A Ultrasound Velocities in 25% Propylene Glycol + Water between 20-40C

AI Thread Summary
The discussion centers on the challenges of using an ultrasonic transit-time flow meter calibrated for water with a 25% propylene glycol and 75% water mixture. The user seeks experimental data on sound velocity in this mixture across temperatures from 20°C to 40°C, as the device requires specific inputs for sound velocity and viscosity. Suggestions include creating calibration procedures and using an in-line electromagnetic flowmeter for reference to back-calculate sound velocity. The conversation highlights the importance of accurately determining the glycol concentration and its impact on flow measurements. Overall, reliable sound velocity data is crucial for ensuring the accuracy of flow readings in this specific application.
MMRite
Messages
2
Reaction score
0
TL;DR Summary
I'm using a transit-time ultrasonic flow meter with water presets, but my fluid is 25% propylene glycol and 75% water. I need data on how the speed of sound in this mixture varies from ~20°C to 40°C to improve measurement accuracy. Any help or references would be appreciated!
Hi everyone, I'm working with an ultrasonic transit-time flow meter that has preset settings calibrated for water. However, my application involves a 25% propylene glycol and 75% water mixture, and I suspect the speed of sound in this mixture deviates enough from water to affect accuracy. I'm looking for experimental data or reliable estimates of how the speed of sound in this mixture changes with temperature (ideally from ~20°C to 40°C). If anyone has access to such data or knows of relevant studies, I’d really appreciate your help!
 
Physics news on Phys.org
The first thing I would check is whether you device has a calibration procedure.
What is the model number? Have you checked the user's manual?

Failing that, I would try to create a pre-measurements calibration procedure - one that you would use to set over the water presets.
Failing that, I would try to create a post-measurement calibration procedure.
 
  • Like
Likes russ_watters
I’m using an Omega FTD-25 flowmeter. I’ve reviewed the user manual, and it doesn’t include a specific calibration procedure. However, it does mention that when working with a fluid not already preset in the device, it requires two user inputs: sound velocity and viscosity.

Since I couldn’t find reliable sound velocity data for my fluid (25% propylene glycol + water) at varying temperatures, I’m planning to use an in-line electromagnetic flowmeter as a reference. My idea is to compare the EM flowmeter’s readings with the ultrasonic meter’s transit time data to back-calculate the actual sound velocity in the fluid.

From what I understand, the flowmeter uses the sound speed and viscosity values primarily as a sanity check to ensure the readings it’s getting are physically plausible.
 
That User's manual is pretty light on calibration information.
On page 17:
§3.12 How to get a scale factor for calibration: A scale factor is the ratio between the ‘actual flow rate’ and the indicated value by the flow meter. The scale factor can be determined by calibration with flow calibration equipment.
So, once you know how far off it is, there is a "scale" value that should allow you to get a roughly correct value reported.

Another issue is going to be how precise that "25%" is. You may need to calibrate for a total of three factors: actual glycol concentration (or other known variable that changes that concentration), temperature, and indicated flow rate.

That "scale" may only be useful in getting approximate flow rates out under the most typical conditions.

Which lead to a couple of other questions - how will that calibrated flow rate value be used? - and how accurate/precise does it need to be to fulfill that purpose?
 
The rope is tied into the person (the load of 200 pounds) and the rope goes up from the person to a fixed pulley and back down to his hands. He hauls the rope to suspend himself in the air. What is the mechanical advantage of the system? The person will indeed only have to lift half of his body weight (roughly 100 pounds) because he now lessened the load by that same amount. This APPEARS to be a 2:1 because he can hold himself with half the force, but my question is: is that mechanical...
Some physics textbook writer told me that Newton's first law applies only on bodies that feel no interactions at all. He said that if a body is on rest or moves in constant velocity, there is no external force acting on it. But I have heard another form of the law that says the net force acting on a body must be zero. This means there is interactions involved after all. So which one is correct?
Thread 'Beam on an inclined plane'
Hello! I have a question regarding a beam on an inclined plane. I was considering a beam resting on two supports attached to an inclined plane. I was almost sure that the lower support must be more loaded. My imagination about this problem is shown in the picture below. Here is how I wrote the condition of equilibrium forces: $$ \begin{cases} F_{g\parallel}=F_{t1}+F_{t2}, \\ F_{g\perp}=F_{r1}+F_{r2} \end{cases}. $$ On the other hand...

Similar threads

Back
Top