MCNP: Can I input the X-Ray tube voltage in MCNP source specification?

AI Thread Summary
In MCNP, the X-ray tube voltage is not directly input; instead, the energy or energy spectrum of the photon source is specified. For example, a 100kV X-ray tube is simulated using a beam of electrons with an energy of 0.1 MeV directed at a target. The current (mA) affects the number of particles generated per second, while the accelerating potential determines the energy per particle. The simulation outputs results per source particle, allowing for statistical analysis without simulating every particle. Therefore, simulations for different mA settings at the same voltage can yield the same results, with post-simulation calculations adjusting for intensity differences.
Anisur Rahman
Messages
11
Reaction score
1
TL;DR Summary
Source Specification
Can I input the X-Ray tube voltage in MCNP source specification?
 
Engineering news on Phys.org
Anisur Rahman said:
TL;DR Summary: Source Specification

Can I input the X-Ray tube voltage in MCNP source specification?
One would put in energy or energy spectrum for a photon source. I am not aware that voltage for an X-ray tube is an input.
 
An electron falling through a potential difference of 1V gains 1eV of kinetic energy.

So an X-ray tube powered at 100kV would be simulated with a beam of electrons of energy 0.1 MeV (SDEF ERG= are in MeV) directed at a target.
 
  • Informative
  • Like
Likes Salman Khan, PSRB191921 and berkeman
Alex A said:
An electron falling through a potential difference of 1V gains 1eV of kinetic energy.

So an X-ray tube powered at 100kV would be simulated with a beam of electrons of energy 0.1 MeV (SDEF ERG= are in MeV) directed at a target.
Even if the current value (mA) differs?
 
mariem said:
Even if the current value (mA) differs?
The current determines the number of particles (mA/e) per sec, were e is the unit charge.

The accelerating potential determines the energy per particle.

The total energy/sec or power would be the number of particles/sec * energy/particle, and the total energy or dose deposited would be related to the integral over time of the (particles/sec * energy/particle).
 
Yes, as @Astronuc says the intensity is the other part of the problem. Almost all answers output by MCNP are per source particle. So no one would attempt to simulate one mA-second (a milli coulomb) of electrons. Enough particles are simulated to get a statistically useful answer and then the results which are per source particle are multiplied by the real world source activity to get flux or dose rate answers, or the total number of real world particles to get dose.

It also means that if you are simulating a 1mA tube and a 10mA tube (the same tube at the same voltage) that only one simulation needs to be done because these situations are the same. Only the maths done afterwards differs.
 
Where can I input the electron source intensity in MCNP? How many particles will I have for an X-ray tube set at an intensity of 10 mA?
 
Do you understand coulombs, do you understand unit charge (the charge on an electron)?
 