Reason for NMR results not being proportional with C-13

Click For Summary

Discussion Overview

The discussion revolves around the differences in NMR (Nuclear Magnetic Resonance) results for C-13 compared to H-1, particularly focusing on why the intensities of C-13 signals are not proportional to the number of equivalent carbon atoms in a molecule. Participants explore concepts related to nuclear relaxation times, the Nuclear Overhauser Effect (NOE), and the influence of surrounding spins on NMR output.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested

Main Points Raised

  • Michael notes that unlike H-1, the NMR output for C-13 is not proportional to the types of carbons, suggesting that the peaks for related hydrogens differ in height compared to those for carbons.
  • Some participants propose that the intensities of C-13 signals depend significantly on the number of surrounding spins, particularly protons, rather than just the number of equivalent C-13 atoms.
  • Another participant suggests that nuclear relaxation times play a crucial role in the clarity of NMR data, indicating that protons can influence the relaxation time of nearby carbons.
  • One participant explains that the NOE involves the transfer of spin energy between spin systems, complicating the analysis of C-13 spectra due to strong coupling with protons.
  • It is mentioned that factors such as molecular structure and concentration of nearby spins can affect the intensity of C-13 signals, making it difficult to correlate intensity directly with the number of carbon configurations.

Areas of Agreement / Disagreement

Participants express differing views on the primary reasons for the non-proportionality of C-13 NMR signals, with some emphasizing nuclear relaxation times and others focusing on the NOE and surrounding spin influences. The discussion remains unresolved with multiple competing perspectives.

Contextual Notes

Participants reference various factors that may influence NMR results, including relaxation times, molecular structure, and the effects of nearby protons, but do not reach a consensus on the primary cause of the observed phenomena.

paranoid times
Messages
2
Reaction score
0
I am trying to figure out why unlike H-1, NMR output on C-13 is not proportional to types of carbons. Which is to say in a molecule with two CH3 and two CH2 The peaks for the related hydrogens would be one unit tall and 2/3 unit tall. Meanwhile the mentioned carbons would not by nessity have the same peaks at all (even though they are even in number).

Some poking around yielded this:
In further contrast to 1H NMR, the intensities of the signals are not normally proportional to the number of equivalent 13C atoms and are instead strongly dependent on the number of surrounding spins (typically 1H). Spectra can be made more quantitative if necessary by allowing sufficient time for the nuclei to relax between repeat scans.
From wikipedia. I think that is referring to the Nuclear Overhauser effect. Is that correct, if so does anyone have any articles explaining what that is, and why it has this effect on NMR output?

Thanks,
Michael
 
Chemistry news on Phys.org
Ok, so if I've got this right, "relaxation time" referrs to the time taken for an excited nuclear particle to return to it's normal spin. And charged nuclear (like hydrogen) can change the time for a spin to return to normal, generally speaking this change is an decrease of time taken to return to normal spin. And the longer the relaxation time, the more clear the data. As such if you have a methyl group data for the attached carbon will have a higher error (due to the protons attached to it, decreasing its relaxation time) than a second or third degree carbon (which has fewer protons near it to change its relaxation time). And it is just that increased degree of error that leaves the spikes in the data non-relational to the number of particular carbon configurations?
 
NOE is the transfer of spin energy through space from one spin system to another. In a C-13 experiment, the protons have a strong through bond coupling that complicates the analysis tremendously so the protons are generally irradiated continuously to mask that particular coupling. This puts some spin energy into the proton systems. It is the transfer of this spin energy into the C-13 manifold (through space) that affects the intensity of the signals. It also removes the through bond coupling information, simplifying the interpretation of the spectra. As you might imagine, since the effectiveness of the spin transfer is through space, things like tertiary structure (tightly coiled vs open structure, for example), concentration of nearby spin systems (methyl vs methinyl), concentration, aggregation and so forth can affect the spin transfer so it is not as simple as counting nearby protons and correlating intensity but it does follow those trends. Just not enough to be able to integrate peaks.
 