Cavity locking using Lockin+PID technique

  • Level: Graduate 
  • Thread starter Hashirama
  • Start date
  • Tags Tags
    Cavity Pid
2 replies · 5K views · last reply
Hashirama
Messages
2
Reaction score
0
TL;DR
In summary I am trying to lock the cavity using the modulation demodulation technique, but I need help with the signals I am getting.
Hello, So I am trying to lock an X cavity to a 1064 nm laser. Laser has a linewidth of ~100KHz, cavity linewidth is 37.65KHz, and the FSR is 3.125 MHz. When I try to scan the cavity, the peaks usually shift and after a few seconds, the resonance comes off. I am scanning the cavity at around 100 Hz, and I assume it has some high-frequency noise that contributes to the jitter.
To lock, I am using Lock In+PID method, but with very slow modulation (20-30 KHz), my piezo where I am sending the modulation (attached to one of the cavity mirrors) can go up to 340KHz.

In the scanning image, the lock-in output is in green, the scanned peaks are in yellow. In PID image, blue one is error monitor and green is the output, I am not sure I am using it correctly.
When I use the modulated signal, the lock-in output behaves strangely, like at the peaks, it gives some signal, but whether it is the derivative or not cant be said. Same when it goes to PID.

I am attaching some pictures of what I got. Can someone please help me identify if it is right?

Thanks a lot!
 
Attachments
  • Oscilloscope screen photo: yellow trace of repeated peaks with a green trace beneath, plus a flat purple trace near the bottom, on a dark grid with readout values at the bottom.
    Modulated.jpg
    61.3 KB · Views: 146
  • Oscilloscope screen showing two overlaid traces: a blue trace with flat-topped steps and a green trace with sharp periodic spikes, on a dark grid. Right panel reads Trigger, Type Edge, Source CH2, Coupling DC, Level -500.0mV.
    PID when scanning(feedback is not connected).jpg
    38.2 KB · Views: 123
  • Oscilloscope screen showing a yellow trace of periodic sharp peaks above a noisy baseline, and below it a noisy green trace with smaller irregular spikes; a flat purple line marks zero level.
    Scanning.jpg
    69.9 KB · Views: 124
berkeman said:
Hello, Greetings.
I have another question. If the cavity is locked at a certain point, What should I see as the cavity transmission from the photodiode? A "constant" voltage signal or the peaks? (While the ramp is off)

I think I managed to do it decently up to the error signal; I am attaching a picture of it.

Blue: cavity transmission, Green: Error from lock-in, yellow: error monitor from pid, setpoint as zero V.
 
Attachments
  • Oscilloscope screen showing four traces: a purple triangle-wave ramp, a blue trace with sharp peaks, a green trace with dips and peaks, and a yellow trace with a rounded waveform, with channel scale readings at the bottom.
    20240420_170227.jpg
    76.2 KB · Views: 126