Why Use Interpolation Over Extrapolation for Calibration Curves?

Join the discussion
Registration is free. Ask a follow-up in this thread, or start your own.
4 replies · 15K views
Beer-monster
Messages
285
Reaction score
0
Could someone help me understand why interpolation rather than extrapolation should be used for calibration curves (with the exception of Standard additions). I know that extrapolation is less precise, but the book I've got doesn't over any more than that, and I think I need a little more detail for an exam.:biggrin:
 
Physics news on Phys.org
The relationship between concentration and signal strength for the analyte.

Would I be right in saying that you can't extrapolate as you can't be sure the curve remains linear (e.g because of self-absorption in ICP-AES)?
 
Beer-monster said:
Would I be right in saying that you can't extrapolate as you can't be sure the curve remains linear (e.g because of self-absorption in ICP-AES)?
That's exactly right !
 
Wow, I feel so very stupid *smacks head*