ECG/EMG Electrode Measurement


by abhiroop_k
Tags: bio-signal, ecg or emg, ecg signal, electrode, emg signal, measurement, signal measurement, signal proccessing
abhiroop_k
abhiroop_k is offline
#1
Feb11-14, 12:03 PM
P: 40
I need to collect data via measuring EMG signals in the upper arm. I have standard AgCl stick-on ECG electrodes. My question is, if I can use normal alligator clip cables for collecting this signal from the electrode and feed it into an instru-amp?
Will this affect the signal strength and/or induce noise and if yes, to what extent?

P.S. I need a decently distinguishable signal, to be used later for pattern recognition.
Phys.Org News Partner Engineering news on Phys.org
PsiKick's batteryless sensors poised for coming 'Internet of things'
Researcher launches successful tech start-up to help the blind
Researchers propose network-based evaluation tool to assess relief operations feasibility
berkeman
berkeman is offline
#2
Feb11-14, 12:23 PM
Mentor
berkeman's Avatar
P: 39,612
Quote Quote by abhiroop_k View Post
I need to collect data via measuring EMG signals in the upper arm. I have standard AgCl stick-on ECG electrodes. My question is, if I can use normal alligator clip cables for collecting this signal from the electrode and feed it into an instru-amp?
Will this affect the signal strength and/or induce noise and if yes, to what extent?

P.S. I need a decently distinguishable signal, to be used later for pattern recognition.
Alligator clip cables as opposed to what? Snap-on cables?
abhiroop_k
abhiroop_k is offline
#3
Feb11-14, 12:24 PM
P: 40
Yes, I will be using button electrodes.

hisham.i
hisham.i is offline
#4
Feb15-14, 01:50 AM
P: 177

ECG/EMG Electrode Measurement


Abhiroop,

Such a signal with this small amplitude is very sensitive to noise, so ofcourse aligator clips will introduce sufficient amount of unwanted signal especially from power line source.

In all cases you will need for band pass filter to filter out the noise & get the usefull info from the signal, however the hardest noise to eleminate is the power line noise.

So give it a try & make sure to use a short wire as much as you can in order to reduce the noise.
ScienceGeyser
ScienceGeyser is offline
#5
Feb15-14, 02:24 AM
P: 23
Alligator clips should work just fine. The only issue is contact resistance. With any connector, the spring constant of the contact mechanism, the surface resistance of the mating metals, and the contact area are what determine contact resistance. If you use strong alligator clips, it should be just as good as any other style connector as long as it is sitting still. If you have a dynamic environment (like a stress test on a treadmill) then you hay have some other issues as the contacts move around.
meBigGuy
meBigGuy is offline
#6
Feb15-14, 05:17 PM
P: 961
ScienceGeyser isn't considering that button cables have very very good shields.

Obviously there is no issue with ohmic contact unless there is a lot of motion.

I think you need very good shielding up to and even around the clip. Even the length of the clip itself will pick up noise.
abhiroop_k
abhiroop_k is offline
#7
Feb15-14, 05:20 PM
P: 40
Okay, a basic doubt but, I have individual button cables that have a main and a shield lead. I need to use 3 of these electrode cables together, so do I connect all their shield leads' to a common ground?
meBigGuy
meBigGuy is offline
#8
Feb15-14, 05:36 PM
P: 961
Depends on your receiver. Do a google search on "ECG amplifier circuit" and look over the images. You will see different methodoligies. Study those and decide what you want/need.
ScienceGeyser
ScienceGeyser is offline
#9
Feb15-14, 10:32 PM
P: 23
Most EMG lead sets are not shielded. If you have problems with ambient noise, the most common solution is to twist the wires for as far as possible. Twisting improves the common mode noise rejection that is inherent in the analog front end of most biosignal instruments.
meBigGuy
meBigGuy is offline
#10
Feb16-14, 02:16 AM
P: 961
I stand corrected. I've seen shielded assemblies, so I assumed they were all shielded


Register to reply

Related Discussions
Difference between 2-electrode and 3-electrode cell Classical Physics 0
To test the stability of a reference electrode in a 3 electrode setup Chemistry 0
Actual earth measurement contradicts measurement predicted by special relativity Special & General Relativity 93
electric double layer near electrode and electric field between electrode Classical Physics 2
Cylindrical electrode to rectangular electrode General Physics 0