Detecting drug in ethanol solvent using UV-Vis

  • Thread starter Thread starter AndrewUV
  • Start date Start date
  • Tags Tags
    Ethanol Solvent
Click For Summary
SUMMARY

This discussion focuses on the challenges of detecting a drug dissolved in ethanol using UV-Vis spectroscopy, specifically at the absorbance peaks of 240 nm for the drug and 190 nm for ethanol. The close proximity of these peaks complicates accurate measurements, particularly when the polymer used for encapsulation may also absorb UV light. The presence of dichloromethane (DCM) in the extraction process is identified as a significant source of interference, as DCM absorbs light below 230 nm, potentially skewing results. Recommendations include using quartz cuvettes and considering alternative solvents with a UV cutoff lower than 230 nm for improved accuracy.

PREREQUISITES
  • Understanding of UV-Vis spectroscopy principles
  • Familiarity with absorbance peak identification
  • Knowledge of solvent properties, specifically dichloromethane (DCM) and ethanol
  • Experience with polymer encapsulation techniques
NEXT STEPS
  • Investigate the absorbance characteristics of the polymer in solvents other than DCM
  • Learn about alternative solvents with UV cutoffs lower than 230 nm, such as trifluoroacetic acid
  • Explore methods for improving drug extraction efficiency from polymers
  • Review best practices for UV-Vis measurements, including cuvette selection and sample preparation
USEFUL FOR

Researchers and laboratory technicians involved in drug formulation, analytical chemistry, and UV-Vis spectroscopy who are seeking to optimize detection methods and improve measurement accuracy in complex solvent systems.

AndrewUV
Messages
4
Reaction score
0
Hey Everyone,

I'm having trouble detecting my drug dissolved in ethanol using UV-Vis spectroscopy. The peak absorbance of the drug is at 240 nm while the peak absorbance of the ethanol seems to be around 190 nm. Are these peaks too close to each other to get an accurate measurement? In the wavelength scan of the drug dissolved in solvent, the graph has a higher peak up to 3 AU for ethanol at 190 nm and a lower peak of 1.5 AU at 240 nm for the drug. The peaks, however, are not separated... from the higher ethanol peak at 190 nm, it drops to only around 1.25 AU at 217 nm before going up to the drug peak of 1.5 AU at 240 nm. I've done a standard and that seems to work out fine. However, when I encapsulate the drug inside polymer and then dissolve that polymer to extract the drug again and detect the amount encapsulated inside, I get a number greater than 100%. So I'm detecting more drug than what I put in... would this have to do with how close the peaks are in absorbance and the ethanol peak possibly contributing to the absorbance? Thanks for your help.

Andrew
 
Chemistry news on Phys.org
Are you measuring your solutions in quartz cuvettes? Glass and plastic cuvettes absorb UV and will interfere with your measurements. I don't think ethanol should have any strong absorbance around 240 nm. Are you sure it's not the polymer that absorbs at 190 nm and is interfering with your measurements? You say your standard looks fine. Is this standard also dissolved in ethanol (but with no polymer)?
 
  • Like
Likes   Reactions: AndrewUV
Thanks for the help! Yes the solutions are measured with a quartz 96 well plate. The standard is dissolved in ethanol with no polymer. Hmm yeah that's a good point... it may be the polymer contributing to the measurements. The polymer, however, is not soluble in ethanol. We are dissolving the polymer with DCM and then adding ethanol so the polymer stays in the DCM and the drug is extracted into the ethanol. The two layers are then centrifuged and supernatant removed to be analyzed. Do you think there is a better way to extract the drug from the polymer?

To see if the polymer absorbs near 240 nm I'm thinking of dissolving the polymer in DCM and comparing that to 100% DCM. Does this make sense?
 
AndrewUV said:
We are dissolving the polymer with DCM and then adding ethanol so the polymer stays in the DCM and the drug is extracted into the ethanol.

There's your problem. DCM will absorb light and interfere with your readings below 230 nm (http://macro.lsu.edu/HowTo/solvents/UV%20Cutoff.htm). It would also be good to check to see whether your polymer also absorbs in the UV/vis (but check in a solvent other than DCM, if possible).
 
  • Like
Likes   Reactions: AndrewUV
So even though we are removing the supernatant (ethanol + drug) there might still be DCM present in that supernatant? And if we are looking at the peak at 240 nm, DCM may interfere with that peak? Also, for using a different solvent, would I want to use something with the UV cutoff lower than 230 nm then? eg. trifluoroacetic acid?
 
DCM is fairly soluble in ethanol, so you are probably getting quite a bit of carry over of DCM into the ethanolic phase after extraction. And, yes, using an extraction solvent with UV cutoff lower than 230 nm would be helpful.
 
  • Like
Likes   Reactions: AndrewUV
Great. Thanks so much for the advice! I'll see if using a different solvent works.
 