Spectrometer sensitivity (photons/count) at different wavelengths

Click For Summary
SUMMARY

The discussion focuses on the sensitivity of spectrometers, specifically how they measure irradiance across different wavelengths. It is established that spectrometers exhibit varying sensitivity (photons/count) at different wavelengths, which can lead to inaccurate relative irradiance readings. Calibration is essential for accurate measurements, particularly using reference spectra from known sources, such as gold for FTIR spectrometers. The discussion emphasizes the importance of understanding the sensor characteristics and the optical setup to ensure reliable results.

PREREQUISITES
  • Understanding of spectrometer operation and calibration techniques
  • Familiarity with FTIR (Fourier Transform Infrared Spectroscopy) principles
  • Knowledge of responsivity curves in photodetectors
  • Basic principles of optical setups and their impact on measurements
NEXT STEPS
  • Research calibration techniques for FTIR spectrometers
  • Learn about responsivity curves for different photodetectors
  • Explore the impact of optical parameters on spectrometer measurements
  • Investigate methods for ensuring wavelength accuracy in spectrometry
USEFUL FOR

Researchers, optical engineers, and laboratory technicians involved in spectrometry, calibration processes, and those seeking to improve the accuracy of spectral measurements.

fog37
Messages
1,566
Reaction score
108
TL;DR
spectrometer sensitivity at different wavelengths does not provide correct relative irradiance....
Hello Everyone,

I am trying to better understand how a spectrometer must be used to measure the wavelength content of the radiation from a specific source.
All spectrometers measure irradiance over a wavelength range (for ex, UV-VIS) but the sensitivity (photons/count) is not the same for all wavelengths. This means that even if the radiation contains the same energy at two different wavelengths ##\lambda_1## and ##\lambda_2##, the spectrometer will show the wavelength with the lower sensitivity to have a higher irradiance even if that is not what is going on...

Does that mean that spectrometers cannot provide a reliable relative irradiance when we compare different wavelengths? How do we solve for that?

Thanks!
 
Engineering news on Phys.org
How you resolve this issue (called responsivity curve in the language of IR detectors and photodiodes, probably called other things in other contexts) depends on what kind of spectrometer you are dealing with. For FTIR, you can calibrate the spectrum by taking a reference spectrum over a very non-dispersive material (gold in the IR, for example). (This procedure also deals with the spectrum of the broadband light source.) There isn't a one-size-fits-all procedure.
 
  • Like
Likes   Reactions: fog37
fog37 said:
This means that even if the radiation contains the same energy at two different wavelengths λ1 and λ2, ...
It also assumes that the aperture of the effective slit is the same width at both wavelengths.

You need to find a source that radiates a known broad band spectrum. Then you can calibrate the sensitivity of your system.

Knowing the predictable characteristics of the sensor employed can resolve the problem. Some sensors, such as bolometers, register thermal energy independent of wavelength.
 
  • Like
Likes   Reactions: fog37 and Twigg
There are two levels of concern here, depending upon the system and your needs. The most direct calibration is to set up the system and compare test sample directly to a known sample . Usually I run known source spectrum --test source scan---known source spectrum. The happy news is that silicon photodetectors are very linear in response, so that is all you really need.
Sometimes you don't have the luxury to calibrate in situ at the time of the test. Then you need to calibrate beforehand and also know a priori how changes in optical setup (calibration vs test) may effect your result. Be aware that optics (slit width optic focal number etc) and temperature can be important. It is easy to get lost in the minutiae.
Spectrometers are exceedingly clever and precise when well used.
 
  • Like
Likes   Reactions: Charles Link and Twigg
I see. So calibration is the solution.

I thought that calibration would only convert the vertical axis, which measures counts, to a correct intensity value ##W/m^2##. But I guess calibration can also provide a correction factor to take care of the different sensitivity across different wavelengths which would result in the relative intensity issues I am describing...
 
  • Like
Likes   Reactions: hutchphd
And of course there is an entirely different method for calibrating the wavelength accuracy which involves sources that emit at known frequencies. These again depend upon your exact requirements but are not particularly arcane nor complicated. Wonderful instruments.
 
  • Like
Likes   Reactions: Twigg

Similar threads

Replies
11
Views
1K
  • · Replies 23 ·
Replies
23
Views
3K
  • · Replies 11 ·
Replies
11
Views
7K
Replies
2
Views
3K
Replies
7
Views
3K
  • · Replies 1 ·
Replies
1
Views
1K
  • · Replies 31 ·
2
Replies
31
Views
4K
Replies
8
Views
5K
  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 18 ·
Replies
18
Views
3K