Why Raman spectroscope doesn't use 532nm laser ?.

  • Context: Graduate 
  • Thread starter Thread starter NEETZGr8
  • Start date Start date
  • Tags Tags
    Laser Raman
Click For Summary
SUMMARY

Raman spectrometers predominantly utilize 785nm or infrared lasers due to the significant interference from fluorescence when using 532nm lasers. While 532nm lasers can reduce costs by allowing the use of silicon CCD detectors and eliminating cooling mechanisms, the fluorescence interference complicates data acquisition. This interference is particularly problematic for organic molecules, which is why 532nm lasers are rarely employed in handheld Raman spectrometers. Alternative laser sources are preferred for applications requiring resonance Raman or higher spatial resolution.

PREREQUISITES
  • Understanding of Raman spectroscopy principles
  • Familiarity with laser types, specifically diode lasers
  • Knowledge of fluorescence interference in spectroscopy
  • Experience with data acquisition and background subtraction techniques
NEXT STEPS
  • Research the impact of fluorescence on Raman spectroscopy data quality
  • Explore the advantages of using 785nm and infrared lasers in Raman applications
  • Learn about resonance Raman spectroscopy and its applications
  • Investigate the design and functionality of handheld Raman spectrometers
USEFUL FOR

Researchers, spectroscopists, and engineers involved in the development and application of Raman spectroscopy, particularly those focused on optimizing laser selection and data acquisition methods.

NEETZGr8
Messages
1
Reaction score
1
Most of the commercial Raman spectrometer uses 785nm or IR laser as excitation source. Though we could use visible lasers like 532 nm in this place, which can reduce the overall cost of the device by

1. Using less power laser( since intensity of raman signal is inversly propotional to the 4th power of wavelength)
2. use silicon ccd detectors, which are less costly compared to the InGAs IR detector
3. No need of detector cooling mechanism

Only issue with 532 nm is the interfernce from fluroscence, which can be removed using proper background subtraction in post data acquisiton.
Then still why 532 nm lasers is not being used in raman spectrometers, especially in hand held raman spectrometer?. Is there any specific reason for not using 532 nm.
Could someone help me with this.
 
Science news on Phys.org
I don't work with Raman spectroscopy but like you said, the usual convention is to avoid fluorescence, especially for organic molecules. I don't think "less power laser" is necessarily true. It largely depends on what kind of laser you are using, but if you are thinking diode lasers, then it generally requires more power to use diode lasers with shorter wavelength than longer wavelength to produce the same number of photons.

There are indeed cases where you want resonance Raman, or working with inorganic material that have significantly larger band gap than organic molecules, or sometimes require better spatial resolution. Then you need different sources.
 