MRI Safety: Question on Potential Burns

  • Context: Graduate 
  • Thread starter Thread starter vabamyyr
  • Start date Start date
  • Tags Tags
    Mri Safety
Click For Summary
SUMMARY

The discussion centers on MRI safety protocols regarding the use of surface coils, specifically the risks associated with leaving unplugged coils in the bore during scans. It is established that unplugged coils can create a conducting loop with the patient and table, potentially inducing high voltage and causing burns or shocks. Additionally, the resonant nature of MRI coils can lead to overheating if not properly connected, resulting in image quality degradation. Properly detuning the coil by connecting it to the system mitigates these risks.

PREREQUISITES
  • Understanding of MRI technology and safety protocols
  • Knowledge of RF fields and their effects on conductive materials
  • Familiarity with coil design and resonance principles in MRI
  • Basic concepts of electrical safety in medical imaging
NEXT STEPS
  • Research MRI coil resonance and its impact on image quality
  • Learn about RF field interactions with conductive materials in MRI
  • Study the safety protocols for MRI scanning with multiple coils
  • Explore the mechanisms of induced voltage in MRI systems
USEFUL FOR

Radiologic technologists, MRI safety officers, medical physicists, and anyone involved in MRI operations and patient safety protocols.

vabamyyr
Messages
65
Reaction score
0
Hi,

MRI safety manual says that whenever you are scanning a patient with a certain surface coil you should remove all unplugged coils from the bore during the scan. For example, if you have a head/neck + spine coil matrix with two coil ports and you are scanning only the spine part then you should either connect the head part also or remove head/neck coil from the bore.
I was wondering what is the physical mechanism that can cause patient burns?
For example, if you scan patient's spine and leave the head coil unplugged. I was thinking in the lines that somehow when the patient is in contact with the unplugged coil it could create a conducting loop between the unplugged coil, patient, patient table maybe and when this loop is in a RF field it causes current to flow in that loop that could cause burns to the patient.
 
Physics news on Phys.org
Thats not burns, but high voltage shock, what is dangerous.

MRI involves very strong magnetic field, so high voltage could get induced on open connectors of the coil as it changes its position with even small patient's moves or when the magnetic field is switched on/off. That voltage could possibly be dangerous if someone touches the connector. The same coil could also catch RF magnetic fields of other coils and thus got inducted voltage.
 
Last edited:
I think the concern would be two-fold, burns and bad images. A MRI coil is a structure which is, by design, resonant at the system's frequency. So, when you are running a pulse sequence that deposits a lot of RF then the coil will be absorbing a lot of energy, perhaps a few kilowatts, and could get hot. Plugging the coil into the table allows the system to "detune" the coil so that it is no longer resonant when the system is transmitting RF. Since it is not resonant, it absorbs little energy, perhaps a few milliwatts, and does not get hot.

With the interference of a resonant structure, even if it doesn't get hot, it should degrade your image quality, i.e. change local flip angles, add noise by inductive coupling to a bigger volume, etc.
 
Thank you DaleSpam for your reply.
 
The administrator at this forum (Bertus) is an MRI specialist who might answer your question authoritatively.

www.allaboutcircuits.com

go well
 

Similar threads

  • · Replies 4 ·
Replies
4
Views
11K
  • · Replies 14 ·
Replies
14
Views
14K
  • · Replies 9 ·
Replies
9
Views
3K
  • · Replies 16 ·
Replies
16
Views
6K
  • · Replies 14 ·
Replies
14
Views
1K
Replies
5
Views
2K
  • · Replies 19 ·
Replies
19
Views
9K
  • · Replies 5 ·
Replies
5
Views
3K
  • · Replies 13 ·
Replies
13
Views
5K
Replies
40
Views
14K