issue with the CRT circuitry, power supply, or AC line noise.
  • #1
jordankersten
22
0
Hello everyone. I am new and have a question. I just got a Tektronix 2246 Mod A Oscilloscope. Everything works great on it. The CRT seems to have some "sparkle" or "noise" in it. It is very slight and it is visible both on the traces and the readout. It looks like the traces and readout displays and cursors have sparkling spots in it. Is this normal? If not, what could be causing it?
 
Engineering news on Phys.org
  • #2
One more thing to add...the problem gets worse the more things that are on the screen: it is worst when all 4 channels are displayed with cursors, etc. It smoothes out when the timing is faster but is still present.
 
  • #3
Sparkling reminds me of dram memory access aliasing. It could be the guy who designed the video memory at Tektronix didn't bother to do the worst case timing analysis. Thing is, you can suffer some of this without crashing the cpu, because its just the user interface, so sloppy design work gets by. Aging might aggravate marginal timing.

But if this scope has a vector drawn display, I don't know how much of this applies.
 
Last edited:
  • #4
Hard to tell from the description.
Could be a leaky HV power supply or noisy transistor in the Z channel.
 
  • #5
It gets worse the more traces you put up on the screen (its a 4 channel scope). The "dashing" is worse on the vertical deflection. A problem in th CRT circuitry? Power supply? AC Line noise?
 
  • #6
jordankersten said:
It gets worse the more traces you put up on the screen (its a 4 channel scope). The "dashing" is worse on the vertical deflection. A problem in th CRT circuitry? Power supply? AC Line noise?
First guess would be to look at the blanking circuitry in the Z channel.
This is supposed to shut the beam off while the vertical beam position is being stepped.
So vertical (or somewhat slanted) dashes on the screen would indicate a problem in this area.
 
  • #7
I guess you already know this but...
The 2246 is an old analogue scope meaning there is probably only a single beam drawing all four channels; this means that there is presumably a CHOP/ALT button somewhere which controls how the channels are drawn.
If this button it set to CHOP the lines will indeed look like they are made up of spots since the beam continuosly jumps between the lines spending 1/4 of its time on each channel meaning only 1/4 of each line is actually drawn (in ALT mode the scope draws one complete channel at a time).
 
  • #8
f95toli said:
I guess you already know this but...
The 2246 is an old analogue scope meaning there is probably only a single beam drawing all four channels; this means that there is presumably a CHOP/ALT button somewhere which controls how the channels are drawn.
If this button it set to CHOP the lines will indeed look like they are made up of spots since the beam continuosly jumps between the lines spending 1/4 of its time on each channel meaning only 1/4 of each line is actually drawn (in ALT mode the scope draws one complete channel at a time).


Not likely to see this because the scope will not be triggering at rate to see these spots. I used to work on analog scopes for a living. Another scope is needed to troubleshoot the defective one.
 
  • #9
Averagesupernova said:
Not likely to see this because the scope will not be triggering at rate to see these spots. I used to work on analog scopes for a living. Another scope is needed to troubleshoot the defective one.
Agreed. It doesn't matter what mode its in (chop or alt) it does the same thing. It gets worse when Line triggering is used. Could it be a power supply filter? Line noise leaking in?
 
  • #10
jordankersten said:
Agreed. It doesn't matter what mode its in (chop or alt) it does the same thing. It gets worse when Line triggering is used. Could it be a power supply filter? Line noise leaking in?

Sounds likely that it is some type of power supply noise. NoTime has already suggested things of this nature.
 
  • #11
Here are some pics of the display problem...one is 10X mag chop and one is 10X mag alt. It smoothes out when you turn the readout off...intensity of the readout has no affect, only if it is on or off.
 

Attachments

  • DSC00771.jpg
    DSC00771.jpg
    42.6 KB · Views: 525
  • DSC00774.jpg
    DSC00774.jpg
    36 KB · Views: 485
  • #12
The problem goes away when the readout is turned off. The readout intensity doesn't affect it at all, but when you turn the readout intensity all the way down (off) the problem disappears.
 
