Can an instrument ever have an accuracy better than its repeatably?

In summary, the pump is more accurate under static conditions, but may have inaccuracies under varying conditions due to hysteresis.
  • #1
rollingstein
Gold Member
646
16
I have a dosing pump spec. sheet from Milton Roy (a fairly reputable manufacturer) which says:

Steady State Accuracy: +/- 1%
Repeatably: +/- 3%

Is this possible? Can an instrument be more accurate than it is repeatable? Intuitively, this didn't make sense to me.

The spec. sheet doesn't say, but I'm assuming that both are expressed as percent of reading or full scale deflection but not a mix.

http://www.miltonroyindia.com/M-Series.pdf (Look at Page 2)
 
Engineering news on Phys.org
  • #2
They are different measures.
Check what they mean by "steady state accuracy".
 
  • #3
I think I now understand it (perhaps).

You mean the "accurate but not precise" case?

accuracy-vs-precision.jpg


I cannot find a definition for "Steady State Accuracy" on their site. I took it to mean the offset measured against some better gold standard callibrating instrument without changing flowrates.

I'm still confused. I was under the impression that repeatablity puts a hard upper bound on how accurate one can be.
 
  • #4
accurate vs precise - that's right.
but without knowing how the manufacturer is arriving at the values - it's anyone's guess.
 
  • #5
Simon Bridge said:
accurate vs precise - that's right.
but without knowing how the manufacturer is arriving at the values - it's anyone's guess.

Funnily, I've never seen a spec. sheet of an instrument that defined either of these terms.

So I guess there's a conventionally agreed upon definition?
 
  • #6
You have presumably googled them to see?
 
  • #7
Well I would assume that steady state means that if you set the instrument to do X then it will continue to do that with an accuracy +/-1%.
-
If the controls are moved and returned to the original position I would assume you can expect it to be within +/- 3% of where it was before you moved it. In other words, with calibration it can always get results of +/-1% repeatable. I look at as a way the manufacturer made the accuracy look better than it is and still avoid liability issues. I used to work for a test equipment manufacturer and I know the games they like to play when it comes to specs.
 
  • #8
Averagesupernova said:
Well I would assume that steady state means that if you set the instrument to do X then it will continue to do that with an accuracy +/-1%.
-
If the controls are moved and returned to the original position I would assume you can expect it to be within +/- 3% of where it was before you moved it. In other words, with calibration it can always get results of +/-1% repeatable. I look at as a way the manufacturer made the accuracy look better than it is and still avoid liability issues. I used to work for a test equipment manufacturer and I know the games they like to play when it comes to specs.

Hmm...I interpreted it to mean for a actual flow of say 100 if many readings are taken it may read 101, 102, 103, 99, 98, 97.

The problem with such an interpretation is that unless they specify how many readings they average over, it's meaningless.
 
  • #9
Specs are always meaningless unless they are defined.
 
  • #10
rollingstein said:
I think I now understand it (perhaps).

You mean the "accurate but not precise" case?

accuracy-vs-precision.jpg


I cannot find a definition for "Steady State Accuracy" on their site. I took it to mean the offset measured against some better gold standard callibrating instrument without changing flowrates.

I'm still confused. I was under the impression that repeatablity puts a hard upper bound on how accurate one can be.

I love the Target example. I'm going to apply my answer in the regard of taking measurements (and not the operation of this pump). It explains the relationship between precision (inherent in the sensor) and accuracy (the true ability to give a correct reading of the condition from the sensor). Accuracy is normally achieved using calibration for the sensor for a known condition.

For static conditions the Target example is correct. However under actual conditions where the measured medium is moving - you have to think of the Target as moving too (and your measurements as the bullets fired at the target).

Taking a measurement is essentially taking a snapshot of a condition in a given time frame. A single reading cannot accurately portray the dynamic condition that may exist at a site. The instrument can be precise, the instrument can be accurate - but without multiple measurements you cannot actually determine the true nature of the condition you are measuring.

Now, on to this Pump. As stated, it is more accurate under static (stable) conditions. Under varying conditions it may have inaccuracies due to hysteresis. That would be my guess for this situation.

My first entry here. Hope it was helpful.

Dave
 

1. Can an instrument's accuracy ever be perfect?

No, it is not possible for an instrument to have perfect accuracy. There will always be some level of error or uncertainty in measurements due to various factors such as human error, environmental conditions, and limitations of the instrument itself.

2. How does repeatability affect an instrument's accuracy?

Repeatability refers to the ability of an instrument to produce consistent results when measuring the same quantity multiple times. While it is important for an instrument to have good repeatability, it does not necessarily guarantee high accuracy. Other factors such as calibration and precision also play a role in an instrument's accuracy.

3. Can an instrument's accuracy improve over time?

Yes, with proper maintenance and calibration, an instrument's accuracy can improve over time. Regular calibration and adjustments ensure that an instrument is performing at its best and can help to reduce errors that may arise over time due to wear and tear.

4. What is the difference between accuracy and precision?

Accuracy refers to how close a measurement is to the true or expected value. Precision, on the other hand, refers to how consistent and reproducible a measurement is. An instrument can be precise but not accurate if it consistently produces the same result that is far from the true value. Similarly, an instrument can be accurate but not precise if it produces different results that are all close to the true value.

5. How can an instrument's accuracy be verified?

An instrument's accuracy can be verified through calibration, which involves comparing the instrument's measurements to a known standard. This can be done using certified reference materials or by sending the instrument to a calibration laboratory. Regular verification of an instrument's accuracy is important for ensuring the reliability of its measurements.