Hello everybody!

Neurology uses Transcranial Magnetic Stimulation (TMS) for research and sometimes diagnostic and treatment.
Introduction there: http://en.wikipedia.org/wiki/Transcr...ic_stimulation
Some documentation there, but other manufacturers exist: http://www.magstim.com/

As I understand it (but beware I'm bad on biology and neurology), not the magnetic field, but the gradient of electric potential and current it induces, creates the desired effect on the brains. The action resembles partially an electro-shock, but:
  • The skull is tranparent to the magnetic field, while it hampers the electro-shock's current;
  • Hence the skin isn't as brutalized by TMS;
  • The path of the current and the intensity are better controlled;
  • This lets concentrate the effect to some cm2.
Since the gradient of electric potential is the time derivative of the magnetic vector potential A, the average of the gradient over a cycle is zero; I understand a net electrochemical effect is obtained because electrochemical reactions are non-linear (they often show a threshold to voltage) and the pulse is made asymmetric, with a sharp current rise that induces a strong short gradient, and a long current tail that induces a weak long gradient.

A pulse can have 100Ás rise time and 1ms fall time. Series of pulses are also used, sometimes of alternate polarity, but always with pulses of asymmetric transition times.

A good gradient of electric potential needs a strong magnetic induction, and in these non-permeable materials, this means kA in a coil, many kW pulse power, and as a result, the coil is very loud and gets warm, which limits the duration of a session or demands active cooling.

-----

I propose to split each pulse into N shorter ones, scaling like that for understanding:
  • Individual pulses are N times shorter, both at rise and fall time;
  • The reactive volts per coil turn are kept the same;
  • Hence the rate: amps*turns per time unit is unchanged;
  • And the amps*turns are divided by N.
The effect on the brains should keep unchanged when introducing N, because:
  • The electric potential gradient, which results from the rate of change of the magnetic potential, is unchanged; (if you prefer, it varies as the volts per coil turn, which are unchanged)
  • The cumulated duration of the electric potential gradient is unchanged;
  • The proportion between rise and fall time, hence the ratio of electric potential gradient is unchanged.
Beware a few things can go wrong here... For instance if some electrochemical reaction needs a consolidation time before it withstands the inversion of polarity of the electric potential gradient. As usual, I didn't check if this was proposed, tried and possibly abandoned - sorry for that - so interested readers should.

Provided the desired effect is still present, splitting the pulse provides big advantages to the apparatus:
  • The current is divided by N, but the duration remains;
  • Consequently, forces in the coil are divided by N2;
  • The coil's deformation speed after N shorter pulses is divided by N, and the noise power by N2;
  • The energy dissipated in the coil is lowered. By less than N2 because losses increase with frequency;
  • Electronics can deliver N times less current if the voltage is kept, and the energy is N times smaller.
Transition times divided by N may cool down the enthusiasm of electronics designers. As well, it requires probably to brake the coil current actively, for instance with an H bridge, or braking diodes reinjecting current in the power supply, possibly by a secondary winding somewhere. A transformer before the coil can bring advantages: match the cable's and coil's impedance to the power supply and switching components' possibilities; have a separate winding to brake the current; produce inverted pulses by an other primary winding and two switches instead of an H-bridge.

Coils should be made of so-called Litz wire (probably a wrong translation for "braided wire"). In fact, if this isn't already done with 100Ás pulses, it should have been: this limits the heat in the coil but may demand a current braker. The current braker could even be a series resistor, easier to cool than the coil. Shorter pulses demand wire braided more finely, which reduce the wire's section filled by copper, so the energy lost in the coil is divided not quite as quickly as N2.

Marc Schaefer, aka Enthalpy
Phys.Org News Partner Engineering news on Phys.org
Comfortable climate indoors with porous glass
Crash-testing rivets
Customized surface inspection
0xDEADBEEF
#2
Nov4-12, 02:26 PM
P: 824
I would guess that they just connect a fat capacitor to the coil. So the rise time and maximum current are an effect of the R,C and L of the system. If you try to disconnect in the first half of the charging cycle you would produce strong back emf and you would also throw away most of the energy in your capacitor. While if you let it swing, it may even work through a few cycles. Maybe you are right and they use rectangular pulses or something like that. But they are neurologists, I think they just want a massive oomph. Like in electroconvulsive therapy our understanding of the brain is so bad that we do the equivalent of smashing it with a hammer in different places to see if it has an effect or cure a disease.
0xDEADBEEF
#3
Nov4-12, 02:33 PM
P: 824
From magstim.com

