Dose calculation for oscilloscope

  • Thread starter Thread starter Salman Khan
  • Start date Start date
  • Tags Tags
    Calculation
Join the discussion
Registration is free. Ask a follow-up in this thread, or start your own.
7 replies · 2K views
Salman Khan
Messages
36
Reaction score
2
If pulse source output is recorded by a detector on oscilloscope, is it possible to calculate dose from this output voltage pulse?
 
Physics news on Phys.org
That depends on the detector and any amplifier. For an ion chamber it would be the total charge transferred, which would be the area under the curve on an oscilloscope trace under some circumstances, or the height of the pulse in others.

A Geiger counter tube would not work. Some other forms of detector might work, but interpreting the result might be challenging.
 
Alex A said:
That depends on the detector and any amplifier. For an ion chamber it would be the total charge transferred, which would be the area under the curve on an oscilloscope trace under some circumstances, or the height of the pulse in others.

A Geiger counter tube would not work. Some other forms of detector might work, but interpreting the result might be challenging.
In case of PMT or other solid state detector?
 
A commercial PMT/Scintillator would probably fail to work. Counters are usually designed with high value resistor chains to save power and with small value capacitors close to the anode. These saturate easily. Spectroscopy grade combinations would probably damage themselves at fairly low exposures.

If you are measuring the dose within the beam I would expect a strong pulse of light from the scintillator. Much more than would be a good fit for a PMT.

A reverse biased diode would probably give a signal, but at these power levels there may be a lot of nonlinear effects. A scintillator connected to a photodiode by a light pipe or fiber optic cable might work well.

I'm bothered that you seem to be trying to reinvent the wheel for instrumentation for a very dangerous device.
 
Reply
  • Like
Likes   Reactions: Salman Khan
Dear Alex, very thanks for sharing your experience.
 
So far as I know ion chamber is gaseous detector having response tim in micro second, how is it possible to measure a pulse dose of nano second pulse using ion chamber ??