Trouble with Humboldt H12090 aspirator vacuum pump

  • Context: Undergrad 
  • Thread starter Thread starter joniverson
  • Start date Start date
  • Tags Tags
    Pump Vacuum
Click For Summary
SUMMARY

The discussion centers on the challenges faced with the Humboldt H12090 aspirator vacuum pump, which fails to achieve the expected vacuum level of 28-30 Hg, only reaching 16 Hg despite correct setup and no apparent leaks. The user transitioned from a glass aspirator to this metal unit, suspecting either a requirement for higher pressure or a defect in the new pump. Insights reveal that aspirators are sensitive to nozzle design and can vary significantly in performance. Cleaning the internal nozzle is recommended as even minor obstructions can severely impact functionality.

PREREQUISITES
  • Understanding of vacuum systems and their operational principles
  • Familiarity with aspirator design and performance characteristics
  • Knowledge of troubleshooting techniques for vacuum pumps
  • Experience with cleaning and maintenance of laboratory equipment
NEXT STEPS
  • Research the specific design features of the Humboldt H12090 aspirator vacuum pump
  • Learn about the impact of nozzle design on vacuum performance
  • Investigate common troubleshooting methods for vacuum pumps
  • Explore maintenance techniques for laboratory aspirators to ensure optimal performance
USEFUL FOR

Laboratory technicians, researchers conducting vacuum experiments, and anyone involved in the maintenance and operation of vacuum systems will benefit from this discussion.

joniverson
I've been in the process of doing experiments requiring 28- 30 Hg vacuum. For the last couple of weeks, everything seemed to be going well with a glass aspirator I acquired as it would pull the necessary vacuum, when hooked up to the tap, in about 45 seconds. Then, as luck would have it, while trying to unhook the tubing, the small wrench I was using fell and smashed the aspirator. Since the same firm I purchased from no longer had any more, I decided to go with a metal unit with the name in the subject. With virtually no change in the setup, just switching out the broken glass unit for the new metal one, I proceeded to try and draw vacuum tonight but despite all efforts, this until will only pull up to 16 Hg, and that's with both faucets turned to full. I'm not sure what could be wrong. There are no leaks and hook up is correct. Is it possible that this unit requires much more pressure than the prior glass unit, or could it possibly be defective? Anything I could try would be welcome. Thanks in advance.
 
Physics news on Phys.org
I've seen apparently identical aspirators pull wildly different vacuum. They are exquisitely sensitive to their nozzle design. Also, IIRC, some may be designed for 'best flow' and others for 'best ultimate vacuum'. Hmm. Spec suggests this model *should* do the trick...

Have you tried cleaning the internal nozzle ? Only takes a trace of dirt or 'flash' to cripple the hydrodynamics...
 

Similar threads

  • · Replies 10 ·
Replies
10
Views
2K
  • · Replies 9 ·
Replies
9
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 7 ·
Replies
7
Views
17K
  • · Replies 4 ·
Replies
4
Views
3K
  • · Replies 24 ·
Replies
24
Views
4K
  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 4 ·
Replies
4
Views
14K
  • · Replies 17 ·
Replies
17
Views
10K
Replies
1
Views
6K