What do the gradient magnets do on MRI scanners?

  • Context: Undergrad 
  • Thread starter Thread starter kkid
  • Start date Start date
  • Tags Tags
    Gradient Magnets Mri
Click For Summary
SUMMARY

Gradient magnets in MRI scanners are essential for spatial localization of protons within the scanned volume. They modify the static magnetic field, allowing the Larmor frequency of spins to vary based on their position. This variation enables the differentiation of signals from different locations, facilitating 3D imaging. By applying a field gradient, the resonance frequencies of spins are spread across the chosen direction, crucial for accurate imaging results.

PREREQUISITES
  • Understanding of Larmor frequency in MRI technology
  • Familiarity with magnetic field gradients
  • Knowledge of proton alignment in magnetic fields
  • Basic principles of radiofrequency (RF) in MRI
NEXT STEPS
  • Research the mathematical principles behind magnetic field gradients in MRI
  • Explore the role of RF pulses in MRI signal acquisition
  • Learn about 3D imaging techniques in MRI
  • Investigate the effects of different gradient strengths on image quality
USEFUL FOR

Medical imaging professionals, MRI technicians, and physicists interested in the technical aspects of MRI technology and imaging techniques.

kkid
Messages
3
Reaction score
0
I am trying to understand MRI scanners.

I know that MRIs work by aligning the protons in the direction of the large magnetic field and the radio frequency sets the frequency of the oscillations to the lamour frequency - also raising its energy level. Then when the RF is switched off, the energy level drops by releasing radio waves instead which are detected by the same coil that emitted them in the first place.


I do not see where the gradient magnets come into this. I know the help with location of the atoms but how?



Thanks.
 
Physics news on Phys.org
RF and the permanent magnetic field strength need some specific relation to get a signal. With a gradient field, this happens in a small region of the scanned volume only - therefore, you know where the signal comes from.
 
The Larmor frequency of a spin is proportional to the magnetic field it is sitting in, call it \mathbf{B}_z=\mathbf{\hat z} B_0. All spins sit in the same field and share the same resonance. If you now add a field gradient onto the static B_0 field \mathbf{B_z}=\mathbf{\hat z}\left( B_0 + \frac{\partial B}{\partial a}\right), where a = x, y or z, then you spread the frequencies along the chosen direction. That is, since you've made the z field depend linearly on position in one direction, spins at different positions in that direction resonate at different frequencies. You have taken the first step towards localizing (i.e. imaging) the spins in a 3D body. Similar games are played for to localize in the other two directions.
 

Similar threads

  • · Replies 9 ·
Replies
9
Views
3K
  • · Replies 6 ·
Replies
6
Views
2K
Replies
4
Views
3K
  • · Replies 2 ·
Replies
2
Views
2K
  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 11 ·
Replies
11
Views
3K
  • · Replies 14 ·
Replies
14
Views
3K