Sterilizing Medical Devices that contain Electronic Components

  • Thread starter Thread starter inforseeker
  • Start date Start date
Click For Summary

Discussion Overview

The discussion revolves around the challenges of sterilizing medical devices that contain sensitive electronic components. Participants explore various sterilization methods, including gamma radiation and ethylene oxide gas, while addressing the potential impacts on electronic components and the design considerations necessary for effective sterilization.

Discussion Character

  • Technical explanation
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • One participant reports failures in electronic components after gamma sterilization at 45 kGy and is considering alternative methods like electron beam sterilization.
  • Another participant questions whether the electronic devices were powered during gamma sterilization, suggesting that this could lead to critical ionization events.
  • There is mention of the need to consider what parts of the device require sterilization, particularly in relation to the electronic components.
  • Ethylene oxide gas is proposed as a sterilization method, with a caution about its explosive nature and the recommendation to involve professionals for its use.
  • A participant shares insights on covering devices with sterile covers as an alternative to direct sterilization, particularly for devices that do not contact patients directly.
  • Another participant discusses the design considerations for ethylene oxide sterilization, including the need for venting internal cavities and avoiding mechanical joints that could trap pathogens.
  • There is a reference to specific materials and adhesives that are compatible with ethylene oxide sterilization, highlighting the importance of material selection in device design.

Areas of Agreement / Disagreement

Participants express differing views on the effectiveness and safety of various sterilization methods, particularly regarding gamma radiation and ethylene oxide. There is no consensus on the best approach, and multiple competing views remain regarding the sterilization of electronic components in medical devices.

Contextual Notes

Participants mention the need for careful consideration of device design, including the impact of sterilization methods on electronic components and the importance of material compatibility. Limitations in knowledge about the effects of radiation on electronics and the specifics of sterilization processes are acknowledged.

inforseeker
Messages
1
Reaction score
0
I'm working on a medical device that contains sensitive electronic components (standard commercial-grade PCBs and surface-mount components). The device needs to be sterilized before clinical use, and I'm running into challenges finding a method that ensures sterility without damaging the electronics.
We've tried gamma sterilization at 45 kGy, but we're seeing failures in some components afterward. I’m now evaluating alternative methods like E beam.
I’d really appreciate insight from anyone who has experience with:
  • Successfully sterilizing electronics in medical devices
  • Specific components or PCB coatings that are more resistant to gamma radiation
  • Known failure mechanisms from radiation or plasma exposure
 
Engineering news on Phys.org
Did one use an accelerator for the gamma sterilization, or 60Co?

Were the electronic devices 'ON' or powered during the gamma sterilization process? That might induce a critical ionization event, even a single event upset (SEU). Gammas are much more penetrating than electrons. Depending on the energy, gammas will ionize atoms by photoelectric effect, Compton scattering, or pair production (e+e-). The photo- and Compton electrons will also cause electron cascades in the matter through which they pass.

Electrons are attenuated quickly, so depending on the device dimensions, one might need to go in two directions from front and back, or top and bottom, i.e.,. from opposite directions.

What needs to be sterilized? Outside of the electronic components?

I have done simulations and dose calculations on medical devices, and various related materials, for photons (X-ray, gamma) and electrons. I have not done any simulations on electronics, but that could be done if a model is available.

I may be able to put you in touch with some people who do similar work, and I believe they are interested in radiation effects in electronics.

If interested, send me a DM. Do not put personal information in this (or any other) thread/post.

There is a separate effort at my organization to look at radiation effects on electronics.
 
Last edited:
  • Informative
  • Like
Likes   Reactions: Lord Jestocost and berkeman
Ethylene Oxide gas is used for sterilization in hospitals for things that can not withstand steam sterilization.

Edit: Problem is that IT IS EXPLOSIVE, so you better farm out that part of the project to the professionals!

Cheers,
Tom
 
Last edited:
  • Like
Likes   Reactions: davenn
If the device does not itself has direct contact with the patient then it may be an option to cover it with a sterile cover instead or as additional protection.

I work with software for OR ultrasound devices and I understand that for those devices the solution is to cover the main computer unit in sterile covers (allowing the surgeon to still see displays and control button) such that only the probes themselves needs to be able to withstand the special sterilization fluids that they are designed for.
 
ethylene oxide gas sterilization is relatively benign. temperatures can be as high as 60 degrees centigrade. The device will also need to withstand the vacuum cycles used to remove the gas. Internal cavities will need to be vented to the exterior so the gas can get in and quickly be removed under vacuum, hermetically sealed, or filled with inert impermeable material. Avoid mechanical joints and parts with crevices (e.g sockets, porous materials) where possible. Fill unavoidable cavities (e.g. under components) by vacuum impregnation with an EtO compatible epoxy. Design/use unavoidable connectors for easy gas flow in and out. Elastomers like silicone rubber are generally permeable to EtO and easy to sterilize, but thick sections require longer cycle times to allow full permeation and purging.

Case in point. An eye surgeon who was treating a teen who had been poked in the eye by a rotten stick and acquired bacterial, fungal, and viral infections brought me literature for a continuous medicine infusion device that had a squashed football shaped silicone rubber retainer connected to a small silicone tube threaded by a needle through the eyelid. It had been removed from the market because pathogens occasionally avoided sterilization in the stainless tubing used to connect the tubing to the retainer. He asked if i could construct one ASAP, because having nurses manually every half hour flush the patient's eye wasn't gaining on his complex infection. I worked late constructing a polished aluminum mold to cast the retainer integral with the end of the tubing, made a part, cleaned it, and gave it to the sterile processing night shift. About a week later, the doctor came by and said, "thank you for your work making that device: having it saved that kids eye."

From the intertubes - "Master Bond multiple ETO resistant adhesives offer a range of properties. High/low viscosity systems feature toughness, optical clarity, fast cures, thermal conductivity, electrical conductivity, low water absorption and vibration/impact resistance."
Also see https://sterigenics.com/technologies/ethylene-oxide/

Good luck; sometimes the intangible rewards are important.
 
Last edited by a moderator:
  • Like
Likes   Reactions: Filip Larsen and berkeman

Similar threads

  • · Replies 11 ·
Replies
11
Views
6K
  • Poll Poll
  • · Replies 1 ·
Replies
1
Views
5K
  • · Replies 152 ·
6
Replies
152
Views
11K
Replies
4
Views
10K
  • · Replies 1 ·
Replies
1
Views
10K