Are My ECG Filters Meeting the Industry Standard for Cuttoff Frequencies?

  • Thread starter samaaa
  • Start date
  • Tags
    Filters
In summary, the filters cuttoff frequencies should be as the ECG filters standard. The noise is reduced so dramatically by such a small shift in cutoff frequency is impressive and likely due to your use of an 8 order filter.
  • #1
samaaa
48
0
Hi guys

i built an ECG system(preamplifier and filters), but i faced a problem in the filters:

the standard of ECG filters is 0.1 Hz cuttoff frequency for highpass filter and 100 Hz for lowpass filter

but i only obtain a good ECG signal when i make the cuttoff frequency of highpass filter 1 Hz and lowpass of cutoff frequency 80 Hz(8 order)

is my ECG filters correct(1 Hz and 80 Hz)?
or the filters cuttoff frequencies should be as the ECG filters standard?

i ask this question because i am writing a research paper.
 
Engineering news on Phys.org
  • #2
When the cutoff frequencies were at 0.1 and 100 Hz, how is the waveform distorted? If you can post schematics and waveform captures we can better analyze the problem.
 
  • #3
skeptic2 said:
When the cutoff frequencies were at 0.1 and 100 Hz, how is the waveform distorted? If you can post schematics and waveform captures we can better analyze the problem.

ok ,this the ECG signal when the cutoff frequency of lowpass filter is 100 Hz
image.png

and this the ECG signal when the cutoff frequency of lowpass filter is 80 Hz
image.png
 
Last edited:
  • #4
The fact that the noise is reduced so dramatically by such a small shift in cutoff frequency is impressive and likely due to your use of an 8 order filter. First I would consider sources of 120 Hz such as the rectified power supply. Of course it may not be 120 Hz noise and it would help to expand the display to see if you can determine the frequency.
 
  • #5
skeptic2 said:
First I would consider sources of 120 Hz such as the rectified power supply. Of course it may not be 120 Hz noise and it would help to expand the display to see if you can determine the frequency.

i am not using power supply , i use two batteries (of 6 volt)

how can i determine the frequency noise?
and how can i prevent it from Interfere with ECG signal?
 
  • #6
samaaa said:
... how can i prevent it from Interfere with ECG signal?
Looks like you've done that. A schematic of the filters would be nice to see though.
 
  • #7
dlgoff said:
Looks like you've done that. A schematic of the filters would be nice to see though.

i use digital filter(matlab) , my hardwere is only preamplifier.
 
  • #8
Can you use Matlab to calculate the FFT of your waveform? That will give you a good idea what coefficients you should use for your filters.
 
  • #9
skeptic2 said:
Can you use Matlab to calculate the FFT of your waveform? That will give you a good idea what coefficients you should use for your filters.

ok dear i will do that as soon as, thank you
 
  • #10
I agree. The FFT will indicate what undesireable frequency components may be causing the noise. But, not where they are coming from.

The next questions will probably be: What is the sample rate of the system, and what are you using for an anti-alias filter?

For the FFT use as many points as you can and experiment with the windowing function for the best results.
 
  • #11
In a region where the mains frequency is 50Hz, it is very difficult to avoid picking up 50 Hz (and harmonics) on body electrodes even if your entire circuit is battery powered and has no reference to the mains.
 
  • #12
skeptic2 said:
Can you use Matlab to calculate the FFT of your waveform? That will give you a good idea what coefficients you should use for your filters.

meBigGuy said:
I agree. The FFT will indicate what undesireable frequency components may be causing the noise.

this is the FFT of ECG signal
image.png

it very clear that the noise 100 Hz com from power line harmonic

meBigGuy said:
The next questions will probably be: What is the sample rate of the system, and what are you using for an anti-alias filter?
sample rate= 400

i am not using anti-alias filter,because the 100 Hz noise i can remove it in matlab

NascentOxygen said:
In a region where the mains frequency is 50Hz

yes the frequency of power line in my country is 50 Hz

NascentOxygen said:
it is very difficult to avoid picking up 50 Hz (and harmonics) on body electrodes even if your entire circuit is battery powered and has no reference to the mains.
No solution :confused:
 
  • #13
Why do you think the standard cutoff frequency is 100 Hz? Is the standard ECG designed for use in a 60 Hz country and the 100 Hz cutoff would then be effective for 120 Hz? I find it hard to believe that even when the heart is in fibrillation that you would have signals above 80 Hz. Can you think of any reason why an 80 Hz cutoff would not work?
 
