The discussion centers on the cutoff frequencies of ECG filters, specifically the highpass filter set at 1 Hz and the lowpass filter at 80 Hz, diverging from the standard 0.1 Hz and 100 Hz. Users highlight that the 80 Hz cutoff effectively reduces noise, particularly from power line harmonics at 50 Hz, while the standard settings may distort the ECG waveform. The conversation emphasizes the importance of using MATLAB for FFT analysis to identify noise frequencies and optimize filter coefficients.

PREREQUISITES
  • Understanding of ECG signal processing and filtering techniques
  • Familiarity with MATLAB for digital signal analysis
  • Knowledge of FFT (Fast Fourier Transform) for frequency analysis
  • Basic electronics principles, including filter design and noise sources
NEXT STEPS
  • Learn how to implement FFT in MATLAB for ECG signal analysis
  • Research the impact of filter order on signal integrity and noise reduction
  • Investigate the effects of sampling rates on filter performance and aliasing
  • Explore the design and application of anti-aliasing filters in ECG systems
USEFUL FOR

Electronics engineers, biomedical engineers, researchers in ECG technology, and anyone involved in the design and optimization of ECG filtering systems.

samaaa
Messages
48
Reaction score
0
Hi guys

i built an ECG system(preamplifier and filters), but i faced a problem in the filters:

the standard of ECG filters is 0.1 Hz cuttoff frequency for highpass filter and 100 Hz for lowpass filter

but i only obtain a good ECG signal when i make the cuttoff frequency of highpass filter 1 Hz and lowpass of cutoff frequency 80 Hz(8 order)

is my ECG filters correct(1 Hz and 80 Hz)?
or the filters cuttoff frequencies should be as the ECG filters standard?

i ask this question because i am writing a research paper.
 
Engineering news on Phys.org
When the cutoff frequencies were at 0.1 and 100 Hz, how is the waveform distorted? If you can post schematics and waveform captures we can better analyze the problem.
 
skeptic2 said:
When the cutoff frequencies were at 0.1 and 100 Hz, how is the waveform distorted? If you can post schematics and waveform captures we can better analyze the problem.

ok ,this the ECG signal when the cutoff frequency of lowpass filter is 100 Hz
image.png

and this the ECG signal when the cutoff frequency of lowpass filter is 80 Hz
image.png
 
Last edited:
The fact that the noise is reduced so dramatically by such a small shift in cutoff frequency is impressive and likely due to your use of an 8 order filter. First I would consider sources of 120 Hz such as the rectified power supply. Of course it may not be 120 Hz noise and it would help to expand the display to see if you can determine the frequency.
 
skeptic2 said:
First I would consider sources of 120 Hz such as the rectified power supply. Of course it may not be 120 Hz noise and it would help to expand the display to see if you can determine the frequency.

i am not using power supply , i use two batteries (of 6 volt)

how can i determine the frequency noise?
and how can i prevent it from Interfere with ECG signal?
 
samaaa said:
... how can i prevent it from Interfere with ECG signal?
Looks like you've done that. A schematic of the filters would be nice to see though.
 
dlgoff said:
Looks like you've done that. A schematic of the filters would be nice to see though.

i use digital filter(matlab) , my hardwere is only preamplifier.
 
Can you use Matlab to calculate the FFT of your waveform? That will give you a good idea what coefficients you should use for your filters.
 
skeptic2 said:
Can you use Matlab to calculate the FFT of your waveform? That will give you a good idea what coefficients you should use for your filters.

ok dear i will do that as soon as, thank you
 
  • #10
I agree. The FFT will indicate what undesireable frequency components may be causing the noise. But, not where they are coming from.

The next questions will probably be: What is the sample rate of the system, and what are you using for an anti-alias filter?

For the FFT use as many points as you can and experiment with the windowing function for the best results.
 
  • #11
In a region where the mains frequency is 50Hz, it is very difficult to avoid picking up 50 Hz (and harmonics) on body electrodes even if your entire circuit is battery powered and has no reference to the mains.
 
  • #12
skeptic2 said:
Can you use Matlab to calculate the FFT of your waveform? That will give you a good idea what coefficients you should use for your filters.

meBigGuy said:
I agree. The FFT will indicate what undesireable frequency components may be causing the noise.

this is the FFT of ECG signal
image.png

it very clear that the noise 100 Hz com from power line harmonic

meBigGuy said:
The next questions will probably be: What is the sample rate of the system, and what are you using for an anti-alias filter?
sample rate= 400

i am not using anti-alias filter,because the 100 Hz noise i can remove it in matlab

NascentOxygen said:
In a region where the mains frequency is 50Hz

yes the frequency of power line in my country is 50 Hz

NascentOxygen said:
it is very difficult to avoid picking up 50 Hz (and harmonics) on body electrodes even if your entire circuit is battery powered and has no reference to the mains.
No solution :confused:
 
  • #13
Why do you think the standard cutoff frequency is 100 Hz? Is the standard ECG designed for use in a 60 Hz country and the 100 Hz cutoff would then be effective for 120 Hz? I find it hard to believe that even when the heart is in fibrillation that you would have signals above 80 Hz. Can you think of any reason why an 80 Hz cutoff would not work?
 
  • #14
How many leads are you using for the EKG pickups on the body? Can you post your schematic (as has been mentioned already). Do you have a ground lead as part of your circuit (like below). It's good that you are just battery-powering your circuit!

http://www.cs.wright.edu/~phe/EGR199/Lab_1/fig3.gif

 
Last edited by a moderator:
  • #15
Oh, and I agree that your upper cutoff frequency should ratio lower since the 100Hz number is likely meant for countries that use the 60Hz AC Mains frequency.
 
  • #16
skeptic2 said:
Why do you think the standard cutoff frequency is 100 Hz?
I read in many book, the ecg lowpass filter is 100 hz cutoff frequency

skeptic2 said:
Why do you think the standard cutoff frequency is 100 Hz? Is the standard ECG designed for use in a 60 Hz country and the 100 Hz cutoff would then be effective for 120 Hz?
Yes, I think that

skeptic2 said:
I find it hard to believe that even when the heart is in fibrillation that you would have signals above 80 Hz. Can you think of any reason why an 80 Hz cutoff would not work?
i also see that 80 Hz cutoff work properly but as i mentioned i am writing a research paper, my work must be
agree with a reference book or with another research paper
 
Last edited:
  • #17
berkeman said:
How many leads are you using for the EKG pickups on the body? Can you post your schematic (as has been mentioned already). Do you have a ground lead as part of your circuit (like below).
i use three leads ,one lead is grounded(right lead),
my schematic is like(similar) the schematic you put it:
http://www.cs.wright.edu/~phe/EGR199/Lab_1/fig3.gif

berkeman said:
Oh, and I agree that your upper cutoff frequency should ratio lower since the 100Hz number is likely meant for countries that use the 60Hz AC Mains frequency.

i agree with you too:smile:
 
Last edited by a moderator:
  • #18
The power line waveform is thought of as being a sine wave, but what if there are harmonics present? What happens to those harmonics at your 400 Hz sampling rate?

Also consider what happens if you are sampling at an integer multiple of the power line frequency.
 

Similar threads

  • · Replies 10 ·
Replies
10
Views
2K
  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K
Replies
2
Views
3K
  • · Replies 6 ·
Replies
6
Views
2K
  • · Replies 6 ·
Replies
6
Views
3K
  • · Replies 8 ·
Replies
8
Views
5K
  • · Replies 6 ·
Replies
6
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 2 ·
Replies
2
Views
4K