  • #13
Ok, the service manual says that When ROR (Readout Request or something like that) input is low, chop blanking is disabled and the ROB (Read Out Blanking) is inverted and controls the blanking. When ROR goes from low to high, the blank output remains connected to the readout blank signal for an additional four to six TC clock periods to mask vertical source-switching transients. My question is this: on my ROR, high input is 5.06V and low is 4.65. Is 4.65 low enough to be considered "low" and enable the readout blanking? When the readout pot is turned all the way off, I get the 5.06 supply voltage on the ROR pin. When the readout is on (doesn't matter to what intensity, just on) then the ROR drops to 4.65V to enable the readout blanking (ROB). Does that voltage difference sound right?
 
  • #14
The 4.65 is unlikely to be a DC voltage, so a DVM is not going to show a meaningful low.

It has been a while since I looked at one of these.
But, The display images you posted don't really look out of line.
So barring another sample that doesn't show this artifact I'm inclined to think there is nothing wrong.
Perhaps someone else here has seen this model more recently.
 
  • #15
If the 7.5 Volt adjust is at exactly 7.5V, the "dashes" produced by the readout scans across the trace. If you bring the 7.5V adjust lower, it scans faster and faster. If you bring the 7.5V up, the "dashes" scan faster and faster disappearing right around 8 volts. All other voltages check out ok. The 15v supply is almost 16v when 7.5v is exactly 7.5. All other voltages are right on.
 
  • #16
One more thing: if there are two readout displays on (CH1 and time) there are two sets of dashes that scan across the trace. If you add other things to the readout (CH1 CH2, time, etc.) another set of dashes scans across the trace for everything you add to the readout.
 
  • #17
From the display, it looks like the same goofy tectronixs i'd been using for the past 10 years.
I don't see the sparkling you're having trouble with, just some segmented traces.
 
  • #18
It's not really sparkle. It's just "dashed" or broken up when the readout is on.
 

1. What is causing the sparkle or noise on my Tektronix 2246 oscilloscope's CRT?

The most common cause of sparkle or noise on the CRT of a Tektronix 2246 oscilloscope is a buildup of dust or debris on the screen. This can be easily resolved by carefully cleaning the CRT with a soft, lint-free cloth.

2. Can the sparkle or noise on my Tektronix 2246 oscilloscope's CRT affect the accuracy of my measurements?

In most cases, the sparkle or noise on the CRT of a Tektronix 2246 oscilloscope will not affect the accuracy of your measurements. However, if the noise is particularly severe, it may interfere with your ability to accurately read the waveform.

3. How can I prevent or reduce sparkle and noise on my Tektronix 2246 oscilloscope's CRT?

To prevent or reduce sparkle and noise on the CRT of your Tektronix 2246 oscilloscope, it is important to regularly clean the screen with a soft, lint-free cloth. Additionally, keeping the oscilloscope in a clean and dust-free environment can also help reduce the buildup of debris on the CRT.

4. If cleaning the CRT does not resolve the sparkle or noise on my Tektronix 2246 oscilloscope, what could be causing it?

In some cases, the sparkle or noise on the CRT of a Tektronix 2246 oscilloscope may be caused by a faulty component or electrical interference. If cleaning the CRT does not resolve the issue, it is best to consult a professional for further troubleshooting and repairs.

5. Is there any way to adjust the settings on my Tektronix 2246 oscilloscope to reduce sparkle and noise on the CRT?

Yes, there are several adjustments that can be made on the Tektronix 2246 oscilloscope to reduce sparkle and noise on the CRT. These include adjusting the focus and intensity controls, as well as the position and size of the trace. Refer to the oscilloscope's manual for specific instructions on how to make these adjustments.

Similar threads

Replies
62
Views
3K
  • Electrical Engineering
Replies
9
Views
4K
  • STEM Educators and Teaching
2
Replies
40
Views
2K
Replies
12
Views
7K
  • Precalculus Mathematics Homework Help
Replies
3
Views
2K
  • General Engineering
Replies
4
Views
2K
  • Electrical Engineering
Replies
4
Views
3K
  • Electrical Engineering
Replies
4
Views
1K
  • Computing and Technology
Replies
15
Views
1K
  • General Engineering
Replies
4
Views
4K
Back
Top