Convert Voltage Readings to Radiation Dose Rate (Sievert)

  • Thread starter Thread starter php
  • Start date Start date
Click For Summary
SUMMARY

This discussion focuses on converting voltage readings from a PIN photodiode to radiation dose rates measured in Sieverts. Calibration against known radiation sources is essential to develop a reliable calibration curve that relates voltage to radiation dose. The conversation highlights the complexities involved in accurately converting the energy deposited in the diode to a dose received by a human, particularly due to differences in material composition. For effective radiation monitoring applications, multiple diodes and calibration techniques are recommended to account for various radiation types and ensure accurate readings.

PREREQUISITES
  • Understanding of PIN photodiode operation and characteristics
  • Knowledge of radiation dose concepts: Sieverts and Grays
  • Familiarity with calibration techniques for radiation detectors
  • Basic principles of radiation interaction with matter
NEXT STEPS
  • Research calibration methods for radiation detectors using known sources
  • Learn about the differences in radiation interaction for various materials
  • Explore techniques for developing calibration curves for dose conversion
  • Investigate the use of multiple detectors for gamma-spectrum analysis
USEFUL FOR

Radiation safety professionals, engineers developing radiation monitoring applications, and researchers interested in accurate radiation dose measurement and calibration techniques.

php
Messages
11
Reaction score
0
how do i convert voltage readings, say from a PIN photodiode, to radiation dose rate equivalent (aka. sievert)?

any equation?

any help will be appreciated!
 
Engineering news on Phys.org
Well, with any detector, one would usually calibrate against known sources of different strengths, then with the known source strength and voltage develop a calibration curve relating voltage to Sieverts (radiation equivalent dose) or perhaps in this case and more appropriately grays (radiation absorbed dose).

Siverts and greys have units of energy (joules) per unit mass, and the energy deposited depends on the type of radiation which affects the interaction per unit length of path that the radiation takes through detector. Beyond the detector would be the geometric correction for distance and size of detector which limits the number of radiation particles interacting with the detector.

I presume this is for beta and/or gamma radiation?
 
It depends on what you want to do. As Astro pointed out already, there are several things to consider, and everything depends on the accuracy you want to reach.

The most difficult point will probably be to convert the dose received by the diode (the energy deposited per unit of mass), into a dose, received by a human exposed to the same flux. This would be ok if the PIN diode had an equivalent composition to the human body, but it isn't - it's made out of higher-Z material. As such, it will over-estimate soft gammas and under-estimate hard gammas.

There are ways to solve this, like having several diodes, certain behind a thin lead shield, and others not, so that you can have a rough estimate of the gamma-spectrum. Other problem: you will seriously under-estimate the neutron-induced dose.

Now, if you just want a rough device of the kind "beep when I have to get out of here", then all these considerations are hair-splitting. But if you want to have a good quantitative dosimeter, then you will have a lot of calibration work.
 
It depends on what you want to do.

actually I'm working on a radiation monitoring application that reads information from the gamma detector, processes it and display the monitored area radiation levels in an understandable format (probably in sievert)! so i'll need conversions like those!
 
Last edited:

Similar threads

  • · Replies 4 ·
Replies
4
Views
3K
Replies
2
Views
1K
  • · Replies 5 ·
Replies
5
Views
3K
Replies
0
Views
830
  • · Replies 8 ·
Replies
8
Views
7K
Replies
0
Views
2K
  • · Replies 1 ·
Replies
1
Views
4K
Replies
12
Views
2K
  • · Replies 7 ·
Replies
7
Views
4K
  • · Replies 1 ·
Replies
1
Views
3K