Is my experiment on spectroscopy conceptually correct?

Click For Summary
SUMMARY

The forum discussion centers on a spectroscopy experiment aimed at measuring CO2 concentrations in air using a photodiode array. The proposed method involves passing radiation with wavelengths from 1 to 50 micrometers through an air sample, where CO2 absorbs radiation between 5 to 20 micrometers. To validate the experiment, it is essential to create various concentrations of CO2 in pure air and ensure the photodiode's stability and sensitivity to detect differences in current produced. Additionally, considerations regarding the type of cell used and potential absorption by materials are crucial for accurate measurements.

PREREQUISITES
  • Understanding of spectroscopy principles
  • Knowledge of photodiode operation and characteristics
  • Familiarity with CO2 absorption wavelengths
  • Experience with gas sampling techniques
NEXT STEPS
  • Research methods for creating CO2-free air using sodium hydroxide solutions
  • Learn about photodiode sensitivity and stability testing
  • Investigate the effects of sample path length on spectroscopic measurements
  • Explore materials suitable for constructing spectroscopic cells that do not absorb in the relevant wavelength range
USEFUL FOR

Researchers, environmental scientists, and students involved in spectroscopy, gas analysis, and atmospheric studies will benefit from this discussion.

FedEx
Messages
318
Reaction score
0
Hi,
I wanted to know that whether my following experiment is conceptually correct or not.

I am making a project on spectrosopy. And I have created my files and chart and everything the only thing which is left is a working model.

Now what i am trying to do is to find out the amount of CO2 in air. Suppose we take a sample of air and we pass radiation having the wavelengths ranging from 1 to 50 micrometers, through it. Now CO2 aborbs radiation ranging from 5 to 20 micrometers . Now behind this sample we would keep a photodiode array.Hence the incident radiation would create electricity.

Now we would rehearse the same experiment again,with the absence of CO2 in the air. And this is what i don't know how to do.

Now both the times the current produced by the photodidodes would be different.So now if we study the voltage and magnitude of the current for both the experiments which would certainly be different,can we have some way in order to find out that how much amount of radiation is aborbed by CO2 so that we can find the total amount of CO2 in the gas?

I kow that i would certainly get wise answers and also suggetions which would help me do this experiment properly.
 
Last edited:
Chemistry news on Phys.org
Yes you could do this. It will require you to make up several concentrations of CO2 in pure (no CO2) air. The range should bracket the expected value that you intend to measure. You must then determine if your new spectrometer is sensitive enough to respond to these differences in a reproducible way. Your photodiode must be stable enough to accurately respond to the different concentrations. If it is not, you must tailor the path length in your sample to give you enough CO2 to 'see'. If too much energy is absorbed by the CO2, you must reduce the path length to get enough energy on the photodiode.

Have you thought about the type of cell you will use for this work? Will plastics absorb in this wavelength range?
 
You can make air free of CO2 by bubbling air through a strong solution of base like sodium hydroxide. Pass the air through a drying column (use drierite or magnesium sulfate, anhydrous) and then into your evacuated cell.
 

Similar threads

  • · Replies 2 ·
Replies
2
Views
3K
  • · Replies 16 ·
Replies
16
Views
2K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 17 ·
Replies
17
Views
3K
Replies
19
Views
4K
  • · Replies 35 ·
2
Replies
35
Views
2K
  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 4 ·
Replies
4
Views
4K
  • · Replies 2 ·
Replies
2
Views
14K
  • · Replies 19 ·
Replies
19
Views
10K