Detecting drug in ethanol solvent using UV-Vis

  • Thread starter Thread starter AndrewUV
  • Start date Start date
  • Tags Tags
    Ethanol Solvent
Click For Summary

Discussion Overview

The discussion revolves around the challenges of detecting a drug dissolved in ethanol using UV-Vis spectroscopy, particularly focusing on the absorbance peaks of the drug and the solvent. Participants explore issues related to solvent interference, extraction methods, and the potential impact of polymer solubility on measurements.

Discussion Character

  • Technical explanation
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • One participant notes that the peak absorbance of the drug is at 240 nm while ethanol has a peak at 190 nm, raising concerns about the proximity of these peaks affecting measurement accuracy.
  • Another participant suggests that using quartz cuvettes is essential, as glass and plastic cuvettes may absorb UV light and interfere with results.
  • There is a question about whether the polymer might be contributing to absorbance at 190 nm, potentially complicating the analysis.
  • A participant mentions that DCM absorbs light below 230 nm, which could interfere with readings at 240 nm, and suggests checking the polymer's absorbance in a different solvent.
  • Concerns are raised about the possibility of DCM carryover into the ethanolic phase after extraction, which could affect the absorbance measurements.
  • Participants discuss the idea of using a different solvent with a UV cutoff lower than 230 nm to avoid interference.

Areas of Agreement / Disagreement

Participants express multiple competing views regarding the sources of absorbance interference and the effectiveness of different extraction methods. The discussion remains unresolved as participants explore various hypotheses and suggestions without reaching a consensus.

Contextual Notes

Limitations include potential carryover of DCM into the supernatant, the solubility of the polymer in different solvents, and the need to verify the absorbance characteristics of the polymer and solvents used.

AndrewUV
Messages
4
Reaction score
0
Hey Everyone,

I'm having trouble detecting my drug dissolved in ethanol using UV-Vis spectroscopy. The peak absorbance of the drug is at 240 nm while the peak absorbance of the ethanol seems to be around 190 nm. Are these peaks too close to each other to get an accurate measurement? In the wavelength scan of the drug dissolved in solvent, the graph has a higher peak up to 3 AU for ethanol at 190 nm and a lower peak of 1.5 AU at 240 nm for the drug. The peaks, however, are not separated... from the higher ethanol peak at 190 nm, it drops to only around 1.25 AU at 217 nm before going up to the drug peak of 1.5 AU at 240 nm. I've done a standard and that seems to work out fine. However, when I encapsulate the drug inside polymer and then dissolve that polymer to extract the drug again and detect the amount encapsulated inside, I get a number greater than 100%. So I'm detecting more drug than what I put in... would this have to do with how close the peaks are in absorbance and the ethanol peak possibly contributing to the absorbance? Thanks for your help.

Andrew
 
Chemistry news on Phys.org
Are you measuring your solutions in quartz cuvettes? Glass and plastic cuvettes absorb UV and will interfere with your measurements. I don't think ethanol should have any strong absorbance around 240 nm. Are you sure it's not the polymer that absorbs at 190 nm and is interfering with your measurements? You say your standard looks fine. Is this standard also dissolved in ethanol (but with no polymer)?
 
  • Like
Likes   Reactions: AndrewUV
Thanks for the help! Yes the solutions are measured with a quartz 96 well plate. The standard is dissolved in ethanol with no polymer. Hmm yeah that's a good point... it may be the polymer contributing to the measurements. The polymer, however, is not soluble in ethanol. We are dissolving the polymer with DCM and then adding ethanol so the polymer stays in the DCM and the drug is extracted into the ethanol. The two layers are then centrifuged and supernatant removed to be analyzed. Do you think there is a better way to extract the drug from the polymer?

To see if the polymer absorbs near 240 nm I'm thinking of dissolving the polymer in DCM and comparing that to 100% DCM. Does this make sense?
 
AndrewUV said:
We are dissolving the polymer with DCM and then adding ethanol so the polymer stays in the DCM and the drug is extracted into the ethanol.

There's your problem. DCM will absorb light and interfere with your readings below 230 nm (http://macro.lsu.edu/HowTo/solvents/UV%20Cutoff.htm). It would also be good to check to see whether your polymer also absorbs in the UV/vis (but check in a solvent other than DCM, if possible).
 
  • Like
Likes   Reactions: AndrewUV
So even though we are removing the supernatant (ethanol + drug) there might still be DCM present in that supernatant? And if we are looking at the peak at 240 nm, DCM may interfere with that peak? Also, for using a different solvent, would I want to use something with the UV cutoff lower than 230 nm then? eg. trifluoroacetic acid?
 
DCM is fairly soluble in ethanol, so you are probably getting quite a bit of carry over of DCM into the ethanolic phase after extraction. And, yes, using an extraction solvent with UV cutoff lower than 230 nm would be helpful.
 
  • Like
Likes   Reactions: AndrewUV
Great. Thanks so much for the advice! I'll see if using a different solvent works.
 