Discussion Overview

The discussion revolves around evaluating the Wittig W2022 Scope, focusing on ideas and procedures for thorough testing and assessment of its performance. Participants share their thoughts on necessary equipment, testing methods, and potential challenges in the evaluation process.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested

Main Points Raised

  • Paul expresses a desire for ideas and procedures to evaluate the Wittig W2022 Scope, emphasizing its importance as a high-precision item.
  • Dave suggests obtaining a service manual to understand the specifications and recommends using a calibrated audio signal generator for testing performance.
  • Another participant mentions the need for a high-precision digital voltmeter (DVM) for accurate DC and sine wave RMS measurements, proposing comparisons against a DC power supply.
  • There is a discussion about the scope's measurement capabilities, questioning whether it can perform measurements independently or if it relies solely on visual assessments.
  • A participant considers using the VI curve of standard signal diodes for evaluation, noting that certain tests can confirm overall functionality without checking every parameter individually.
  • Square wave response testing, both unfiltered and filtered, is proposed as a method to assess the scope's performance while accounting for potential shortcomings in the test equipment.
  • Concerns about firmware issues are raised, with references to external resources for firmware projects and the common challenges associated with software in low-end scopes.

Areas of Agreement / Disagreement

Participants present various ideas and methods for evaluating the scope, but there is no consensus on a single approach or solution. Multiple viewpoints and suggestions remain in the discussion.

Contextual Notes

Participants mention the importance of calibration and precision in testing, indicating that some features may be difficult to evaluate accurately. The discussion also highlights the potential complexity of firmware issues without resolving them.

Who May Find This Useful

Individuals interested in evaluating oscilloscopes, particularly those with a focus on precision testing and performance assessment, may find this discussion relevant.

Windadct
Messages
1,452
Reaction score
411
Hello PF'ers...

I just picked up a Wittig W2022 Scope on E-bay.

I would like to give it a thorough evaluation, however, this is to be the highest precision item on the bench - does anyone have some good ideas or procedures for running this through its paces?

Perhaps even an eval form?

Thanks!
Paul
 
Engineering news on Phys.org
Hi Windact

have you got or have tried to get a service manual for it ?
if not you could try here ebaman

he has masses of manuals for all sorts of gear
A manual is at least going to give you the specs the scope should be able to achieve
(taking its age/amount of use into account)

a decent quality audio sig-gen/function generator with calibrated level output
would be an essential signal source for testing any scope's performance

cheers
Dave
 
Windadct said:
Hello PF'ers...

I just picked up a Wittig W2022 Scope on E-bay.

I would like to give it a thorough evaluation, however, this is to be the highest precision item on the bench - does anyone have some good ideas or procedures for running this through its paces?

Perhaps even an eval form?

Thanks!
Paul

You should have a DVM that is higher precision, at least for DC measurements and sine wave RMS measurements... :smile:

Compare the DC accuracy against your DVM using your DC power supply. Then use an AC square wave source to be sure that you can compensate the 'scope probes to flat signal response.

Does the 'scope have any measurement capabilities of its own, or is it all visual measurements made by you?
 
Yes - I'll download the manuals- I and many have some self cal, my DMM is OK - but a few features like bandwith would be tricky.

I was thinking of something like comparing VI curve for a few standard, signal diodes. Their performance should be pretty consistent. Many times when you look at cal you are checking each parameter at once, but IMO many of the digital devices, there are often one or two "tests" that can confirm that pretty much everything is working.

Another woud be Square Wave response - both unfiltered and then filtered. By overlapping results of different test - shortcomings of the test equipment can be ruled out and attributed to the UUT.

Just thinking out loud

Also - thanks for the firmware project - it looks interesting - probably the most common issue/complaint with low end scopes is the software, but in general all software projects are far more difficult then consumers ( and many manufactures) understand. So I want / Need the PC connection - but will see how well the OEMs works and then check the FW Project out.
 

Similar threads

  • · Replies 19 ·
Replies
19
Views
10K
  • · Replies 4 ·
Replies
4
Views
3K
  • · Replies 0 ·
Replies
0
Views
3K
  • · Replies 2 ·
Replies
2
Views
611
  • · Replies 9 ·
Replies
9
Views
10K
  • · Replies 1 ·
Replies
1
Views
5K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 6 ·
Replies
6
Views
4K
  • · Replies 10 ·
Replies
10
Views
5K