A vacuum furnace setup is being designed to fill with argon after reaching a vacuum of about 10^-5 mbar. The proposed configuration includes a KF tee, a valve, a pressure regulator set to 0.5 bar, and an argon bottle. Concerns were raised about whether a single valve would suffice to maintain vacuum and if the pressure regulator could allow gas to flow back into the bottle due to the vacuum. It was clarified that as long as the argon bottle pressure remains higher than the furnace pressure, backflow won't occur, but for continuous argon supply, a mass flow controller may be necessary. The discussion emphasizes the importance of maintaining a slight underpressure during processes involving high-entropy alloys to minimize manganese evaporation.
dudeekeen
Messages
2
Reaction score
0
TL;DR Summary
I want to fill my vacuum furnace with argon after pumping down. How to set up valves and regulator to avoid vacuum leaks and gas backflow?
Hi everyone,


I have a vacuum furnace that can reach about 10^-5 mbar. On the KF flange, I currently have a capped tee. I want to build a system that, after pumping down the vacuum, allows me to fill the furnace with argon from a gas bottle through a pressure regulator (e.g., set to 0.5 bar), so the furnace atmosphere is argon instead of air.


Here’s the rough idea for the setup:
vacuum furnace → KF tee → valve (e.g., ball valve) → pressure regulator → argon bottle


I have some questions:


  • Will one valve be enough to maintain the vacuum once closed?
  • Could the pressure regulator at 0.5 bar cause gas to be sucked back into the bottle due to the vacuum in the furnace?
  • How can I prevent the vacuum from pulling gas through the regulator backwards?
  • Should I use a check valve or some other type of valve for better protection? If so, which kind?

I’d really appreciate any advice, tips, or experiences with similar setups.


Thanks a lot!
 
Engineering news on Phys.org
dudeekeen said:
Could the pressure regulator at 0.5 bar cause gas to be sucked back into the bottle due to the vacuum in the furnace?
If the pressure in the argon bottle remains greater than in the furnace, then argon cannot flow, backwards through the regulator, into the bottle.
 
You didn't say what you're doing with the furnace. If you just want to 'backfill' the furnace with Argon, your setup will work. Gas won't flow from low pressure (furnace) to high pressure (bottle).
The furnaces that I've been involved with weren't as simple as just 'backfill' - they required a continuous supply of argon - 'flushing' the furnace to maintain composition (melting Titanium, Plasma etching, etc.). In this case (flushing) a mass flow controller and a throttle valve are typically used to maintain target pressure - "fixed flow / variable throttle" or "fixed throttle / variable flow" both work.
 
Dullard said:
You didn't say what you're doing with the furnace. If you just want to 'backfill' the furnace with Argon, your setup will work. Gas won't flow from low pressure (furnace) to high pressure (bottle).
The furnaces that I've been involved with weren't as simple as just 'backfill' - they required a continuous supply of argon - 'flushing' the furnace to maintain composition (melting Titanium, Plasma etching, etc.). In this case (flushing) a mass flow controller and a throttle valve are typically used to maintain target pressure - "fixed flow / variable throttle" or "fixed throttle / variable flow" both work.
I produce large single crystals of metals to study the fundamental mechanisms of plastic deformation. The process is carried out in a vacuum furnace, and it works beautifully for most materials — except for high-entropy alloys. These alloys contain a significant amount of manganese, which tends to evaporate excessively under reduced pressure.

My idea was to first create a high vacuum, then backfill the chamber with argon, repeat the vacuum step to further purge residual gases, and finally fill with argon again. However, it’s important for the process to still maintain a minimal underpressure (slightly below atmospheric) during the crystal growth, to limit manganese evaporation while preserving a clean environment.
 
Hi all, I have a question. So from the derivation of the Isentropic process relationship PV^gamma = constant, there is a step dW = PdV, which can only be said for quasi-equilibrium (or reversible) processes. As such I believe PV^gamma = constant (and the family of equations) should not be applicable to just adiabatic processes? Ie, it should be applicable only for adiabatic + reversible = isentropic processes? However, I've seen couple of online notes/books, and...
Thread 'How can I find the cleanout for my building drain?'
I am a long distance truck driver, but I recently completed a plumbing program with Stratford Career Institute. In the chapter of my textbook Repairing DWV Systems, the author says that if there is a clog in the building drain, one can clear out the clog by using a snake augur or maybe some other type of tool into the cleanout for the building drain. The author said that the cleanout for the building drain is usually near the stack. I live in a duplex townhouse. Just out of curiosity, I...
I have an engine that uses a dry sump oiling system. The oil collection pan has three AN fittings to use for scavenging. Two of the fittings are approximately on the same level, the third is about 1/2 to 3/4 inch higher than the other two. The system ran for years with no problem using a three stage pump (one pressure and two scavenge stages). The two scavenge stages were connected at times to any two of the three AN fittings on the tank. Recently I tried an upgrade to a four stage pump...
Back
Top