  • #14
How many leads are you using for the EKG pickups on the body? Can you post your schematic (as has been mentioned already). Do you have a ground lead as part of your circuit (like below). It's good that you are just battery-powering your circuit!

http://www.cs.wright.edu/~phe/EGR199/Lab_1/fig3.gif

 
Last edited by a moderator:
  • #15
Oh, and I agree that your upper cutoff frequency should ratio lower since the 100Hz number is likely meant for countries that use the 60Hz AC Mains frequency.
 
  • #16
skeptic2 said:
Why do you think the standard cutoff frequency is 100 Hz?
I read in many book, the ecg lowpass filter is 100 hz cutoff frequency

skeptic2 said:
Why do you think the standard cutoff frequency is 100 Hz? Is the standard ECG designed for use in a 60 Hz country and the 100 Hz cutoff would then be effective for 120 Hz?
Yes, I think that

skeptic2 said:
I find it hard to believe that even when the heart is in fibrillation that you would have signals above 80 Hz. Can you think of any reason why an 80 Hz cutoff would not work?
i also see that 80 Hz cutoff work properly but as i mentioned i am writing a research paper, my work must be
agree with a reference book or with another research paper
 
Last edited:
  • #17
berkeman said:
How many leads are you using for the EKG pickups on the body? Can you post your schematic (as has been mentioned already). Do you have a ground lead as part of your circuit (like below).
i use three leads ,one lead is grounded(right lead),
my schematic is like(similar) the schematic you put it:
http://www.cs.wright.edu/~phe/EGR199/Lab_1/fig3.gif

berkeman said:
Oh, and I agree that your upper cutoff frequency should ratio lower since the 100Hz number is likely meant for countries that use the 60Hz AC Mains frequency.

i agree with you too:smile:
 
Last edited by a moderator:
  • #18
The power line waveform is thought of as being a sine wave, but what if there are harmonics present? What happens to those harmonics at your 400 Hz sampling rate?

Also consider what happens if you are sampling at an integer multiple of the power line frequency.
 

What is the industry standard for cutoff frequencies in ECG filters?

The industry standard for cutoff frequencies in ECG filters is typically between 0.05 Hz and 150 Hz.

How can I determine if my ECG filters meet the industry standard for cutoff frequencies?

You can determine if your ECG filters meet the industry standard by consulting the specifications provided by the manufacturer or by performing frequency response tests.

What are the consequences of not meeting the industry standard for cutoff frequencies in ECG filters?

Not meeting the industry standard for cutoff frequencies may result in inaccurate or distorted ECG readings, which can potentially lead to misdiagnosis or incorrect treatment.

What factors can affect the cutoff frequencies in ECG filters?

The cutoff frequencies in ECG filters can be affected by factors such as component tolerances, aging of components, and external interference.

Are there any alternative standards for cutoff frequencies in ECG filters?

Yes, there are alternative standards for cutoff frequencies in ECG filters such as the American Heart Association's recommended frequency range of 0.05 Hz to 100 Hz.

Similar threads

  • Electrical Engineering
Replies
4
Views
796
  • Electrical Engineering
Replies
10
Views
2K
  • Electrical Engineering
Replies
1
Views
2K
  • Electrical Engineering
Replies
6
Views
1K
Replies
2
Views
2K
  • Electrical Engineering
Replies
8
Views
5K
  • Electrical Engineering
Replies
6
Views
2K
  • Introductory Physics Homework Help
Replies
6
Views
2K
Replies
9
Views
1K
Replies
1
Views
3K
Back
Top