In summary, the purpose of the R/2 resistor in the augmented limb lead voltages measurement is to balance the input bias current and create a signal-weighting network that transforms the RA-LA-LL "coordinate system" to another coordinate system with increased sensitivity to certain abnormalities and disease states. This is achieved by combining the RA-LA-LL electrodes with the resistor network to produce sensitivity vectors that are equally spaced in angle. The combination of all six leads is useful in detecting diseases and abnormalities. Other lead systems, such as the Frank VECG, also contribute to a more comprehensive understanding of cardiac activation currents. Overall, the resistor network plays a critical role in producing accurate and reliable ECG readings for medical diagnosis and treatment.
  • #1
pumas
15
0
Hello,

To measure the augmented limb lead voltages aVr, aVl and aVf two resistors of value R are used on the (-) side and a resistance R/2 on the (+) side. Could anyone please explain me what is the purpose of the R/2 resistor. I attached a picture.


Thank you
 

Attachments

  • Augmented limb lead voltages.doc
    31 KB · Views: 293
Biology news on Phys.org
  • #2
The directional sensitivity of the standard lead system to cardiac activation currents lies in the anterior or coronal plane, but is rather skewed: I is horizontal, II is approximately 60 deg (down and to left), III is 120 deg (down and to right). The augmented limb leads use the same electrodes (LA, RA and LL) but weight and combine them with resistor networks to produce sensitivity vectors that are equally spaced in angle. Diseases and abnormalities will often show up best in one lead or another, so the combination of all six is useful. See the lead sensitivities towards the beginning, and diagnostic examples towards the end, of this:

"medresidents.stanford.edu/TeachingMaterials/EKG%20Basics/EKG%20Basics%20-%20Long.ppt"[/URL]

There are other lead systems, too. The Frank VECG (vector ecg) is notable because it includes a lead with sensitivity in the sagittal plane (front to back).
 
Last edited by a moderator:
  • #3
marcusl said:
The directional sensitivity of the standard lead system to cardiac activation currents lies in the anterior or coronal plane, but is rather skewed: I is horizontal, II is approximately 60 deg (down and to left), III is 120 deg (down and to right). The augmented limb leads use the same electrodes (LA, RA and LL) but weight and combine them with resistor networks to produce sensitivity vectors that are equally spaced in angle. Diseases and abnormalities will often show up best in one lead or another, so the combination of all six is useful. See the lead sensitivities towards the beginning, and diagnostic examples towards the end, of this:

"medresidents.stanford.edu/TeachingMaterials/EKG%20Basics/EKG%20Basics%20-%20Long.ppt"[/URL]

There are other lead systems, too. The Frank VECG (vector ecg) is notable because it includes a lead with sensitivity in the sagittal plane (front to back).[/QUOTE]


Awesome post, marcusl. And thanks for the link to the PPT file -- very helpful.
 
Last edited by a moderator:
  • #4
Glad to be of help :eek:)
 
  • #5
marcusl said:
Glad to be of help :eek:)

You were more than basic help, marcusl. I followed some of the tree links from your post and found great study materials. I'm an EMT outside of my regular EE job. Thank you very much.
 
  • #6
it's been a really long since I've studied this, and looking through an old book wasn't a complete help so take it for what it's worth. but you've essentially got a two-lead signal feeding into an instrumentation op-amp. and even though the input impedance to the amp is high, the signals are small and the isolation resistors are high also. the two R leads balance out the input bias current so that your equivalent input resistance there is R/2. and so the single resistor lead also becomes R/2.

anyhoo, it's the sort of thing that only biomedical instrumentation engineers worry about. i may have the course notes on it somewhere...
 
