measuring volume flow rate


Hi
I have a question about measuring volume flow rate with a laminar flow element.
I read the pressure difference across the element (mmH2o) and I get volume flow rate (cfm) from a calibration law. If that calibration law was obtained at standard conditions, would I have to correct it for different conditions? Thanks!
PhysOrg.com
PhysOrg
science news on PhysOrg.com

>> Intel's Haswell to extend battery life, set for Taipei launch
>> Galaxies fed by funnels of fuel
>> The better to see you with: Scientists build record-setting metamaterial flat lens
Sep12-08, 06:31 AM   #2
 
Recognitions:
Science Advisor Science Advisor
Not that I don't believe you, but I find it hard to believe that the unit was calibrated at standard conditions. The calibration was probably referenced to standard conditions. You might want to check the documentation with your unit. The ones we have have correction factors for calculating the actual flow for non standard conditions.
Sep12-08, 10:00 AM   #3
 
calb law is not at standard coditions ;-) I meant calibrated at certain conditions and then used at others.
Sep12-08, 10:21 AM   #4
 

measuring volume flow rate


It depends. Usually if you calibrate a measurement device its for a range of operating states or measurements. Otherwise there wouldn't be much of a point in calibrating it would there? What kind of change in conditions are you referring too? Pressure, temperature, less gravity, what? Are you talking about a wind tunnel or flow venturi or something?
Sep13-08, 04:48 AM   #5
 
Recognitions:
Science Advisor Science Advisor
A laminar flow element is a type of air flow meter. They are sometimes used as calibration masters.
Sep13-08, 09:53 AM   #6
 
Mentor
The biggest thing that typically affects airflow measurements is the density of the air. But what determines if you need to make a correction is how big the error is. If it is less than, say, 5%, then it is probably inside the error margin and repeatability of the device and system.
Sep14-08, 11:48 AM   #7
 
Recognitions:
Science Advisor Science Advisor
That's one of the things about these particular instruments. They are often used as calibration standards. The 5% number is huge. I have one on my desk right now that measures up to 150 scfm. It's not much but it does require corrections even for small variations if you need that level of accuracy. These guys are well beyond pitot tubes.
Sep17-08, 02:27 AM   #8
 
thanks for replying.
Should I correct the air volume flow rate for the inlet pressure ratio? I mean, if the calib. law is
Q [cfm] = (calib. coeff.) * Delta_h [mmH2O]
should it be Q [cfm] = (calib. coeff.) * Delta_h [mmH2O]* P_upstream/P_atm ?
Sep17-08, 01:34 PM   #9
 
Recognitions:
Gold Membership Gold Member
The equations for many properties of air as a function of temperature can be found in ASTM C-680
Sep17-08, 01:46 PM   #10
 
Recognitions:
Gold Membership Gold Member
Also, be very careful with "Standard Conditions" my "Standard" for air properties is 60 F but with a slight change of 8 degrees I almost lost my whole heat exchanger design once.
Thread Closed
Thread Tools


Similar Threads for: measuring volume flow rate
Thread Forum Replies
Flow rate Introductory Physics Homework 1
Flow rate Introductory Physics Homework 1
[SOLVED] Simple volume flow rate calculation Introductory Physics Homework 2
flow rate Introductory Physics Homework 2
Measuring rate of reaction via rate of temperature change Chemistry 3