Analyzing signals from a 57Co source using a 5 GS/sec ADC is feasible, but the effectiveness depends on the detector type, with NaI(Tl) being suitable. The pulse width of approximately 50 microseconds means that a 20 ns resolution may not significantly enhance the data quality, necessitating signal amplification before ADC input. For coincidence counting between the 122 keV and 14.4 keV emissions, a more complex electronic setup with multiple channels is required. While some suggest using conventional NIM modules for this purpose, it is possible to utilize a flash ADC with the right configuration. Additional resources on coincidence setups, including TAs, CFDs, and TACs, would be beneficial for further understanding.
Rajini
Messages
619
Reaction score
4
Hello All,
Is it possible to analyze signals/photons/pulses from a 57Co source (strength 2 microCi) using a ADC (multihit readout) with 5 GS/sec? Just to look at pulse height spectrum..also I am interested to make a coincidence counting between 122 adn 14.4 keV levels! I heared that these ADCs are used to count pulse duration of less than 20 ns. But 57Co sources emits pulses in the order of microsec.
Thanks for your input.
Cheers.
 
Physics news on Phys.org
It will depend on what kind of detector you are using; the ADC itself can't detect pulses directly.
Also; what do you mean by "pulse duration" in this context? I assume you are interested in gamma radiation in which case there IS not pulse duration as such (only single photons which do not have a "shape"); the pulse shape will depend entirely on the properties of the detector.
 
Hello,
Yes I use NaI(Tl) detectors. Can I connect these detectors to ADC with 5 GS/sec? Will it work? When you view the signals from detector using a oscilloscope, the pulses look as in page 71 of this pdf document http://www.lanl.gov/orgs/n/n1/panda/00326399.pdf. The x-axis scale is what I meant as pulse duration..in that fig it is 50microsec.
Thanks
 
If your pulse width is 50 us, or 50,000 ns, having 20 ns time resolution will not help you much.
 
You're going to want to amplify your signal before putting it into an ADC, this is described in the document you provided and is a standard procedure and will give a nice signal as shown in fig 4.6. The pulse duration isn't to do with 57Co itself, it's the time response of the detector. A standard spectroscopy amplifier and ADC will have no issue with this.

If you wish to look at coincidences, you'll have to do some more complicated electronic setup than you are attempting at the moment, for a start, you'll need multiple channels in which to do a coincidence.
 
Hello, vanadium/e.bar.qoum:
Yes I want to set a coincidence instrument. 122 keV as start signal (it is the start of the emission of 14.4 kev) and 14.4 keV as stop signal. Although, I have CFDs, SCAs, TAs, etc, I prefer to use the flash ADC with 5 GS/sec. So thought of asking help! Some say it is only possible with conventional NIM modules..Is it true?
Regards,
Rajini
PS: How coincidence setup are done? How the signals are selected using TAs, CFDs, TACs,? Any info/materials/books for this works with pictures and exmaples..
 
I'm setting up an analog power supply. I have the transformer bridge and filter capacitors so far. The transformer puts out 30 volts. I am currently using two parallel power transistors and a variable resistor to set the output. It also has a meter to monitor voltage. The question is this. How do I set up a single transistor to remove whats left of the ripple after the filter capacitors. It has to vary along with the control transistors to be constant in its ripple removing. The bases of the...