Spectrometer sensitivity (photons/count) at different wavelengths

AI Thread Summary
Spectrometers measure irradiance across various wavelengths, but their sensitivity (photons/count) varies, affecting the reliability of relative irradiance comparisons. Calibration is essential to address this issue, often involving reference spectra from known sources to correct sensitivity discrepancies. Different spectrometer types, like FTIR, require specific calibration methods tailored to their characteristics. Additionally, understanding the sensor's predictable behavior, such as that of silicon photodetectors, can enhance measurement accuracy. Overall, proper calibration ensures that spectrometers can provide precise and reliable data across different wavelengths.
fog37
Messages
1,566
Reaction score
108
TL;DR Summary
spectrometer sensitivity at different wavelengths does not provide correct relative irradiance....
Hello Everyone,

I am trying to better understand how a spectrometer must be used to measure the wavelength content of the radiation from a specific source.
All spectrometers measure irradiance over a wavelength range (for ex, UV-VIS) but the sensitivity (photons/count) is not the same for all wavelengths. This means that even if the radiation contains the same energy at two different wavelengths ##\lambda_1## and ##\lambda_2##, the spectrometer will show the wavelength with the lower sensitivity to have a higher irradiance even if that is not what is going on...

Does that mean that spectrometers cannot provide a reliable relative irradiance when we compare different wavelengths? How do we solve for that?

Thanks!
 
Engineering news on Phys.org
How you resolve this issue (called responsivity curve in the language of IR detectors and photodiodes, probably called other things in other contexts) depends on what kind of spectrometer you are dealing with. For FTIR, you can calibrate the spectrum by taking a reference spectrum over a very non-dispersive material (gold in the IR, for example). (This procedure also deals with the spectrum of the broadband light source.) There isn't a one-size-fits-all procedure.
 
fog37 said:
This means that even if the radiation contains the same energy at two different wavelengths λ1 and λ2, ...
It also assumes that the aperture of the effective slit is the same width at both wavelengths.

You need to find a source that radiates a known broad band spectrum. Then you can calibrate the sensitivity of your system.

Knowing the predictable characteristics of the sensor employed can resolve the problem. Some sensors, such as bolometers, register thermal energy independent of wavelength.
 
  • Like
Likes fog37 and Twigg
There are two levels of concern here, depending upon the system and your needs. The most direct calibration is to set up the system and compare test sample directly to a known sample . Usually I run known source spectrum --test source scan---known source spectrum. The happy news is that silicon photodetectors are very linear in response, so that is all you really need.
Sometimes you don't have the luxury to calibrate in situ at the time of the test. Then you need to calibrate beforehand and also know a priori how changes in optical setup (calibration vs test) may effect your result. Be aware that optics (slit width optic focal number etc) and temperature can be important. It is easy to get lost in the minutiae.
Spectrometers are exceedingly clever and precise when well used.
 
  • Like
Likes Charles Link and Twigg
I see. So calibration is the solution.

I thought that calibration would only convert the vertical axis, which measures counts, to a correct intensity value ##W/m^2##. But I guess calibration can also provide a correction factor to take care of the different sensitivity across different wavelengths which would result in the relative intensity issues I am describing...
 
And of course there is an entirely different method for calibrating the wavelength accuracy which involves sources that emit at known frequencies. These again depend upon your exact requirements but are not particularly arcane nor complicated. Wonderful instruments.
 
Hi all, I have a question. So from the derivation of the Isentropic process relationship PV^gamma = constant, there is a step dW = PdV, which can only be said for quasi-equilibrium (or reversible) processes. As such I believe PV^gamma = constant (and the family of equations) should not be applicable to just adiabatic processes? Ie, it should be applicable only for adiabatic + reversible = isentropic processes? However, I've seen couple of online notes/books, and...
I have an engine that uses a dry sump oiling system. The oil collection pan has three AN fittings to use for scavenging. Two of the fittings are approximately on the same level, the third is about 1/2 to 3/4 inch higher than the other two. The system ran for years with no problem using a three stage pump (one pressure and two scavenge stages). The two scavenge stages were connected at times to any two of the three AN fittings on the tank. Recently I tried an upgrade to a four stage pump...