Planobilly said:
The manual indicates the above chart is based on the probe being set at 10X.
can you scan and post that ?
I'm from old days
some scopes were 50 Ω in
most were 1 meg and a few pf as your chart says
since the 1X-10X switch was in the probe, the scope stayed 1 meg but there was 9 more meg in the probe
and that's why the probe had a trimmer cap to adjust out the pfs from 'scope and probe's coaxial cable.
By the time i got old, the 'scope acquired another ring on its BNC connector through which the probe apprised the 'scope of whether it was selected for 1X or 10X, and the 'scope's internal microcomputer adjusted the numbers on the screen to match. .
Now scopes have a computer menu so complex that i need to have it interpreted by young people.
So
I think
your scope is 1 meg on X1 and the probe makes it 10 meg on X10
and the max voltage is 400 at the 'scope input
and
the manual that comes with your
probe will tell how for much voltage the probe remains safe... when selected to 10X it divides that voltage by 10 before handing it to 'scope, but i doubt it's good for 4kv.
Here's a "bargain " probe i bought
http://www.mpja.com/20MHz-Oscilloscope-Probe/productinfo/15085+TE/
Quality 200MHz oscilloscope probes with a BNC connector to match to most equipment.
Max Input: 600V AC P/P + DC.
in other words, 600V at any instant, even peak of a sinewave atop DC.