Questions about a parallel plate capacitor apparatus for lab experiments
DDesulgon
Hello, I am starting to teach about plate capacitors and have a few questions. I am using this model: https://www.phywe.com/physics/elect...tric-field/plate-capacitor-d-260mm_1000_1931/
1) As I understand it, the left plate is electrically insulated, while the right plate is conductively connected to the housing? Does that mean the right plate should definitely be grounded? Is it then safe to touch the housing? This is an important question, especially because the right plate is quite difficult to move and you actually have to hold the housing in place to do so.
2) Is it realistic to assume that there could be any danger? The radius is 13 cm. Assuming a plate distance of a few millimeters, calculating the capacity and the discharge energy with E=1/2*C*U^2, the result is a few mJ (at voltages of several kV). However, discharge energies above 350 mJ are considered dangerous to touch. The power supply is also current-limited. Am I overlooking something here?
3) The charge flows off the plates very quickly. This can be observed very clearly with a static voltmeter. The manufacturer suggests that one possible cause is that only one of the plates is electrically insulated. What is the point of this? Isn't only one plate electrically insulated anyway, while the other is connected to the housing and grounded with a cable?
What measures could be taken to prevent the charges from flowing away so quickly?
Sorry if these are obvious questions—I have little experience with the device and want to be on the safe side.
And sorry again for the clumsy language, English is not my native language.
Thank you in advance!
Science news on Phys.org
The area of you capacitor is about 0.53 square meters.
The minimum non-zero distance is 0.1mm.
That works out to about 4.7pF.
So, with this device set at its minimal resolution, you would only need 12KVDC to store 350mJ.
Thanks for your reply! Unfortunately, I can't follow your math :)
I get 0.053 m^2 for the area. Then, with a plate spacing of 0.1 mm, I get a capacity of 4.7 nF (assuming it's filled with air). However, I only get to 350 mJ at a voltage of approx. 12 kV. Shouldn't it be possible to stay well below this value in all experiments?
It's true that the current limit feature of the power source no longer applies here. But that's what the 350 mJ limit is for, right?
Anyway, if you ground the plate connected to the housing, you shouldn't be at risk when touching the housing, right?
You are right. It is 12KV. In a classroom environment, I would still keep the voltage at or below 50V.
@DDesulgon, you might find the following link useful - it describes an experiment with the sort of capacitor similar to the one you have. It might help to answer some of your questions:
https://ucscphysicsdemo.sites.ucsc.edu/physics-5c6c-demos/parallel-plate-capacitor/
Check the resistance between the insulated plate and the housing. Dirt can sometimes provide a short-circuit at high voltages. So can damp air. (My old physics teacher wouldn't do electrostatic experiments in damp weather as charge would quickly leak (through the damp air) to ground.)
AFAIK the current-limiting feature of the power supply is to limit the maximum continuous current you can get if you get a shock from the supply.
But a (high voltage) power supply can have a significant output capacitance. You can get a short initial shock from the supply's capacitance before the current-limiting kicks in. Maybe that's what the 350mJ refers to.
Also, a capacitor charged to a high voltage (and not to connected to a power supply) can still deliver a nasty shock. The '350mJ' might refer to that.
It sounds like you are unfamiliar with high-voltage electrical experiments, please proceed with extreme care/caution!
Edit: typo'.
Thanks for your reply.
What order of magnitude would be expected for the resulting resistance, and what actions could be taken if the value is lower?
DDesulgon said:
What order of magnitude would be expected for the resulting resistance, and what actions could be tak1n if the value is lower?
If you are using a medium quality general purpose multimeter it will typically measure resistances up to 20M##\Omega##.
I would expect the resistance between the (uncharged and isolated) capacitor plates to be much greater than 20M##\Omega## - more like a few G##\Omega##.
So you shouldn't get an actual reading, just an over-range indication.
If you do get a resistance reading, clean the surface of the mounting (of the fixed plate) with whatever safe solvent you have (e.g. isopropyl alcohol). Then see if that changes the resistance.
Have you got a damp air problem? Or are you in a low humidity environment?
Let us know:
1) what happens;
2) what specific experiments you want to do.
  • · Replies 16 ·
  • · Replies 18 ·
  • · Replies 20 ·
  • · Replies 103 ·
20K
