I have used three ways. All are difficult.
1) Integrating coil. Wrap 100-200 turns of #40 wire into a coil with ID the same as the magnet OD. Count # turns. Build an integrating circuit with about 1 minute time constant. R series = 100 k, C feedback= 600 uF. Put a 10 meg bleed resistor in parallel with C. choose a low bias current, low
voltage offset op amp.
2) Use a Hall Effect Sensor. Make good differential amplifier op amp to amplify signal. Calibration accuracy of sensor is only about +/- 20%
3) Proton resonance NMR. very accurate. Very difficult to set up.