Can a type K thermocouple be used with cryo?

  • Thread starter Thread starter jonhswon
  • Start date Start date
  • Tags Tags
    Thermocouple
Join the discussion
Ask a follow-up here, or get your own question answered by working scientists, mathematicians and engineers — people, not an autocomplete.
Real named experts · corrections over time · the nuance an AI answer skips
4 replies · 4K views
jonhswon
Messages
10
Reaction score
4
Hi.

I can not understand why type K tc can not be used with cryogenic temperatures?

If for example calibrated with liquid nitrogen and ice, can it be used from 0 C to -200 C accurately?

I understand that there is a non-linear characteristic a lower temperatures, but then one could make a voltage-to-temp plot by calibrating against ice, liquid nitrogen, solid ice, etc.. and obtain a curve.

Does the inaccuracy arise simply because they are usually calibrated with ice and boiling water only, then linear interpolation/extrapolation line is drawn, OR is there something else that I am missing?
 
Engineering news on Phys.org
Maybe the metals or the alloy junction are superconductors at cryogenic temperatures.

Does the Seebeck effect vanish, or the calibration change, in superconductive metals?
 
Baluncore said:
Maybe the metals or the alloy junction are superconductors at cryogenic temperatures.

Does the Seebeck effect vanish, or the calibration change, in superconductive metals?

No, not at the temperatures mentioned in the OP (liquid nitrogen). There are very few metalsa/alloy that become superconducting even at liqduid helium temperatures (Nb, NbN, Pb and some alloys); there are none at liquid nitrogen temperatures.
For the type of metals that thermocouples are made you typically have to go to mK temperatures; and not all metals become superconducting even then.
 
The Lakeshore link has an error (I think): The page claims a Type K sensitivity of 4.1 mV/K @ 20K. I think they meant 'uV.'
 
Reply