Determination of Ca/Mg in orthophosphate-rich solutions

Click For Summary
SUMMARY

The discussion centers on measuring calcium (Ca) and magnesium (Mg) concentrations in orthophosphate-rich solutions, specifically those containing up to 250 mg/L of PO43-. Complexometric methods, particularly using EDTA, are preferred despite their interference due to high pH conditions that lead to precipitation of calcium and magnesium salts. The participants highlight the lack of a standard method for addressing these interferences and suggest exploring alternative chelating agents that function effectively at lower pH levels. A reference to a 1961 paper proposing an alternative titration method was also mentioned.

PREREQUISITES
  • Understanding of complexometric titration techniques
  • Familiarity with EDTA as a chelating agent
  • Knowledge of pH effects on solubility and precipitation
  • Experience with analytical methods for measuring metal ion concentrations
NEXT STEPS
  • Research alternative chelating agents for calcium and magnesium analysis at lower pH
  • Investigate the 1961 paper on alternative titration methods for orthophosphate interference
  • Learn about the impact of pH on complexometric titrations
  • Explore troubleshooting techniques for atomic absorption spectrophotometers
USEFUL FOR

Chemists, laboratory technicians, and environmental scientists involved in water quality analysis and those seeking to optimize methods for measuring calcium and magnesium in the presence of orthophosphates.

Baho Ilok
Messages
45
Reaction score
5
I want to measure the Ca and Mg concentration in a solution that contains up to 250 mg/L of PO43-. Apparently, orthophosphates interfere with both flame photometric methods and complexometric methods with EDTA (or so I've read). Is there a standard method for measuring Ca and Mg in this type of solution?

I am more preferring complexometric methods at the moment because our atomic absorption spectrophotometer is not functioning well. Thank you!
 
Chemistry news on Phys.org
Complexometry is a gold standard here.

Problem with orthophosphates is that for EDTA titration you need a high pH, and in high pH and in the presence of orthophosphates you can expect precipitation of weakly soluble magnesium and calcium salts. I am not aware of any standard method of dealing with such situations (doesn't mean there is no one, its been ages since the last time I had to deal with these titrations). I would check if there is no other chelating agent that works nicely in lower pH.
 
Borek said:
Complexometry is a gold standard here.

Problem with orthophosphates is that for EDTA titration you need a high pH, and in high pH and in the presence of orthophosphates you can expect precipitation of weakly soluble magnesium and calcium salts. I am not aware of any standard method of dealing with such situations (doesn't mean there is no one, its been ages since the last time I had to deal with these titrations). I would check if there is no other chelating agent that works nicely in lower pH.
I found one paper that proposed an alternative titration method that eliminates orthophosphate interference but it was published in 1961. I was wondering though that there was a standard method that I have not yet read about.
 