The D702 has at least 25% extra stimulating power compared with previous generations of Magstim coil*, while also offering an 85% increase in the number of stimulation pulses feasible at 50% power when energised at 1Hz. This has been validated by independent research in which testing was carried out double blind.
Maybe I am too sceptical of physicians, but this sounds a lot like 25% more oomph to me.

trini
#4
Nov5-12, 09:47 PM
P: 208
Transcranial Magnetic Stimulation

The only thing I wonder about is whether higher voltages are unsuitable for neuro-related devices. The neurologists probably didn't design the devices, they would have told electrical engineers they wanted a field of XX strength. The engineers probably would have done as you suggested, unless the neurologists warned them otherwise. Its just a guess as to why they use higher currents instead though.

Oh and if this was your public disclosure you've got a year to patent, clock's ticking bro ;) Perhaps look at some of the patents on the matter already, see if they mention these issues. gogogogo.
Enthalpy
#5
Nov6-12, 03:49 PM
P: 661
And here's a diagram intended to clarify the changes in the waveforms. Log in if needed, lick to magnify.

I've sketched N=5 for clarity, but real apparatus will probably use a bigger N.

Wave forms are idealized; real-life ones are smoother, and shall be so to protect the components.

I've written Ruhmkorff mode on the last waveforms, by analogy with an ignition coil, but terms like Forward and Flyback mode would sound more familiar to power electronics engineers.
Attached Thumbnails
Waveforms.png  
Enthalpy
#6
Nov6-12, 03:56 PM
P: 661
A well designed TMS coil can have 1ms time constant, real ones have a bit less, but forward operation at N=10 or N=100, with fall time shortened accordingly, needs to brake the current actively. The brake circuitry (a lossy freewheel if you wish) is less than obvious, because the brake voltage is (say 10 times) less than the supply voltage, and the brake circuitry shall not misfire during pulse rise, despite users want pulses of both polarities with few ms between inversions.

Referring to the circuit example sketched below (log in if needed, click to magnify):

An IGBT opened long before the pulse decides the polarity. The TMS coil could have been connected between these collectors, but the second transformer insulates the TMS coil and permits a direct gate control. A big E ferrite core suffices for such a transformer, so I feel this combination easier.

The left IGBT defines the rise time of the current pulse, and the supply voltage defines the rise rate.

On this sketch the resistor sets the current fall time in the coil. Operation at big N is meant to reduce losses in the TMS coil, but here most supply power is dissipated in the resistor. While this power is smaller than at N=1, and the resistor easier to cool than the TMS coil, one may prefer to reinject the energy in the supply. A third winding can help this, but since the voltage and current must already challenge the existing main switch technology, this winding shouldn't step up the voltage, so the power dump rail must have a lower voltage than the supply.

An adjustable power dump rail would control the fall time independently, and a braking voltage constant over the pulse may improve the TMS operation by limiting to a uniform plateau the induced electric potential gradient. The rail must accept and transform power but be pre-loaded before the first pulse. Several braking windings of different voltage ratios may help. I preferred to draw a resistor on the sketch

At high N operation, hence with shorter commutation of a smaller current, MOS must be better than IGBT. Soft commutation (resonance) would be nice, except for the clarity of the sketch

Marc Schaefer, aka Enthalpy
Attached Thumbnails
SwitchSL.png  
Enthalpy
#7
Nov6-12, 04:05 PM
P: 661
Up to now, supplied power increases the current quickly in a TMS coil, which must be the neuro-active phase when the induced electric potential gradient is strong, and a passive circuit lets the current decrease more slowly, in which phase I postulate the electric potential gradient must be below a threshold so it doesn't cancel out the effect obtained in the active phase.

If this model holds, the Flyback operation mode must produce the same result, where the powered phase increases the current slowly in the TMS coil, inducing a limited electric potential gradient, and the neuro-active phase brakes the coil current quickly to induce a strong electric potential gradient. (This must apply as well when the pulse is not split.)

