Determining resistance of oscilloscope.

In summary: Based on the graphs and figures I've attached, it seems as though probe 1 (the 10:1 probe) was used. The input impedance is 1 Mohm ± 1% (18 ±3 pF), which is what is specified in the user's guide. However, another reference I have for some unexplained reason states that the input impedance should be 50 ohm. To ascertain this, I would need to access the device itself and measure the input impedance. Alternatively, someone else who is familiar with the 'scope could measure it for me. In addition, I would like to estimate the error in time and voltage. The error in time is 50ppm, and the error in voltage is 3% of full scale. Based
  • #1
peripatein
880
0
Hi,
I'd appreciate some help in determining the internal resistance of the oscilloscope Agilent DSO1002A. Based on the user's guide the input impedance is: 1 Mohm ± 1% || 18 ±3 pF, but another reference I have for some unexplained reason states it should be 50 ohm. Which is correct?
 
Physics news on Phys.org
  • #2
Some oscilloscopes let you choose between 1 MOhm and 50 Ohm impedance.
Maybe the other reference suggested to use 50 Ohm (externally)?
 
  • #3
How may I ascertain this, whilst having no access to the device itself? I must use this in my calculations.
 
  • #4
It is a simple RC circuit.
 
  • #5
peripatein said:
Hi,
I'd appreciate some help in determining the internal resistance of the oscilloscope Agilent DSO1002A. Based on the user's guide the input impedance is: 1 Mohm ± 1% || 18 ±3 pF, but another reference I have for some unexplained reason states it should be 50 ohm. Which is correct?

peripatein said:
How may I ascertain this, whilst having no access to the device itself? I must use this in my calculations.

What do you mean you don't have access? Does that mean you did some experiments with it, and are now writing the report somewhere other than the lab?

Do you have any screenshots from the 'scope and your work? The coupling impedance may be noted somewhere on the screen. Alternately, can you call someone who can look at the 'scope for you? Just have them select that channel's coupling information menu, and it will state which input impedance is selected.
 
  • #6
I mean precisely that - I am currently working on a lab report somewhere other than the lab. It is also the weekend, mind you, so there is no one who could gain access to the device. I am attaching both the sketches of the circuits themselves and some of the graphs generated by the oscilloscope. Hopefully, via these you will be able to answer some of my questions.
 

Attachments

  • RCL.jpg
    RCL.jpg
    9.6 KB · Views: 847
  • RC.jpg
    RC.jpg
    6.8 KB · Views: 774
  • Waveform_2_EY1.jpg
    Waveform_2_EY1.jpg
    13.2 KB · Views: 891
  • #7
peripatein said:
I mean precisely that - I am currently working on a lab report somewhere other than the lab. It is also the weekend, mind you, so there is no one who could gain access to the device. I am attaching both the sketches of the circuits themselves and some of the graphs generated by the oscilloscope. Hopefully, via these you will be able to answer some of my questions.

You were using 10:1 oscilloscope probes, correct? They assume an input impedance of 1 MOhm into the 'scope's BNC connections. If you switch on the 50 Ohm input impedance option, that will attenuate the input waveform significantly. From the looks of your waveforms, the inputs were set to 1 MOhm.

Or were you directly connecting your test circuits to the 'scope BNC inputs with only coax cable and not 'scope probes?
 
  • #8
To the best of my knowledge we were instructed to use probe 1, i.e. plausibly 1:1 (and not 10:1). Moreover, were it 1 MOhm, would it not function as an open circuit (nearly infinite resistance)? We did use coax cables.
 
  • #9
So are you still inclined to believe it was 1 MOhm?
Furthermore, I would like to estimate the error in time and voltage. Based on the user's guide, the error in time is given as 50ppm whereas the error in voltage is 3% of full scale. Is the error in time then 50 * 10^-6 * resolution? But what is the resolution? 10^-6 sec (based on the graph I previously attached)?
As for the voltage, what is the "full scale"?
 
  • #10
peripatein said:
To the best of my knowledge we were instructed to use probe 1, i.e. plausibly 1:1 (and not 10:1). Moreover, were it 1 MOhm, would it not function as an open circuit (nearly infinite resistance)? We did use coax cables.

Here is a good intro from Tektrtonix about 'scope probe fundamentals:

www.ece.vt.edu/cel/docs/Probe_Fundamentals_Tektronix.pdf

A discussion of 'scope probe loading on circuits starts on page 32.

A 10:1 'scope probe has 9 MOhm in series with the input, to divide against the 1 MOhm input of the 'scope, as you can see in the attached figure from http://zone.ni.com/cms/images/devzone/ph/37d4325357.gif

That's how you get a nice high 10 MOhm input impedance to help avoid loading your test circuit. I don't know what a 1:1 probe would be, except maybe just straight coax. If that's what you used, you will need to look at your test circuit's impedances and the signal generator's output impedance and voltage level, to tell what the input impedance of the 'scope was. You should be able to figure it out from that.

http://zone.ni.com/cms/images/devzone/ph/37d4325357.gif
37d4325357.gif
 
Last edited by a moderator:
  • #11
And what about the errors (time and voltage)?
 
