Air pressure difference between floor & ceiling

  • #1
Can't figure out how to get my digital manometers to read the pressure difference between the air near the floor vs. at the 8' high ceiling. It reads 0 IWC no matter how I do it. One of my instruments is a micromanometer (reads as low as .001 IWC pressure differential), and it reads 0 IWC there too.
I'm zeroing the meter at floor level, and have no hose on the "reference" port. And I gave a 9 foot manometer hose on the "signal" port with the meter at floor level and the open end of the hose a few feet from the ceiling.
I'm thinking that there be about a 0.108 IWC PD there with that 8' difference in height. Why am I getting 0? It isn't an air tight room, and no fans are on.
 

Answers and Replies

  • #2
russ_watters
Mentor
20,967
7,598
Are you accounting for the air pressure in the manometer tube...?
 
  • #3
Are you accounting for the air pressure in the manometer tube...?

I guess not, if the tube you're referring to is the hose going from the digital manometer to the ceiling. I assumed that pressure in the hose was the same as the air pressure at the ceiling where the open end of that hose is.
Never had to do that just measuring static pressure, i.e. in an air duct. But I guess that was at about the same height as the meter.
How would I account for the pressure in that hose?
 
  • #4
AlephZero
Science Advisor
Homework Helper
7,002
293
I assumed that pressure in the hose was the same as the air pressure at the ceiling where the open end of that hose is.

No. The air pressure in a vertical length of hose will vary with height, exactly the same way as the air pressure in the room.

If you position the hose vertically, seal the end, and then bring it down to floor level you should see a difference in reading. But correcting the reading to give you an accurate measurement would be complcated, especially if the hose is flexible.

The straightforward way to measure this is with an instrument that measures absolute pressure, not relative, and is sensitive enough to show the small pressure difference.
 
  • #5
137
0
Thank you, AZ, as I too was unsuccessfully "measuring" height air dP in my room with a diff. manometer. Now I see why - so simple. I have tried just now, and exactly, as you imply, it is not possible with a flexible hose as a slightest deformation of its walls causes huge pressure fluctuations.
 

Related Threads on Air pressure difference between floor & ceiling

Replies
6
Views
5K
Replies
2
Views
2K
Replies
15
Views
1K
  • Last Post
Replies
3
Views
1K
Replies
2
Views
655
Replies
7
Views
72K
Replies
6
Views
9K
  • Last Post
Replies
4
Views
8K
  • Last Post
Replies
3
Views
1K
Replies
3
Views
3K
Top