MRI and Electrons

by dangus
Tags: electrons
Share this thread:
marcusl
#19
Apr1-12, 11:04 AM
Sci Advisor
PF Gold
P: 2,059
Quote Quote by K^2 View Post
And yet, very little of it is used in actual MRI. For the most part, what MRI systems are built to measure are water concentrations and water diffusion.
Not so. Virtually all clinical diagnosis is based on T1- and T2-weighted contrast enhancement, arising from spin-lattice and spin-spin couplings, respectively. To oversimplify and exemplify: T1-weighted images particularly highlight differences in "solidity" between tumor and tissue, meniscus and bone, etc. while T2 weighting emphasizes fluid due to inflammation and is used in spinal imaging because it highlights cerebral spinal fluid so that disks, impingements and bulges are visible.
marcusl
#20
Apr1-12, 11:12 AM
Sci Advisor
PF Gold
P: 2,059
Quote Quote by dangus View Post
I have another question about MRI physics. I was going to make a new thread, but it seemed easier to just post it here since you all seem to know a lot about the topic.

When the perpendicular RF pulse is applied in a MRI machine the net magnetic moment tilts from the Z axis into the X-Y plane. What happens to the individual magnetic moments of all the protons though? Do the individual magnetic moments tip as well?

My previous understanding was that the individual magnetic moments dont change orientation, instead the tipping of the net magnetic moment is caused by the precession of all the protons in phase (which creates the X-Y component) and the elevation of some protons from the low energy state to the high energy state (thus shrinking the net Z component). Is this true?
The classical and quantum mechanical pictures of NMR are sufficiently different to cause confusion if you mix and match the concepts. In the classical view, the spins are tipped away from the z axis by the RF field, whereupon they precess. The longer the field is applied, the greater the polar angle from the z axis. A CW field will send the spins from z to the xy plane (90) to -z (180) and back up to 360 or 0, and then it continues on. To tip them into the xy plane, the RF is turned off.

In the quantum picture, the spins are a two-state quantum system (spin up or down) with an energy gap (the Zeeman splitting). Transitions accompany the absorption or emission of a photon of correct energy. The spins in this view are always aligned or anti-aligned with B0.
dangus
#21
Apr1-12, 11:46 AM
P: 12
Quote Quote by marcusl View Post
The classical and quantum mechanical pictures of NMR are sufficiently different to cause confusion if you mix and match the concepts. In the classical view, the spins are tipped away from the z axis by the RF field, whereupon they precess. The longer the field is applied, the greater the polar angle from the z axis. A CW field will send the spins from z to the xy plane (90) to -z (180) and back up to 360 or 0, and then it continues on. To tip them into the xy plane, the RF is turned off.

In the quantum picture, the spins are a two-state quantum system (spin up or down) with an energy gap (the Zeeman splitting). Transitions accompany the absorption or emission of a photon of correct energy. The spins in this view are always aligned or anti-aligned with B0.
It would seem then that my previous description of the tipping net magnetic moment was the quantum mechanics description of the system then. Correct?

Assuming my description is correct, I am having trouble rationalizing something. In the quantum mechanics view T1 relaxation occurs as the high energy protons drop back down to their low energy state, this increases the number of low energy protons relative to high energy protons and thus regrows the Z component of the net magnetic moment. T2 relaxation occurs as the protons stop precessing in phase and thus their horizontal components are no longer additive.

In this quantum mechanics description the vertical and horizontal components of the magnetic moment are completely independent. Therefore it would seem that T1 relaxation can have no impact on the signal. Even if we wait 0 time between excitation cycles such that no T1 relaxation occurs, the next RF pulse will still bring all the protons into phase (both the high and low energy protons) and the horizontal component will be the same as if we waited a long time between cycles. You see what Im saying? Does it make sense?
dangus
#22
Apr1-12, 12:55 PM
P: 12
Im having another problem with the quantum mechanics view. The classical view says that we can perform a 90 pulse, a 180 pulse a 270 pulse etc with increasingly strong or long RF pulses. But in the quantum mechanics view this doesnt seem feasible. Once the protons are all elevated into the high energy state by a 180 pulse, it is impossible to continue the rotation. All the protons are already elevated to the high energy state, so it doesnt seem like further RF pulsing would do anything.

