Checking Wiring for 240V 50Hz Mains Variac

In summary: However, if the transformer is saturated, it will still produce excessive current. The light bulb trick will help. Let us know how it works out.In summary, the user is seeking advice on the proper wiring for a mains variac with an input of 240V and 50Hz. They mention that once connected, it blows up the fuse and suspect it may be due to inrush current. They have also tried to experiment with different connections but have not been successful. Other users suggest using a light bulb in series for testing purposes and mention that the variac can be wired for either 125 or 220 volt input. They also mention that the slider may go up to 250 volts with either input and there is a risk of saturation
  • #1
aladinlamp
44
1
hi

can you tell me what should be the proper wiring for this mains variac?
input 240V 50Hz

Once connected, it keeps blowing up the fuse.
I think it is because of the inrush current, so i want to double check with you, that i have it wired correctly.

Thanx
 

Attachments

  • 100_2297.jpg
    100_2297.jpg
    54.3 KB · Views: 675
  • 100_2298.jpg
    100_2298.jpg
    35.1 KB · Views: 647
  • 100_2299.jpg
    100_2299.jpg
    39.5 KB · Views: 657
Engineering news on Phys.org
  • #2
I've worked with variacs before but I've never seen one like this. For testing purposes you can connect a 220 volt light bulb or two 120 volt light bulbs in series with the primary. This will prevent the fuse from blowing and you can experiment with different connections. If the variac is shorted the bulbs will come on with full brightness even with no load on the secondary.
 
  • #3
aladinlamp said:
hi

can you tell me what should be the proper wiring for this mains variac?
input 240V 50Hz

Once connected, it keeps blowing up the fuse.
I think it is because of the inrush current, so i want to double check with you, that i have it wired correctly.

Thanx

The first picture shows *very* clearly that this variac expects a 120V input, and can produce up to a 240V output. The diagram at the right of the first picture shows how to hook it up. Do not put in 240V.
 
  • #4
I think that the Variac can be wired for either 125 or 220 volt input, on the terminals in center picture.. The label very clearly shows this. The common for either is the tab with the "0". Using 240 volt input on thr 220 tab might push the iron into saturation. Use the light bulb in series as per above post by Turtlemeister. For the output, I think "S" in the third picture is the slider, and "0" is the common, which is hard-wired to the input common. The label implies that the slider may go to 250 volts with either 125 or 220 (not 240) volt input. The input is rated at 12 A for either 125 or 220 volt input. There is a lot of inductance here, and I get surges when I turn my 1 KVA Variac on.
 
  • #5
I don't think S is the slider. Look at the first picture. It seems to me that "0" is the common, "A" is for 125 volts, "S" is for 220 volts, "E" is the Earth ground. Don't know where the slider is however; from the pictures that is.

Edit: After a second look, I'm not sure about "S" now. Why wouldn't it be shown in parenthesis like the other terminals? Is that an arrow I see?
 
Last edited:
  • #6
Now I think it is like Berkeman says. Wired for 125 volt with the slider "S" going up to 220 volts. It is an auto-transformer.
 
  • #7
Bob S said:
I think that the Variac can be wired for either 125 or 220 volt input, on the terminals in center picture.. The label very clearly shows this. The common for either is the tab with the "0". Using 240 volt input on thr 220 tab might push the iron into saturation. Use the light bulb in series as per above post by Turtlemeister. For the output, I think "S" in the third picture is the slider, and "0" is the common, which is hard-wired to the input common. The label implies that the slider may go to 250 volts with either 125 or 220 (not 240) volt input. The input is rated at 12 A for either 125 or 220 volt input. There is a lot of inductance here, and I get surges when I turn my 1 KVA Variac on.

I think you might be right, Bob. I've never used a variac like that, putting in a voltage near the top of the autotransformer, but I suppose it may work.

To the OP -- you can use a resistance meter with the variac not connected to any input power source, to measure the resistance of the various windings, and verify that the slider S is doing what is expected.
 
  • #8
berkeman said:
I think you might be right, Bob. I've never used a variac like that, putting in a voltage near the top of the autotransformer, but I suppose it may work.
.
I think my Variac had the option of either having a 0 to 120 v or a 0 to 140 v slider range, depending on how I hooked up the AC. I think there were 3 input tabs; comm, 120 v and 140 v. I built it into a box in 1958, and I don't want to take it apart now (it is still working)..
 
  • #9
It is not uncommon for variacs to have multiple input voltages. The Superior Electric model 236B for example has 120/240 inputs. As Bob has pointed out, putting 240 on the 220 input may be to much and could be what's causing the excessive inrush current. Try connecting 120v to the 125 input. That should give you the same voltage output range but at reduced current on higher voltages.
 

1. What is a 240V 50Hz mains Variac?

A 240V 50Hz mains Variac is an adjustable transformer that is used to vary the voltage and frequency of an electrical circuit. It is commonly used in laboratory and industrial settings to control the voltage and frequency of electrical equipment.

2. Why is it important to check the wiring for a 240V 50Hz mains Variac?

It is important to check the wiring for a 240V 50Hz mains Variac because incorrect or faulty wiring can lead to safety hazards and damage to equipment. It is crucial to ensure that the wiring is properly connected and insulated to prevent electrical shocks and fires.

3. How do I check the wiring for a 240V 50Hz mains Variac?

The first step is to turn off the power supply to the Variac and disconnect it from any other electrical equipment. Then, visually inspect the wiring for any signs of damage or loose connections. Use a multimeter to test the continuity of the wires and make sure they are connected to the correct terminals.

4. What should I do if I find faulty wiring in my 240V 50Hz mains Variac?

If you find faulty wiring in your 240V 50Hz mains Variac, do not attempt to use it until the wiring has been repaired or replaced by a qualified electrician. Using a Variac with faulty wiring can be extremely dangerous and may cause damage to the equipment connected to it.

5. How often should I check the wiring for my 240V 50Hz mains Variac?

It is recommended to check the wiring for your 240V 50Hz mains Variac at least once every six months. However, if the Variac is used frequently or in a harsh environment, it may be necessary to check the wiring more frequently to ensure safety and proper functioning.

Similar threads

Replies
6
Views
2K
Replies
19
Views
2K
  • Electrical Engineering
Replies
9
Views
2K
  • Electrical Engineering
Replies
1
Views
863
  • Electrical Engineering
3
Replies
83
Views
4K
Replies
14
Views
5K
Replies
6
Views
5K
Replies
22
Views
2K
  • Electrical Engineering
Replies
16
Views
4K
Replies
8
Views
2K
Back
Top