  • #12
peripatein said:
So are you still inclined to believe it was 1 MOhm?
Furthermore, I would like to estimate the error in time and voltage. Based on the user's guide, the error in time is given as 50ppm whereas the error in voltage is 3% of full scale. Is the error in time then 50 * 10^-6 * resolution? But what is the resolution? 10^-6 sec (based on the graph I previously attached)?
As for the voltage, what is the "full scale"?

Does the user's guide say anything more specific than just "50ppm"? That sounds like a tolerance on its internal timebase clock. 50ppm is 0.005%, so any time measurements that you made with the 'scope could be off by +/-0.005%.

I think by full scale they mean the Vpp value for the waveform. When you measure that on the scope, the number can be off by +/-3% of its actual value.

BTW, did this scope have a current calibration sticker on it? If it's out of cal, the error in amplitude could be bigger. The timebase might be slightly off too, but for your experiments, any timebase errors are probably negligible.
 
  • #13
First of all, should the error in voltage indeed be 0.03*full scale, and not 0.04*full scale? Could you kindly confirm that provided that 4V were applied? Second, since we used 4V, does that mean that delta V is +/- 0.12V (in case it is 3% of full scale, and not 4%)? Third, what would delta t be then? Would it be 0.005% * 1 microsecond (as inferred from the graph I attached, unless I am not analysing the image properly)?
 
  • #14
peripatein said:
First of all, should the error in voltage indeed be 0.03*full scale, and not 0.04*full scale? Could you kindly confirm that provided that 4V were applied? Second, since we used 4V, does that mean that delta V is +/- 0.12V (in case it is 3% of full scale, and not 4%)? Third, what would delta t be then? Would it be 0.005% * 1 microsecond (as inferred from the graph I attached, unless I am not analysing the image properly)?

Sorry, what exactly were you measuring?

The % errors multiply your Vpp measurement, and any time measurements you do. The timebase error should be negligible for measurements like this. Human errors in making the measurements, or noise in the setup would likely dominate.
 
  • #15
It might be negligible, but I still need them for MATLAB to plot a graph and yield errors etc.
Is the time measurement the 200 microseconds at the leftmost corner of the plot I attached? If so, does that mean that the error in time measurement would be 0.005%*200*10^-6??
As we used 4V, would the error in voltage be 3% or 4%? I am not sure I am reading the user's guide properly!
 
  • #16
peripatein said:
It might be negligible, but I still need them for MATLAB to plot a graph and yield errors etc.
Is the time measurement the 200 microseconds at the leftmost corner of the plot I attached? If so, does that mean that the error in time measurement would be 0.005%*200*10^-6??
As we used 4V, would the error in voltage be 3% or 4%? I am not sure I am reading the user's guide properly!

200us/div is not a measurement. What delta-t were you measuring? Maybe a delay from the RC?

I would just use the 3% number on all of the voltages you measure. You should look around at the Agilent website for app notes that talk about DSO accuracy and errors. That will probably make this a lot easier for you. You can even reference the Tek probe app note and any Agilent app notes in your lab report.
 
  • #17
I have looked it up in the user's manual, yet all I could come up with was the 50ppm. I am not exactly sure how to correctly apply that 0.005%! Could you possibly help?
 

1. What is the purpose of determining the resistance of an oscilloscope?

The resistance of an oscilloscope is an important measurement that helps determine the accuracy and reliability of the instrument. It also allows for proper calibration and troubleshooting of the device.

2. How do you determine the resistance of an oscilloscope?

To determine the resistance of an oscilloscope, you will need a multimeter. Connect the multimeter to the input of the oscilloscope and set it to the resistance mode. The reading on the multimeter will indicate the resistance of the oscilloscope.

3. What is the ideal resistance for an oscilloscope?

The ideal resistance for an oscilloscope can vary depending on the specific model and manufacturer. However, most oscilloscopes have a resistance in the range of 1 megaohm (MΩ). This is considered to be the standard resistance for most oscilloscopes.

4. Why is it important to periodically check the resistance of an oscilloscope?

Periodically checking the resistance of an oscilloscope is important because it can change over time due to wear and tear or other factors. This can affect the accuracy and reliability of the instrument, so it is important to calibrate it regularly by checking its resistance.

5. Can the resistance of an oscilloscope be adjusted?

Most modern oscilloscopes have a fixed resistance that cannot be adjusted. However, some older models may have a variable resistance that can be adjusted for calibration purposes. It is important to refer to the manufacturer's instructions for proper calibration procedures.

Similar threads

  • Atomic and Condensed Matter
Replies
5
Views
1K
Replies
6
Views
2K
  • Electrical Engineering
Replies
10
Views
4K
Replies
31
Views
3K
Replies
11
Views
3K
Replies
7
Views
4K
Replies
105
Views
7K
  • Electrical Engineering
Replies
9
Views
898
  • Electrical Engineering
Replies
2
Views
1K
  • Introductory Physics Homework Help
Replies
8
Views
1K
Back
Top