Gamma spectrometry: Calibration using X-ray emissions

Click For Summary
SUMMARY

The discussion centers on the efficiency calibration of an HPGe detector used in a whole-body counter, utilizing Ba-133 and Eu-152 as calibration nuclides. The participant observed discrepancies in the X-ray emissions, particularly a significantly smaller peak at 45.6 keV from Eu-152 compared to the peaks at 30.8 keV and 35.2 keV from Ba-133. The participant utilized a brick phantom with rod sources and employed Canberra's Genie 2000 software for Gaussian fitting to estimate count rates. The abundance values for the X-ray emissions were sourced from nucleide.org and nucleonica.net, indicating a potential issue with the calibration setup or source distribution.

PREREQUISITES
  • Understanding of gamma spectrometry principles
  • Familiarity with HPGe detector calibration techniques
  • Knowledge of X-ray emission characteristics of nuclides
  • Experience with Canberra's Genie 2000 software for data analysis
NEXT STEPS
  • Research the impact of source distribution on calibration accuracy in gamma spectrometry
  • Learn about the statistical methods for estimating count rates in spectrometry
  • Investigate the calibration procedures for low-energy X-ray emissions
  • Explore the differences in X-ray emission abundances between various nuclides
USEFUL FOR

Gamma spectroscopists, radiation safety professionals, and researchers involved in nuclear medicine or environmental monitoring will benefit from this discussion.

Oliver-BfS
Messages
6
Reaction score
0
I have just done an efficiency calibration measurement for the HPGe detector of a whole-body counter. Ba-133 and Eu-152 were the calibration nuclides. For a calibration at low energies I thought of using the X-ray emissions of both nuclides as well. However, the result looks quite bad - see the attachment.

The peak at 45.6 keV originating from Eu-152 is far too small compared to the X-ray peaks at 30.8 and 35.2 (Ba-133) and at 39.9 (Eu-152). A difference between the emissions of Eu-152 and Ba-133 is possible because of a slightly different distribution of the sources within the phantom. (I used a brick phantom with rod sources.) However, what might be the reason for the difference between the two emissions of Eu-152? The problem occurs also in repeated calibrations so the small number of counts at 45.6 keV is not a random event. The abundance of the X-ray emissions at 45 keV was entered as 14.82%, that at 39.9 keV as 58.5% - values that I found both on nucleide.org and on nucleonica.net. Thank you for your help!
 

Attachments

  • calibration_curve.png
    calibration_curve.png
    2.8 KB · Views: 702
Engineering news on Phys.org
How did you estimate your count rates?
Oliver-BfS said:
The abundance of the X-ray emissions at 45 keV was entered as 14.82%, that at 39.9 keV as 58.5%
Do they both refer to the same base value?
 
The count rates were estimated by fitting Gaussian functions to the peaks. This is done automatically by the software (Canberra's Genie 2000). Summing up the numbers of counts per channel and subtracting the continuum background yields a similar number of counts.

Nucleide.org (PDF-File, page 8) reports the abundance as emitted photons per 100 disintegrations.
 

Similar threads

  • · Replies 17 ·
Replies
17
Views
4K
  • · Replies 2 ·
Replies
2
Views
11K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 2 ·
Replies
2
Views
8K
  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 8 ·
Replies
8
Views
5K
  • · Replies 7 ·
Replies
7
Views
4K
  • · Replies 1 ·
Replies
1
Views
4K
Replies
5
Views
2K