Understanding Absorbance Values in UV Spectroscopy

Join the discussion
Ask a follow-up here, or get your own question answered by working scientists, mathematicians and engineers — people, not an autocomplete.
Real named experts · corrections over time · the nuance an AI answer skips
2 replies · 2K views
kokololo
Messages
1
Reaction score
0
Hi guys, please can you kindly help me. I am getting an absorbance of over 6.0Abs for my experiment. However, when I am reading posts they mention that an absorbance above 1 is not reliable. Please can you explain so that I can understand. Thanks
 
Chemistry news on Phys.org
What it is that you don't understand here? It is hard to help on a dry facts not knowing why you have problems with them.

Generally speaking to measure that high absorbance you would need to be able to accurately measure intensity of radiation over 6 orders of magnitude, that's not trivial.
 
An abosorbance value of 6.0 means that if you shine a light through the solution, only 0.0001% of the light passes through. Because it's difficult to accurately measure such a tiny amount of light, the absorbance values are not very reliable. To measure the absorbance of the solution, you should first dilute your sample so that the absorbance falls within a reliable range (generally ~ 0.1-1.0).