OSL Dosimetry and Build Up Region

In summary, the "build up" region is necessary to measure dose in surface of a patient because charged particle equilibrium does not exist there at a given dose.
  • #1
AndresPB
28
1
I understand the following image as follows:

B9781416053163002513_f248-003-9781416053163.jpg


There is a build up region that varies with photons energy, the Dmax for 6 MV photons is around 1.5 cm. The build up region means that kerma > dose in that region, meaning that photons are creating electrons and when the Dmax is reached it is going to be the point where the photons make the most electrons and then after this point the dose will begin to decrease because the intensity of the beam (photons) will diminish because of body attenuation.

I am having troubles relating this explanation with the way Optical Stimulated Light Dosimetry and TLD's work. Isn't it suppose that this kind of dosimetry absorb photons to work (Photons excite electrons in the conduction band of the crystal and they move to the valence band)? Then why is it necessary to use some kind of material for the "build up" region to measure dose in surface of a patient?
With this I mean: In many cases, medical physicists use a bolus equivalent of 1.5 cm of water to simulate this "build-up" region and they put the detector underneath it. Why is this necessary if what we are measuring are photons and not electrons?

Thanks a lot!
 

Attachments

  • B9781416053163002513_f248-003-9781416053163.jpg
    B9781416053163002513_f248-003-9781416053163.jpg
    10.5 KB · Views: 1,539
Biology news on Phys.org
  • #2
Interesting question.

Adding bolus is not "necessary" as a matter of need for the measurements. It's done when one is interested in measuring the Dmax dose, as opposed to the surface dose.

If you're trying to say, make a measurement on a patient to verify a calculation made by your treatment planning system, it's much easier to verify dose to a point where charged particle equilibrium exists. Most modern treatment planning systems will calculate dose at Dmax and downstream very accurately. But look at your graph over the first few mm from the surface. The slopes are rather steep. The steep slopes, plus, uncertainties in your source model (electron contamination, low energy spectra, etc.), plus partial volume effects on your planning CT, can lead to high uncertainties in treatment planning system predictions of surface dose. And you also have to account of the physical properties of your OSL detector and its casing perturbing the beam. This makes it difficult to expect a dosimetry match at the surface.

You can still make an OSL detector measurement without any bolus or buildup material. And there are many reasons why someone might want to do this.

Does that make sense?
 
Last edited:
  • Like
Likes jim mcnamara

1. What is OSL dosimetry?

Optically Stimulated Luminescence (OSL) dosimetry is a method used to measure the amount of radiation exposure an object or person has received. It involves using a special material that can store and release light energy when stimulated by a specific wavelength of light.

2. How does OSL dosimetry work?

In OSL dosimetry, a material is exposed to radiation and some of the energy from the radiation is trapped in the material's crystal lattice. When the material is later exposed to a specific wavelength of light, the trapped energy is released as visible light, which can be measured to determine the amount of radiation exposure.

3. What is the build up region in OSL dosimetry?

The build up region in OSL dosimetry refers to the area around the radiation source where the dose received by the material is not directly proportional to the distance from the source. This is due to the scattering and absorption of radiation in the surrounding materials, which can affect the accuracy of the dose measurement.

4. How is the build up region accounted for in OSL dosimetry?

To account for the build up region in OSL dosimetry, a correction factor is applied to the measured dose. This factor takes into account the effects of scattering and absorption in the surrounding materials and ensures a more accurate measurement of the radiation dose received by the material.

5. What are some applications of OSL dosimetry?

OSL dosimetry is commonly used in medical and industrial settings to measure radiation exposure. It is also used for personal dosimetry for workers who may be exposed to radiation, such as nuclear power plant employees. Additionally, OSL dosimetry can be used to date archaeological artifacts or geological samples by measuring the amount of radiation they have been exposed to since their creation.

Similar threads

  • Quantum Physics
Replies
17
Views
1K
  • STEM Career Guidance
Replies
2
Views
2K
Replies
152
Views
5K
  • Quantum Physics
3
Replies
81
Views
7K
  • STEM Academic Advising
Replies
1
Views
1K
  • Sci-Fi Writing and World Building
Replies
6
Views
2K
Replies
5
Views
3K
Replies
4
Views
8K
  • Other Physics Topics
Replies
10
Views
2K
  • Sci-Fi Writing and World Building
Replies
2
Views
2K
Back
Top