  • #7
Proton Soup said:
it's been a really long since I've studied this, and looking through an old book wasn't a complete help so take it for what it's worth. but you've essentially got a two-lead signal feeding into an instrumentation op-amp. and even though the input impedance to the amp is high, the signals are small and the isolation resistors are high also. the two R leads balance out the input bias current so that your equivalent input resistance there is R/2. and so the single resistor lead also becomes R/2.
This description does not get to the purpose of the resistor network, and in any case the reality is more complicated. First, the input impedance of the instrumentation amps used in this application are incredibly high and bias currents are incredibly low--10 Gohm and 500 pA are typical even for cheap plastic-packaged IC's--so resistors in the leads are not there for balance. In fact the I, II and III leads are usually direct-coupled. (Of course there are three amplifiers, one for each lead pair).

Since patient safety is critical, the amps are commonly followed by optically-coupled isolation amps eliminating any possible connection to Earth ground or power even if the ECG machine is AC powered.

Finally, the input amps are ground-referenced to yet another electrode which is standardized to be the right leg. It is common to actually drive the right leg to minimize common-mode potentials, since the patient is floating with respect to Earth ground.

Which brings us back to the resistors; they form a signal-weighting network that transforms the RA-LA-LL "coordinate system" to another coordinate system with increased sensitivity to certain abnormalities and disease states, as I stated earlier.
 
  • #8
marcusl said:
This description does not get to the purpose of the resistor network, and in any case the reality is more complicated. First, the input impedance of the instrumentation amps used in this application are incredibly high and bias currents are incredibly low--10 Gohm and 500 pA are typical even for cheap plastic-packaged IC's--so resistors in the leads are not there for balance. In fact the I, II and III leads are usually direct-coupled. (Of course there are three amplifiers, one for each lead pair).

Since patient safety is critical, the amps are commonly followed by optically-coupled isolation amps eliminating any possible connection to Earth ground or power even if the ECG machine is AC powered.

Finally, the input amps are ground-referenced to yet another electrode which is standardized to be the right leg. It is common to actually drive the right leg to minimize common-mode potentials, since the patient is floating with respect to Earth ground.

Which brings us back to the resistors; they form a signal-weighting network that transforms the RA-LA-LL "coordinate system" to another coordinate system with increased sensitivity to certain abnormalities and disease states, as I stated earlier.

if there's no current flow going on here, then how do you suppose the weighting of the resistors factors in? should be simple to show how changing R/2 to R will change it, no? i didn't see where your ppt presentation addresses this.

rather, i think it's about CMRR.
 
  • #9
These resistors aren't attached to the patient, they are located after the amplifiers in the ECG instrument. Their relative values are chosen to produce the desired output weighting.
 
  • #10
marcusl said:
These resistors aren't attached to the patient, they are located after the amplifiers in the ECG instrument. Their relative values are chosen to produce the desired output weighting.

link me to a circuit that shows this.
 
  • #11
I suspect it's in Geddes and Baker, Applied Biomed Instrumentation, but I don't have my copy handy. You can also look at Fig. 2, p. 23 of
http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1991-10.pdf"
showing one of the 3 channels in a clinical product.
 
Last edited by a moderator:
  • #12
marcusl said:
I suspect it's in Geddes and Baker, Applied Biomed Instrumentation, but I don't have my copy handy. You can also look at Fig. 2, p. 23 of
http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1991-10.pdf"
showing one of the 3 channels in a clinical product.

thanks. the book I'm looking at is Medical Instrumentation Application and Design, Webster, editor, 1992, so it's from about the same era as that HP equipment.

the first stage of that circuit appears to be an instrumentation amplifier, which is used primarily because it has very high Common Mode Rejection Ratio (CMRR) and impedance. the precision resistors probably correspond to the resistors in the instrumentation amp. they have to be laser trimmed to maximize CMRR.

http://en.wikipedia.org/wiki/Instrumentation_amplifier

the resistors that have to be trimmed in this stage are the R1 resistors in the above link. and when i say "instrumenation amp", i don't mean the whole network in the link above, just the two input amps, the R1's, and Rgain. the exact value of Rgain is not critical.

at this point they break the circuit up with muxes so that they can switch between other pairs of instrumentation amps and add the driven right leg circuit feedback (which also improves CMRR, fwiw). the part they have labeled as a differential amp corresponds to the amp with resistors R2 and R3 in the wiki link. resistor values aren't as critical now. and then they do some post processing on the signal.

if lead resistances are automatically chosen in a lead selector circuit, they are not shown in that HP literature figure. that figure appears to address only the ASIC they designed for the amplifier.

i'm sure the input(lead) resistors are still there no matter what. they are a low-tech failsafe to protect the patient in case you were to fry your amp and accidentally provide a low-impedance path to ground. making the aVr input R/2 simply balances the input impedance because the aVl and aVf combined in parallel work out to R/2 back to your signal reference.
 