Furthermore, by the quantum mechanics view, the 180 pulse would also bring all the protons precession into phase and this 180 pulse should have a magnetic moment with a horizontal component, but it doesnt.
K^2
#23
Apr1-12, 03:54 PM
Sci Advisor
P: 2,470
Quote Quote by marcusl View Post
In the quantum picture, the spins are a two-state quantum system (spin up or down) with an energy gap (the Zeeman splitting). Transitions accompany the absorption or emission of a photon of correct energy. The spins in this view are always aligned or anti-aligned with B0.
No, they are not. Take a spin in m=-1/2 in B0, apply a 90 pulse. If you now measure <Sx>, you'll get (1/2)Sin(ωt), while <Sz> is 0.

The orientation of spin in quantum picture follows perfectly the classical picture.
Not so. Virtually all clinical diagnosis is based on T1- and T2-weighted contrast enhancement
In other words, you do proton density measurement with and without spin-echo. You're still not measuring any of the actual spin-spin interactions you've been talking about.

Though, yes, I'll grant you, the difference between T1 and T2 does depend on tissue, so you do get different contrasts because of it. Thing is, you also have T1 and T2 in EPR, and you'd get roughly the same difference between contrasts with and without the echo.
dangus
#24
Apr1-12, 04:41 PM
P: 12
Quote Quote by K^2 View Post
No, they are not. Take a spin in m=-1/2 in B0, apply a 90 pulse. If you now measure <Sx>, you'll get (1/2)Sin(ωt), while <Sz> is 0.

The orientation of spin in quantum picture follows perfectly the classical picture.
I was also under the impression that in the quantum view the spins are only either aligned or antialigned. I obviously dont know enough to argue it one way or the other though, but I am curious to hear what marcusi says. I will share a link though, which is one of the sources I got my information about MRI from and it seems to imply that the spins are only aligned or antialigned. That being said this source could very easily be wrong.

http://en.wikibooks.org/wiki/Basic_P...e#Nuclear_Spin
K^2
#25
Apr1-12, 05:00 PM
Sci Advisor
P: 2,470
Quote Quote by dangus View Post
I was also under the impression that in the quantum view the spins are only either aligned or antialigned.
Not at all. This only happens after a measurement. If I measure the energy of a spin in the magnetic field, I'll get results corresponding to either +1/2 or -1/2, with spin state collapsing to one or the other. But if the magnetization was only aligned or anti-aligned with magnetic field, you'd never get transverse magnetic field, which is what you actually measure in resonance experiment.
marcusl
#26
Apr1-12, 09:58 PM
Sci Advisor
PF Gold
P: 2,059
Quote Quote by marcusl View Post
Not so. Virtually all clinical diagnosis is based on T1- and T2-weighted contrast enhancement, arising from spin-lattice and spin-spin couplings, respectively.
Quote Quote by K^2 View Post
In other words, you do proton density measurement with and without spin-echo. You're still not measuring any of the actual spin-spin interactions you've been talking about.
I don't follow you here. Both T1- and T2-weighted images are typically taken with spin echoes. The interplay of the echo time TE and the repetition time TR determine the contrast.
Quote Quote by K^2 View Post
Though, yes, I'll grant you, the difference between T1 and T2 does depend on tissue, so you do get different contrasts because of it.
Thank you, I'll take your grant--especially since it lies behind most of the clinical utility of MRI.
Quote Quote by K^2 View Post
Thing is, you also have T1 and T2 in EPR, and you'd get roughly the same difference between contrasts with and without the echo.
Same comment as above regarding spin echo. You'd have to look at something special like free radicals to get an EPR signal in the body as Mike H mentioned, giving a very differently slanted image. I've heard of trying to image free radicals created from radiation exposure in cancer therapy using EPR at very low field strength so that RF penetrates. It's not clear to me, though, why you'd get the same tissue contrast in EPR as in MRI.
marcusl
#27
Apr1-12, 10:03 PM
Sci Advisor
PF Gold
P: 2,059
Quote Quote by Mike H View Post
I'm not sure I'd entirely agree with EPR being a sideshow - it really depends on what one's show is in the first place, after all. It's something of a specialized niche, I will agree, although I get the impression more and more people are becoming intrigued by its benefits. It's been invaluable in inorganic & solid state chemistry, and as I've alluded to earlier in this post, for understanding the nature and role of free radicals in a number of contexts. With the development of pulsed EPR methods, one can obtain longer distance constraints for materials and biological assemblies, typically between 20 to 80 Angstroms. Also, it's a natural choice for examining dynamics at the nanosecond time scale.
Sorry for my choice of words. I was thinking of body imaging, and I think most would agree that EPRI is in a preliminary research stage that is very small compared to MRI research and clinical practice.
K^2
#28
Apr1-12, 10:36 PM
Sci Advisor
P: 2,470
I don't follow you here. Both T1- and T2-weighted images are typically taken with spin echoes. The interplay of the echo time TE and the repetition time TR determine the contrast.
You can measure the T2 with a single experiment, without any echo at all. Simply apply 90 pulse, and measure resonance. But yes, if you need to improve signal-to-noise, you can get the same result with an echo and long TR. T1, on the other hand, you can't measure without spin echo at all, because T2 is much shorter.