In this other operation mode, the circuitry first supplies a moderate voltage to the coil to accelerate the current, then interrupts the current brutally, which results in a strong electric potential in the coil as well. Very similar to an engine's spark ignition circuit.
This operation mode seems advantageous:
  • The smaller electric potential gradient, which I suppose is the more delicate phase, is better controlled as it results from the supply voltage;
  • The brutal phase, when both U and I are strong, can flow in the circuitry through a diode, which is more robust than a transistor;
  • The circuit example here is simpler.
Referring to the circuit diagram sketched below (log in of needed, click to magnify):

The diodes and transistors must withstand the inductive overvoltage when cutting the current; the supply voltage is much lower, roughly in the same ratio as current fall and rise times in the TMS coil, for instance 10 times lower. The inductive overvoltage is defined through the brake voltage in this example.

The power supply provides less peak power now, but the brake circuit absorbs more, and the transistors must cut the full I at full U, uncomfortable. If the brake windings have as many turns as the accelerating windings, these windings can be coupled very closely to protect the transistors, and can even be the same winding - but with the same number of turns, the brake voltage is much bigger than the supply voltage.

A different choice uses fewer turns at the brake windings, and injects the brake current directly in the supply. This limits to a fixed ratio between the current rise and fall time, which can be switched if additional brake windings provide different numbers of turns. The transistors demand a separate protection then. Anyway, circuitry to protect the components (and preferably soften the transitions) is necessary, though not drawn on the sketch.

Marc Schaefer, aka Enthalpy
Attached Thumbnails
SwitchLS.png  
Enthalpy
#8
Nov6-12, 04:17 PM
P: 661
Quote Quote by 0xDEADBEEF View Post
I would guess that they just connect a fat capacitor to the coil...
O yes! Because their maximum pulse is nearly 3kV and 7kA, leaving no room for extra flourish. Their circuit must be very similar to the one I built to make SmCo magnets: fat capacitor, overloaded thyristor, coil, freewheel diode - except that my pulse was a 10ms half-sine and theirs has to rise in 100Ás because they seek dA/dt and I wanted H.

So the waveforms in the first case are just a simplification of my drawing. Their rise is much a true quarter-sine, and the fall nearly an exponential.

On the improved operation I propose, I'd like the lower-voltage phase to have a true plateau, not an exponential, because this limits the peak voltage there for a given flux and duration, and I suppose the neurologic effect of the pulse comes from non-linear electrochemical response to a strong a short phase versus a weaker and longer one.
Enthalpy
#9
Nov6-12, 04:25 PM
P: 661
Quote Quote by 0xDEADBEEF View Post
...the equivalent of smashing it with a hammer in different places to see if it has an effect...
Sure. The induced electric potential gradient extends well over 5cm for its strongest portion, and 10cm away it's only divided by 3...

In one use, neurologists check whether some muscles can be activated by induction in the brain. It tells if the spine transmits properly. The apparatus is good enough for that, and shows usefulness.

Which does not mean that neurologists are happy with said hammer. While the pulse sequences I propose don't improve the field pattern, the weaker peak voltage during the weak phase may improve operation. And anyway, avoiding the noise and heat associated with 7kA is better.
Enthalpy
#10
Nov6-12, 04:42 PM
P: 661
Quote Quote by trini View Post
...whether higher voltages are unsuitable... they use higher currents instead...
Well, with 3kV and 7kA they are from both sides. I hadn't mentioned the voltage...

My guess is: power components (thyristors) exist for 3kV, several in series is a difficulty, but a <3kA thyristor can dump 7kA, so this choice wasn't bad. They had to keep the cable's weight under control as well.

The coil would begin at 200V more or less, it's the voltage per turn I estimate - but 100kA then, worse choice. More than 3kV would make the apparatus even more dangerous, for neurology just as for any use, and insulator thickness gets fat then.

