Gas Field Detectors: Explaining Electron-Balance-Compensation & Chamber Types

Click For Summary
SUMMARY

This discussion focuses on gas field detectors, specifically addressing electron-balance-compensation in gamma ray exposure measurements and the distinctions between free air ionization chambers and wide range ionization chambers. The conversation highlights that ionization chambers convert radiation interactions into electronic signals for dose measurement, with design parameters such as gas type, pressure, electrode geometry, and voltage playing critical roles. The participants emphasize the importance of understanding these factors to optimize detector performance and accuracy in radiation detection.

PREREQUISITES
  • Understanding of ionization chamber technology
  • Familiarity with gamma ray exposure and measurement techniques
  • Knowledge of gas types used in radiation detection
  • Basic principles of electric fields in detection systems
NEXT STEPS
  • Research the principles of electron-balance-compensation in radiation detection
  • Explore the design differences between free air ionization chambers and wide range ionization chambers
  • Investigate the impact of gas pressure and electrode geometry on detector performance
  • Learn about the various types of gas filled detectors and their applications in radiation measurement
USEFUL FOR

Radiation physicists, nuclear engineers, and professionals involved in radiation safety and measurement who seek to deepen their understanding of gas field detectors and their operational principles.

Nucengable
Messages
42
Reaction score
0
Hi, I would appreciate it if someone could explain these expressions to me.
I was reading about Gas field detectors from " Radiation Detection and Measurements" for "Knoll" and paticulary in chapter five
"The gas field detectors are commonly used in dose measurements. In this case, the
detector response should be directed toured the conversion of radiation interactions to detector response in dose scale"

but what I didn't understand what exactly he mean by :

"electron-balance-compensation in the measurement of gamma ray
exposure." ?
and what's the difference between
"free air ionization chamber (for up to 250 keV)" ?
"wide range ionization chamber (cavity chamber)" ?
..
I couldn't find anything on Google and the book didn't explain very well.
Thanks in advance.
 
Engineering news on Phys.org
You might have more luck googling for gas FILLED detectors.

Basically, ionization of some detection gas is turned into an electronic signal that is then interpreted to find the dose.

The type of gas determines what kind of radiation you are sensitive to
 
yeah I know "filled" and "field" , I made copy and paste for the paragraph above , I thought field would mean something different, not sure of a misspelling.
..
I thought the same at the first place , if we have two detectors have the same gas and the same volume , one is detecting a high energy gamma-rays and the other for gamma-rays up to 250 kev , what's the different in the design? this might be a better question ..
..
I think there's could be some collectors from gamma rays inside the detector works somehow to collect gamma-rays or the other particles that produced from the interaction between the gamma-rays and the collector material as indication for the presence of gamma-ray ,
...
and if we have a small volume detector comparing with the range of gamma-rays how possible that detector could give us good measurements about the dose , I guess all these are design parameters...? if you have any idea about this I would appreciate it.
 
One thing that might be different is the gas pressure, the next is the geometry of the electrodes and the voltage across them.

Ionization chambers operate with constant electric fields that just collect the charges generated by the radiation.

Geiger counters operate with much higher fields such that the primary charges get accellerated to the point where they themselves ionize gas and generate more charges (a charge avalanche), thus amplifying the signal, but loosing energy resolution...

There are many different types of gas filled detectors that work with different field geometries and voltages/electric fields strengths. And that is just gas filled detectors. You also have scintillators, semiconductor based detectors, ...
 
Thank you.. :)
 

Similar threads

  • · Replies 24 ·
Replies
24
Views
3K
  • · Replies 1 ·
Replies
1
Views
2K
  • Poll Poll
  • · Replies 1 ·
Replies
1
Views
5K
Replies
4
Views
4K
  • · Replies 44 ·
2
Replies
44
Views
18K
  • · Replies 6 ·
Replies
6
Views
5K
  • · Replies 8 ·
Replies
8
Views
6K
  • · Replies 7 ·
Replies
7
Views
4K
  • · Replies 4 ·
Replies
4
Views
31K
  • · Replies 2 ·
Replies
2
Views
4K