Can a Photomultiplier be Calibrated for Extremely Low Light Levels?
  • Thread starter Thread starter MereA.
The discussion revolves around the calibration of photomultipliers for measuring extremely low light levels. Participants explore various methods and alternatives for achieving accurate measurements in conditions where light is nearly absent.
  • One participant inquires about the feasibility and methods for calibrating a photomultiplier to detect very low light levels, expressing uncertainty about its effectiveness.
  • Another suggests using an Fe55 source or a pulsed LED source with neutral density filters for calibration.
  • A request for resources on calibration methods using LED sources and ND filters is made.
  • Participants discuss the importance of defining the physical quantity for calibration, with one expressing a preference for lux as the measurement unit.
  • Concerns are raised about the validity of using lux for non-visible light, emphasizing the need to reference the SI definition of candela.
  • One participant notes the challenges of accurately calibrating light levels, suggesting a comparison with a light meter and the use of ND filters to achieve low light conditions.
  • A detailed explanation is provided regarding the calibration process using a dark box, pulsed light sources, and the Poisson distribution of photoelectrons.
  • Alternatives to photomultipliers, such as avalanche photodiodes (APDs), are discussed, highlighting their advantages in certain applications.
  • Concerns are raised about the limitations of photomultipliers in measuring absolute light levels, particularly regarding statistical uncertainties and variations in quantum efficiency.
  • One participant emphasizes the need to clarify the amount of light being measured and the required accuracy for the measurements.
Participants express differing opinions on the effectiveness of photomultipliers for low light measurements, with some advocating for their use while others highlight significant limitations. The discussion remains unresolved regarding the best approach for calibration and measurement.
Participants note the complexities involved in calibrating photomultipliers, including issues related to dark current, thermionic emission, and the statistical nature of photoelectron yield. There are also discussions about the appropriateness of using lux as a measurement unit for various light conditions.
MereA.
Does anyone know how to calibrate a photomultiplier? I need to measure very, very low levels of light (almost no light) and am considering using a photomultiplier. However, I don't know how easy (or difficult) it would be to calibrate a photomultiplier to be sensitive to such a low level of light. If a photomultiplier won't work, are there other options? I've looked at lux meters but could not find any that would measure below 10^-2 lux, which is too much light. Thank you for any help.
Typically an Fe55 source which gives an x-ray photon of very consistent energy which will generate a known number of electrons in the photo cathode.
An simpler method is a pulsed LED source and a stack of ND filters
Do you know if there is a good source (website, book) that would describe the process of calibration using an LED source and ND filters?
In what physical quantity do you want it calibrated?
If possible, I'd like to express the amount of light in terms of lux.
You do know that the lux is a photometric unit of light intensity, right? As such, it is only valid for visible light and you have to find the SI definition of candela (through one standard source that I do not remember from the top of my head).
Light is annoyingly difficult to calibrate accurately - better than the few % level.
Easiest way to do it is to compare a source with a light meter then use a stack of photographic ND filters to drop the source by a known amount for your low light sensor.
Okay. Thank you for your help, NobodySpecial and Dickfore.
2This url has a table that converts lux to photons.
http://www.egc.com/useful_info_lighting.php
For example, from the table, 54 sunlight photons (per m2 per sec?*) is equal to 1 lux. It is possible to calibrate the gain of a photomultiplier tube (PMT) by putting in a dark box, and by using a pulsed light source and neutral density filters (NDFs), get the light level so low that the distribution of number of photoelectrons from the PMT cathode is given by the Poisson distribution (i.e., less than 10 photoelectrons per light pulse) with discrete pulse heights. You will need a gated pulse height analyzer for this. By attenuating the light another factor of 10 using a NDF, nearly every output pulse is due to a single photoelectron. measuring the Coulombs per photoelectron is straight forward. For example, a single photoelecton in a 10-stage PMT might create about 10-13 Coulombs at the anode.
* Something is very fishy. 1 watt of sunlight per square meter is about 2.5 x 1018 photons per square meter, from basic physics.
[added] the table in the url is in mcromoles of photons per m2 per sec, or 6 x 1017 photons per m2 per sec. Using the conversion to watts in the table gives 4.57 x 6 x 1017 = 2.74 x 1018, in agreement with the above physics example.
Bob S
MereA. said:
...If a photomultiplier won't work, are there other options?...
These devices can compete:
http://en.wikipedia.org/wiki/Avalanche_photodiode
You may also want to browse in our own selection of instruments, just in case.
www.gentec-eo.com (from fW to kW)
APDs are generally easier than photo-multipliers if you don't need a very large area. They don't need scary HV supplies, have better red sensitivity and are pretty immune to damage from high light levels
The electronics for them is pretty simple - I had breadboard circuits doing individual photon counting with avalanche mode APDs 20years ago.
Here is a good selection of Hamamatsu photomultipliers:
http://sales.hamamatsu.com/en/produ...s/photomultiplier-tubes/productlist.php?#form
As one example, the Hamamatsu R760 tube has a gain of about 1 x 106, and a dark current of about 1 nA, meaning the dark current is equivalent to about [STRIKE]600[/STRIKE] 6000 cathode photoelectrons per second. If you could gate the signal from the PMT, you could get a clean signal without a dark current background.
Bob S
PMTs are terrible at measuring absolute light levels, particularly high gain tubes. Truly horrible. First, you are limited by the statistics of photoelectron yield at the first dynode, and second, the quantum efficiency of your photocathode varies enormously from tube to tube and even across the face of a single tube. Calibrating on single photons and scaling up introduces its own uncertainties - you have to worry about dark current and thermionic emission (i.e. current from non-light) and emission off the second dynode (i.e. reduced gain).
The only time people use them is when they have no choice. If you go that route, you will be spending 95% of your time understanding your instrumentation and 5% of your time doing your measurement.
One question you should have been asked - "How much light are you measuring?" The other is "How accurately do you need to measure it?"
  • · Replies 21 ·
  • · Replies 35 ·