So within their pulse shape, duration and number, 3kV and 7kA is one reasonable combination. What I hope is that dA/dt matters, not A or B, and a shorter t can be offset through repetition, so A and B and I can be smaller (and then U as well, since the coil's number of turns gives some freedom between I and U).
trini
#11
Nov6-12, 09:17 PM
P: 208
https://www.google.com/?tbm=pts#q=tr...w=1440&bih=799

Have a look there, I filtered it to include utility patents only. Only about 8 pages of results so you can check through them all to see what is applicable to you. I haven't read them but check these:

http://www.google.com/patents/US7854...ed=0CDsQ6AEwAg

http://www.google.com/patents/US7998...ed=0CD4Q6AEwAw

http://www.google.com/patents/US8018...ed=0CDgQ6AEwAQ

You should look at the patents asap before you get into any more theory. Good luck
Enthalpy
#12
Nov7-12, 05:53 PM
P: 661
Here are the losses I computed for eddy currents in wires, because these predominate if thick wire is used at moderate or high frequency.

Omega is 2*pi*F
B the RMS induction
rho the resistivity (18e-9 ohm*m for cold copper)
d the diameter of individual wires
D the diameter of the bundle
l the length and V the volume of the conductors
eta the filling factor
mu the total permeability (pi*4e-7 H/m for vacuum).

A wire, thin enough to let an external field pass through nearly unchanged, dissipates as on the first attached formula (provided I didn't botch it, of course).

A round bundle of thin wires that creates its own induction dissipatesas on the second formula. This holds approximately for an annular coil BUT beware it's for 1 turn... Multiply by the turns squared. Note losses don't depend on the bundle's diameter.

The limit where eddy currents lose as much power as the DC resistance does is on the third forula - still when the bundle creates its own induction. It can be worse with a magnetic core, especially near an air gap.

The parry to eddy currents in conductors is a braided wire ("Litz wire") composed of insulated thinner wires twisted together. Round wires can occupy this fraction of the cylindrical section at best:
7/9 ~78% for a 7-wire Litz
19/25 ~76% for a 19-wire Litz
49/81 ~60% for a 7*7-wire Litz, and so on
knowing that a cylindrical section occupies at best 79% of the winding area of a coil.

Marc Schaefer, aka Enthalpy
Attached Thumbnails
EddyLoss_ThinWire_ExternalField.png   EddyLossResistance_BundleOwnField.png   EddyLossEqualsJoule.png  
Enthalpy
#13
Nov9-12, 09:06 AM
P: 661
A coil can be made for N=100, that is 1Ás and 10Ás rise and fall time (or 10Ás and 1Ás in flyback mode) with small losses. It takes tiny wire, like d=71Ám. Then, braiding it ("Litz wire") in several steps, like 19* and 19* and 7*, achieves a good copper section.

The coil may then have for instance 4 turns, so 500V suffice to induce the desired electric potential gradient. Because of the shorter times, the coil takes only some 350A despite having fewer turns (existing coils must have about 15 turns).

This can simplify the electronics - but times are shorter! Above all, it reduces the heat in the coil by >100 and the noise by a similar factor.

The cable between the generator and the coil must be braided also. It can consist of 4+4 strands, themselves braided in several steps, slightly smaller than the coil┤s one, assembled in an 8-braid that is flexible. Pairs of strands in the 8-braid carry opposite currents. This cable is lighter and loses little power thanks to the smaller current.

Aluminium wire would improve a lot the cable┤s weight, but braided insulated thin aluminium wire must be tailor-made.

Marc Schaefer, aka Enthalpy

p.s. My Internet access is broken, hope to read you all soon.
trini
#14
Nov9-12, 04:39 PM
P: 208
enthalpy, i know what you're working on is just the coil, but maybe what is most needed is a way to focus the field so that it is concentrated in select areas rather than something which hits the whole brain. perhaps look into some kind of wave guide or focusing device? It could also be beneficial in terms of what you're doing as it would require less power if you are focusing the field.
Enthalpy
#15
Nov11-12, 10:50 AM
P: 661
My Internet access is broken right now, my apologies for bad reactivity.

As shorter pulse times reduce the induction far below the present 1-2T, a ferrite core gets possible. Only outside the skull, but it can cut by two the magnetic path length and the current needed.

Smaller current and induction also enable coil shapes not circular. Though, neither induction nor the vector potential can be focussed at distance; at best, a subtractive pattern of currents could make the vector potential sharper. As creating the desired fields gets easier, we may consider subtractive patterns...

In every case (also at 1ms and 10ms), the turns of the coils should be spread apart where a concentrated field is not desired. This reduces the self-inductance hence the necessary voltage - or current if increasing the number of turns.