Last edited by a moderator:
  • #13
Thanks, I am familiar with instrumentation amps and CMRR. The authors clearly indicate in the article's text that lead weighting is done after the input amps in the "precision resistor network" in the figure. Having said that, however, I agree with you that the lead weighting resistors can be put before the amplifiers, and maybe they was a time (before my time) when they always were. Current limiting resistors are actually a must to prevent instrument damage should a patient be defibrillated, and I always included them. And if resistors are used at the patient ports for either of these reasons, then I agree that the impedances need to be balanced. I mistakenly thought you were saying that they had to be there to balance the input bias currents.

As a historical note, Burr-Brown's 3656 was the first isolation amp IC available, and it made ECG design easy when it came out. The isolation was good up to 8kV and its isolated power supply easily powered three instrumentation input amps. An instrument designed around this unit, with proper layout and a plastic case, offered excellent patient safety. I found that the data sheets are still available online:
http://www.datasheetcatalog.com/datasheets_pdf/3/6/5/6/3656BG.shtml"
 
Last edited by a moderator:
  • #14
no, i just think it's the value it is to maximize CMRR.

and yeah, i remember Burr-Brown being very popular back then.
 
  • #15
Proton Soup said:
no, i just think it's the value it is to maximize CMRR.
Fair enough!
Proton Soup said:
and yeah, i remember Burr-Brown being very popular back then.
So you're an old-timer too!:biggrin:
 

1. What is an augmented limb lead?

An augmented limb lead is an electrical recording of the heart's activity that provides a different perspective than the standard 12-lead ECG. It is created by using three electrodes instead of two, and measuring the difference in electrical potential between each electrode and a central reference point.

2. Why is augmented limb lead used in ECG testing?

Augmented limb lead provides a more complete view of the heart's electrical activity, allowing for a more accurate diagnosis of heart conditions. It also helps identify abnormalities in certain areas of the heart that may not be visible on the standard 12-lead ECG.

3. How is augmented limb lead different from standard 12-lead ECG?

The standard 12-lead ECG uses 10 electrodes placed on the arms, legs, and chest to record the heart's electrical activity from 12 different perspectives. Augmented limb lead, on the other hand, uses three electrodes and measures the difference in electrical potential between each electrode and a central reference point, providing a different perspective on the heart's activity.

4. What conditions can be diagnosed using augmented limb lead?

Augmented limb lead can help diagnose a variety of heart conditions, including myocardial infarction (heart attack), cardiac arrhythmias (irregular heartbeats), cardiomyopathy (enlarged or weakened heart muscle), and conduction abnormalities (problems with the electrical signals that control the heart).

5. Is augmented limb lead safe?

Yes, augmented limb lead is a non-invasive and safe procedure. It does not involve any radiation, and the electrodes are simply placed on the skin's surface. However, as with any medical procedure, there may be rare cases of allergic reactions or skin irritation at the electrode sites.

Similar threads

  • Electrical Engineering
Replies
9
Views
4K
Replies
105
Views
7K
  • Classical Physics
Replies
6
Views
2K
Replies
2
Views
806
Replies
17
Views
1K
  • Electrical Engineering
Replies
4
Views
1K
  • Engineering and Comp Sci Homework Help
Replies
7
Views
880
Replies
3
Views
517
  • Introductory Physics Homework Help
Replies
3
Views
760
  • Electrical Engineering
Replies
6
Views
932
Back
Top