It's not clear to me, though, why you'd get the same tissue contrast in EPR as in MRI.
I would expect similar dependence of T1 and T2 on tissue type. I'd have to think about it to make sure I can construct a convincing argument. A lot of the same factors play role, though. Diffusion, for example, which contributes to T2 contrast of fluids, should be having similar impact on EPR T2. I'm not as certain about T1.

P.S. I have no experience with medical MRI, and I'm sure it shows. Everything I've done was with NMR spectrometers and some micro-imaging. Though, we did have a sample of a mouse spinal column that fit in the bore of the spectrometer, so the concepts aren't completely alien.
dangus
#29
Apr2-12, 11:43 AM
P: 12
I apologize for the 10 million questions. Hopefully this will be my last. Can someone explain to me exactly how the 180 pulse used to convert a T2* -> T2 signal works. I understand the concept I think.

The analogy of two cars driving at two different constant speeds is often used. If they both start from the same spot and drive for x seconds, then stop, turn around, and drive for x seconds in the opposite direction, they will both arrive back at the start at the same time. They are now "in phase" again.

This makes sense to me. The problem though, is that I dont see how the 180 pulse actually accomplishes this. The protons get out of phase because each experiences a slightly different magnetic field and thus has a slightly different precession frequency. The 180 pulse though doesnt actually reverse the direction of spins, so I dont see how it will bring everything back into phase.
marcusl
#30
Apr2-12, 12:17 PM
Sci Advisor
PF Gold
P: 2,059
Actually the 180 does reverse the direction of spin (that's what a 180 rotation does, after all), rephasing the spins into an echo. The video cartoon in Wikipedia's Spin Echo article is one of many website diagrams that shows how it works.
K^2
#31
Apr2-12, 02:03 PM
Sci Advisor
P: 2,470
I think his confusion stems from the fact that precession direction doesn't change, unlike in the car analogy. But the thing is, if you flip the track, along with the cars, and have cars keep driving in the same direction, it's the same thing as if the cars turned around. That's a bit closer to what the 180 pulse does.
marcusl
#32
Apr2-12, 02:48 PM
Sci Advisor
PF Gold
P: 2,059
Hmm, even that's not quite the best picture. Remember that the slowest car (spin) must end up closest to the finish line after the flip...
I think the wiki cartoon shows it clearly.
dangus
#33
Apr2-12, 06:00 PM
P: 12
Quote Quote by K^2 View Post
I think his confusion stems from the fact that precession direction doesn't change, unlike in the car analogy. But the thing is, if you flip the track, along with the cars, and have cars keep driving in the same direction, it's the same thing as if the cars turned around. That's a bit closer to what the 180 pulse does.
Correct. That was my confusion. After drawing some pictures and staring at it for a bit I see it now. The analogy of flipping the track around was actually quite helpful.

Thank you
K^2
#34
Apr4-12, 10:28 PM
Sci Advisor
P: 2,470
I talked to an actual NMR theorist about all this. He set me straight on some off the issues. While my understanding of T2 contrast wasn't far off, the thing I overlooked with EPR is the frequencies. Basically, molecular movement that's relevant to T2 in NMR is irrelevant in EPR. So EPR T2 contrast is going to be effectively useless.

T1 contrast in EPR will give you some information about the tissues, and it might or might not be useful, but it would be different from information you'd get from T1 using NMR. Again, different frequencies mean that lattice relaxation happens due to completely different vibrational modes.

So marcusl was absolutely correct. As far as medical imaging goes, EPR wouldn't be able to replace NMR.
marcusl
#35
Apr4-12, 11:09 PM
Sci Advisor
PF Gold
P: 2,059
Thanks for the new information, K^2. It's nice to know the reasons behind the differences.


Register to reply

Related Discussions
Electron capture vs internal conversion electrons vs auger electrons High Energy, Nuclear, Particle Physics 1
Are protons and electrons electrical charges? or do protons and electrons HAVE an... General Physics 4
Which force seperates the electrons and nucleus and causes the electrons to orbit? General Physics 14
Max # of electrons in an atom whose highest-energy electrons have n=5 Biology, Chemistry & Other Homework 2
Feynman diagram interpretation - 2 electrons in and 2 electrons out Advanced Physics Homework 2