Discussion Overview
The discussion revolves around the calibration of photomultipliers for measuring extremely low light levels. Participants explore various methods and alternatives for achieving accurate measurements in conditions where light is nearly absent.
Discussion Character
- Exploratory
- Technical explanation
- Debate/contested
Main Points Raised
- One participant inquires about the feasibility and methods for calibrating a photomultiplier to detect very low light levels, expressing uncertainty about its effectiveness.
- Another suggests using an Fe55 source or a pulsed LED source with neutral density filters for calibration.
- A request for resources on calibration methods using LED sources and ND filters is made.
- Participants discuss the importance of defining the physical quantity for calibration, with one expressing a preference for lux as the measurement unit.
- Concerns are raised about the validity of using lux for non-visible light, emphasizing the need to reference the SI definition of candela.
- One participant notes the challenges of accurately calibrating light levels, suggesting a comparison with a light meter and the use of ND filters to achieve low light conditions.
- A detailed explanation is provided regarding the calibration process using a dark box, pulsed light sources, and the Poisson distribution of photoelectrons.
- Alternatives to photomultipliers, such as avalanche photodiodes (APDs), are discussed, highlighting their advantages in certain applications.
- Concerns are raised about the limitations of photomultipliers in measuring absolute light levels, particularly regarding statistical uncertainties and variations in quantum efficiency.
- One participant emphasizes the need to clarify the amount of light being measured and the required accuracy for the measurements.
Areas of Agreement / Disagreement
Participants express differing opinions on the effectiveness of photomultipliers for low light measurements, with some advocating for their use while others highlight significant limitations. The discussion remains unresolved regarding the best approach for calibration and measurement.
Contextual Notes
Participants note the complexities involved in calibrating photomultipliers, including issues related to dark current, thermionic emission, and the statistical nature of photoelectron yield. There are also discussions about the appropriateness of using lux as a measurement unit for various light conditions.