Marc schaefer, aka enthalpy
Enthalpy
#16
Nov11-12, 11:03 AM
P: 661
Even shorter pulse times seem possible.

With a single thick turn, an 8-shaped coil with both D=70mm loops in series shows about 240nH. It still needs about 200V per turn, but limiting this to 5ns reduces the peak current to 4A approximately, and this is accessible to some RF transmitter transistors.

Then, the transistor, capacitor (or a cable┤s capacity) and flywheel can sit in the coil head, and the cable transmit DC power instead of the strong pulse. Much gets simpler.

Can we exaggerate the pulse duration further? Yes, but with an electromagnetic pulse then, not just magnetic. It must have unsymmetric durations if my explanation holds. This one would fully enable to concentrate the field. More later, maybe.

Marc Schaefer, aka Enhalpy
Enthalpy
#17
Nov18-12, 03:24 PM
P: 661
Here under is a coil enabled by the smaller current and induction: neither round nor flat. The turns are packed close where the effect is sought, and spread apart elsewhere to reduce the inductance.

Usual 8-shaped coils induce undesired potential gradients nearly half as strong under their return paths as under the central zone; the coil sketched here shall minimize them by placing the return paths farther away from the cortex, but near enough to reduce the effect of the main current path outside the central zone.

The angles of the turns (possibly more than 4, possibly of different sizes and shapes) should hence be optimized. I won't do it; a small software needs only to sum over the current path the contributions to the vector potential A to various positions at the target. The Biot and Savart formula is expected in the next message, and in
http://de.wikipedia.org/wiki/Biot-Savart-Gesetz

The active zone being ~45 mm long and the return paths ~85 mm, I estimate the inductance to 560nH with for instance 4 turns, using 4*500 nH/m where the turns are grouped and 1*500 nH/m where they're spread. To achieve the same 2,9 GA*turn/s as the 2* D70mm coil from Magstim fed with 2800 V, the design example takes only 725 MA/s and 406 V, and if the pulse's active duration is just 1Ás, the peak current is only 725 A. Power components may prefer more turns.

The coil's wire consists for instance of commercial Litz wire (visual impression here under) of 33*Awg41 or 71Ám individual wires. 7 such threads are twisted to a strand (for 2m in a lab, fasten them to the ceiling, put weights, and turn) and 7 strands to a conductor of 6.4 mm2 in D~4.5 mm, filled to 40%. Over the coil, 1.5 mohm DC resistance lose 2.8 mJ over one pulse of 1+10 Ás * 725 A, while eddy currents lose roughly 1.6 mJ. Copper weighs 30g, so if hundred 1+10 Ás pulses are one 0.1+1 ms pulse worth, the coil can absorb 2000 equivalents in 80K heating.

The cable may use the same 7*7*33* D=71Ám conductors; 4+4 of them, insulated and braided, plus shielding and mechanical protection, make a 2 m bipolar cable. Copper weighs 0.91 kg; 2.8 mohm DC resistance loses 5.4 mJ and eddy currents some 1.8 mJ. The coil plus the cable lose 15mJ from a 150mJ pulse. Aluminium Litz wire would be highly welcome.

A pair of Toshibas's MG300Q1US51 - and certainly others - look capable of switching the pulse: 1200 V, 300 A, tr=50ns tf=100ns. The forward mode would reduce conduction losses in the IGBT. Regenerative braking reduces the supply's capacitors to ~1 dm3.

Marc Schaefer, aka Enthalpy
(Log in and click on the pictures to magnify)
Attached Thumbnails
CoilDspreadSide.png   CoilDspreadAxial.png   Litz.png  
Enthalpy
#18
Nov18-12, 03:26 PM
P: 661
And here's the fourth picture needing a second message: Biot and Savart's law for the vector potential. You know, A, from B = rot(A). And mu/4pi is 10-7 in S.I.
Attached Thumbnails
BiotSavart.png  


Register to reply

Related Discussions
Artificial polymer muscles that contract with electronic/magnetic stimulation? Materials & Chemical Engineering 2
Stimulation artifact. Biology 0
Between which layers first stimulation energy Atomic, Solid State, Comp. Physics 2
Deep brain stimulation Medical Sciences 4
Stimulation Mg-O2? Chemistry 9