Flow meter and mass flow rate

  1. Nov 6, 2017 #1
    Hello all,

    I have an in-house designed flow meter with roughly the configuration in the figure. In the setup, a restriction is placed in the flow stream (the rows of horizontal lines in the figure). Pressure measurements upstream and downstream of the restriction, P1 and P2, respectively are continuously monitored.

    For calibration (determining the relationship between P1-P2 and the mass flow rate), the air flowing through the system is supplied to a nozzle with a flow straightener upstream of the nozzle exit. Velocity measurements with hot-wire anemometry allows me to calculate the mass flow rate at the exit of the nozzle at discrete P1-P2 values.

    38159109296_00225e9f7e_b.jpg

    So that's how far the calibration goes. Now on to how it's used in experiments.

    For some of the experiments that I am conducting there can be a rather significant resistance in the flow (a filter). The backpressure created by this resistance is monitored by a pressure gauge mounted just upstream of P1 and may reach up to 1 bar (gauge). I am uncertain as to how this backpressure affects the flowmeter calibration. For example:

    Let's say from the flow meter calibration, a mass flow rate, m-dota is provided by P1-P2 = Pa.

    If a resistance is fitted downstream, causing backpressure, is m-dota still provided by Pa?

    Thank you in advanced for any response.
     
  2. jcsd
  3. Nov 6, 2017 #2

    DoItForYourself

    User Avatar
    Gold Member

    I think that if the difference ΔP is the same, the m-dota will almost be the same.

    In your case, practically the volumetric flow (Q) depends only on the square root of ΔP. If the density does not change significantly, then the mass flow (m-dota), which equals (density)x(volumetric flow), will be the same.
     
  4. Nov 6, 2017 #3

    BvU

    User Avatar
    Science Advisor
    Homework Helper
    Gold Member

    For an ideal gas the density doubles when the pressure goes from 1 Bar to 2 Bar...

    You have some ##\Delta p## that follows from ##c_d {1\over 2} \rho v^2 ## so you can bet that ##\Delta p(\dot m)## is a different function at different ##p## !
     
  5. Nov 6, 2017 #4

    DoItForYourself

    User Avatar
    Gold Member

    I agree with you, the function Δp(m) is not the same at different pressure.

    But in the case of bigger pressures (for example from 20 to 21 bars), the difference is not so big.
     
  6. Nov 6, 2017 #5

    BvU

    User Avatar
    Science Advisor
    Homework Helper
    Gold Member

    OP is talking about 1 - 2 Bar for the pressure upstream of some filter if I am not mistaken.

    If the gas is ideal enough, the ##\ \Delta p = c_d {1\over 2} \rho v^2\ ## can be worked around using the ideal gas law ##\ \rho = {pM\over RT}\ ## and ##\ v ={ \dot m\over \rho A\ }## leaving something like ##p\,\Delta p \approx {\rm constant} \times\dot m^2##
     
  7. Nov 6, 2017 #6
    Does the flowmeter have turbulent flow, where the pressure drop is proportional to velocity squared? Or is it laminar, where the pressure drop is proportional to velocity?
     
  8. Nov 7, 2017 #7
    The flow is turbulent.
     
  9. Nov 7, 2017 #8

    BvU

    User Avatar
    Science Advisor
    Homework Helper
    Gold Member

    yes. Perhaps you can find some guidance on the site of a flowmeter company like Brooks. Working principle is a little different but calulations are similar.
     
  10. Nov 7, 2017 #9
    Since you have calibration data, you can fit your data to the following equation: Q = K * (delta P / rho)^0.5, where:
    Q is volume flow rate
    K is a least squares fit to your test data
    Delta P is P1 - P2
    rho is the fluid density

    Get the density of air from a psychrometric chart. If your flowmeter operates with P2/P1 greater than 0.99, you can ignore the effects of compressibility. A good source on this is Measurement Systems Application and Design, by Doebelin.
     
  11. Nov 8, 2017 #10
    Thank you all for the responses. You have been very helpful.

    I would like to go back to the original question, if I may:

    I will continue to look deeper into the physics, but is it safe to consider at this point that the answer is no?
     
  12. Nov 8, 2017 #11

    BvU

    User Avatar
    Science Advisor
    Homework Helper
    Gold Member

    No it is not. If that isn't clear from the answers you received, you should seek some assistance with a person who understands these answers.
     