K Type Thermocouple Probe short Circuited with Metallic Body

  • Thread starter Thread starter arham1026
  • Start date Start date
  • Tags Tags
    Insulation Sensor
AI Thread Summary
The K-type thermocouple probe is experiencing a temperature drop to zero when placed in a metallic box, likely due to short-circuiting with the box. Insulating the probe with a thin epoxy coating can help maintain heat transfer while preventing shorting. Users suggest checking the thermocouple's connections and grounding to diagnose potential issues, as an open circuit may also cause erroneous readings. It is recommended to verify the serial data for any indicators of an open circuit condition. Proper grounding and insulation are crucial for accurate temperature measurements in this setup.
arham1026
Messages
1
Reaction score
0
TL;DR Summary
I am Measuring the temperature of metallic box with K thermocouple probe. As soon as probe is place in cavity of metallic box, the temperature readings goes to 0C.
I measure the temperature of metallic box using K thermocouple probe along with MAX6675IC to digitize its output. As soon as probe is placed in the cavity, the temperature drops to zero. I suspect that probe is being shorted with metallic box.Suggest some way to insulate the probe or suggest some other new sensor(contact based )
 
Engineering news on Phys.org
I used to glue small TCs onto electronics (transistor cases and such) with epoxy. A very thin coating was enough insulation, but didn't interfere with heat transfer too much. Not a "probe" just the wires, which is mostly all we ever used. A roll of TC wire is cheap enough that you can just cut the end off and make a new junction or a whole new probe. It wasn't unusual to see a power supply in temperature tests with lots of TC wires attached.
 
Welcome to PF.

One side of the thermocouple pair is grounded, so you should expect a short to result in an offset or noise, rather than zero. Test that condition by grounding the TC probe to T- and to ground.
I expect it is more likely that the TC pair is open circuit, when stretched. Check bit D2 in the serial data. "D2 is normally low and goes high when the thermocouple input is open". Does an open TC circuit return zero data?
Check that T- has a ground connection.
 
Thread 'Weird near-field phenomenon I get in my EM simulation'
I recently made a basic simulation of wire antennas and I am not sure if the near field in my simulation is modeled correctly. One of the things that worry me is the fact that sometimes I see in my simulation "movements" in the near field that seems to be faster than the speed of wave propagation I defined (the speed of light in the simulation). Specifically I see "nodes" of low amplitude in the E field that are quickly "emitted" from the antenna and then slow down as they approach the far...
Hello dear reader, a brief introduction: Some 4 years ago someone started developing health related issues, apparently due to exposure to RF & ELF related frequencies and/or fields (Magnetic). This is currently becoming known as EHS. (Electromagnetic hypersensitivity is a claimed sensitivity to electromagnetic fields, to which adverse symptoms are attributed.) She experiences a deep burning sensation throughout her entire body, leaving her in pain and exhausted after a pulse has occurred...