Calculating Quantum Yields for Fluorescent Dyes

  • Thread starter Thread starter LtStorm
  • Start date Start date
  • Tags Tags
    Quantum
Click For Summary
SUMMARY

This discussion focuses on calculating quantum yields for fluorescent dyes, specifically using anthracene and Rhodamine B as standards. The method involves measuring fluorescence and absorbance, followed by cross-calibrating the two standards using the equation ΦX = ΦST(GradX/GradST)(ηX2ST2). The user reports obtaining quantum yields of 0.346 for anthracene and 0.38 for Rhodamine B, which differ from the literature values of 0.27 and 0.49, respectively. The discussion highlights the importance of using accurate slopes for calibration and maintaining consistent measurement conditions.

PREREQUISITES
  • Understanding of quantum yield calculations
  • Familiarity with fluorescence spectroscopy techniques
  • Knowledge of absorbance measurements
  • Experience with using anthracene and Rhodamine B as fluorescent standards
NEXT STEPS
  • Research the methodology for cross-calibrating fluorescent standards in detail
  • Learn about the effects of solvent refractive index on quantum yield measurements
  • Explore advanced fluorescence spectroscopy techniques for improved accuracy
  • Investigate common pitfalls in quantum yield calculations and how to avoid them
USEFUL FOR

Researchers and laboratory technicians involved in photochemistry, fluorescence spectroscopy, and those working with fluorescent dyes in various applications.

LtStorm
Messages
72
Reaction score
0
So, I'm trying to work out a method of calculating quantum yields for some samples that contain fluorescent dyes.

I've dug around and found a procedure for how to do it, which was essentially the same as doing an extinction coefficient (take a sample, run its fluorescence and absorbance, dilute it, run it again, continue until you have at least six data points) in that you end up with a graph and its slope that you can plug into an equation.

But where I'm having trouble here is that to get an accurate quantum yield, you need to cross-calibrate two different standards. I'm using anthracene (0.27 in ethanol) and Rhodamine B (0.49 in ethanol).

The equation for doing this is;

\Phi_{X}=\Phi_{ST}\frac{Grad_{X}}{Grad_{ST}}\frac{\eta_{X}^{2}}{\eta_{ST}^{2}}

Where 'GradX' is the sample's slope and 'GradST' is the standard's slope. The Etas are for the refractive index of the solvent, which shouldn't be an issue as I've run my standards in ethanol as that was what I could find the literature values for anthracene and Rhodamine B in.

My understanding of what I need to do at this point is thus;

For Anthracene, I use the literature value for Rhodamine B's quantum yield for \Phi_{ST}, then use my slope for Anthracene as GradX and my slope for Rhodamine B for GradST.

Then do the reverse for Rhodamine B. What I feel I'm missing here is the standard's slope, GradST. It feels odd that would also come from my data. But I don't know what else I could possibly use.

Anyone know how to do a cross-calibration for something like this properly? The literature has been little help, and my advisor hasn't been able to work out the issue either.

Also, I have not been getting garbage numbers entirely, they just are quite different from the literature values. Doing my cross-calibration, I get a QY of 0.346 for anthracene (compare to 0.27 lit value) and 0.38 for Rhodamine B (compare to 0.49 lit value).

Here is the exact passage from the instructions I have been following as well;

First, the two standard compounds are cross-calibrated using this equation. This is achieved by
calculating the quantum yield of each standard sample relative to the other. For example, if the
two standard samples are labelled A and B, initially A is treated as the standard (ST) and B as the
test sample (X), and the known ΦF for A is used. Following this, the process is reversed, such
that B is now treated as the standard (ST) and A becomes the test sample (X). In this manner,
the quantum yields of A and B are calculated relative to B and A respectively.


I'm putting this here as it isn't homework, though the moderators may feel free to move it to the homework section if they feel it is more appropriate for that forum still.
 
Last edited:
Chemistry news on Phys.org
The way I have always measured quantum yields is to use :
\Phiunk = \Phistd * (Iunk /Aunk )*(Istd /Astd )*(\etaunk /\etastd )2
where \Phi is the quantum yield, I is the integrated emission intensity, A is the absorbance at the excitation wavelength, and \eta is the refractive index of the solvent
Measure an emission spectrum of the standard sample and the unknown sample, keeping all settings (like excitation wavelength, integration time, slit widths etc.) the same. Apply whatever instrument corrections are necessary. Integrate the area under both emission curves and use in the formula above. For the most accurate results choose a standard that emits in the same wavelength range as your unknown and make the absorbance of the standard and unknown samples at the excitation wavelength between .1 and .2 (optically dilute)
 

Similar threads

  • · Replies 2 ·
Replies
2
Views
3K
  • · Replies 3 ·
Replies
3
Views
3K
  • · Replies 1 ·
Replies
1
Views
1K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 0 ·
Replies
0
Views
2K
Replies
4
Views
9K
  • · Replies 8 ·
Replies
8
Views
2K
Replies
3
Views
4K
  • · Replies 8 ·
Replies
8
Views
3K
  • · Replies 2 ·
Replies
2
Views
4K