What are Lambda and Z Parameters in Ultrasound B-mode Imaging?

Join the discussion
Ask a follow-up here, or get your own question answered by working scientists, mathematicians and engineers — people, not an autocomplete.
Real named experts · corrections over time · the nuance an AI answer skips
2 replies · 2K views
nao113
Messages
68
Reaction score
13
Homework Statement
1. The transducer array of a linear-array real-time imaging instrument has 32 unfocused elements. Each element is 0.5 cm wide, and there is a nonradiating gap of 0.5 cm between neighboring elements. Only one element at a time is excited.
a) At a frequency of 3 MHz, find the lateral resolution of this instrument at a depth of 1 cm and a depth of 8 cm. The lateral resolution of a linear- array imager is given by the beam size or the line spacing, whichever is larger.
Use relationship:
c = 1540 m/s, c =frequency x l
b) If echo information out to a depth of 12 cm is desired, calculate the minimum time required to scan the entire array.
Relevant Equations
Z < ZR => LR = D
Z > ZR => LR = 2 tan(sin-1 lambda/D) x Z
Screen Shot 2022-05-26 at 19.16.11.png


a. lambda = 513 X 10^-6 m
ZR = 1, 2 cm
Z = 1 cm => Z < ZR => LR = D = 0.5 cm
Z = 8 cm => Z > ZR => LR = 2 tan(sin^-1 (lambda/D)) x Z
= 2 tan(sin^-1 (513 x 10^-6 m/0.005 m)) x 0.08
= 1.65 cm

b. t = Z/c => 0.12/1540 = 77.9 x 10^-6 s
 
on Phys.org
It won't hurt to actually explain what all these letters mean in your equations. This in case you have some question, eventually. So far I don't see the purpose of your post.
 
nasu said:
It won't hurt to actually explain what all these letters mean in your equations. This in case you have some question, eventually. So far I don't see the purpose of your post.
Hello, thank you for the advice, I already changed it, I wonder whether I answered them correctly or not.
 