Accuracy of an Optical Spectrum Analyzer (OSA)

Click For Summary

Discussion Overview

The discussion revolves around the accuracy of an Optical Spectrum Analyzer (OSA), specifically addressing the interpretation of accuracy specifications in dB and their implications for measurements at different power levels. Participants explore the relationship between accuracy, input power levels, and the potential confusion arising from the use of dB as a unit of measurement.

Discussion Character

  • Technical explanation
  • Debate/contested

Main Points Raised

  • One participant questions the meaning of a 0.4 dB accuracy specification, suggesting that it may lead to high relative errors at low power levels, particularly around 0 dBm.
  • Another participant suggests providing a reference to the datasheet for clarity and notes the importance of understanding the minimum detectable power.
  • A participant identifies the specific model of the OSA and mentions that the minimum detectable power is -65 dBm, indicating that accuracy specifications may vary with wavelength.
  • One participant recalls that the accuracy specification is given as +/- 0.4 dB at an input level of -23 dB and discusses the linearity of accuracy with changing input levels.
  • Another participant argues that expressing accuracy in dB is a valid alternative to percentage, noting that uncertainty may remain consistent across different scales.
  • A later reply acknowledges the previous contributions and suggests that a linear relationship between accuracy and input level may resolve the initial confusion.

Areas of Agreement / Disagreement

Participants express differing views on the interpretation of accuracy in dB and its implications for measurements at various power levels. While some participants propose that a linear relationship may clarify the issue, others highlight the potential for confusion in the specifications.

Contextual Notes

Limitations include the dependence on specific input levels and the potential variability of accuracy specifications across different models and conditions.

zambadeos
Messages
3
Reaction score
0
Hey!
I've a simple (and maybe stupid) question about the accuracy of an OSA. The datasheet report a level accuracy of 0.4 dB but this seems quite strange to me. Indeed level measurements are done in dBm thus, as long as dB measures a relative power, I assume that even the 0.4 dB are referred to 1 mW as in the case of dBm. Am I correct?
If yes what sounds strange to me is that if I measure a power near to 1 mW than the relative error will be quite high (I'm measuring something like 0 dBm) and as the power that I'm measuring decreases also the relative error decreases.
An example to explain my problem:
1) P=0.4 dBm, the relative error is equal to 100%. Expressed in mW means that I'm measuring a power of 1.096 mW
2) P=-65 dBm (the minimum detectable level for my OSA), the relative error is equal to 0.6%. Expressed in watt: [itex]P= (0.316 \pm 0.002)~nW[/itex], that would mean an accuracy of 2 pW: quite amazing!

So where is the bug in my reasoning?

Thanks to everyone
 
Science news on Phys.org
Hello
It might be a good idea to give a reference to that particular data sheet. I agree that what you seem to have read is a bit confusing. There should be some indication of the lowest power value it can register, at least.
 
Ok, my OSA is MS9710C from Anritsu, you can find the datasheet here (appendix A).
In any case the minimum detectable power should be -65 dBm as i wrote above (for the sake of completeness i have to say that it changes accordingly to the wavelength range in which you are working, but we can focus on that value). Maybe my example is not correct because I'm working at the lower limit of my detector but the conceptual problem is still present even if we consider an higher value.
Moreover I've read datasheets from some other osa and i saw that is quite common to express accuracy in terms of dB, so there must be an error in my reasoning!
In any case thank you for the answer!
 
It's been a while since I've used this type of instrument, but the level specification (page A-3) is given as +/- 0.4 dB with an input level of -23 dB. Since the linearity spec is +/- 0.05 dB with input levels between -50 and 0 dB, I expect the accuracy spec to be approximately linear with changing input levels- curse you, log scales!
 
  • Like
Likes   Reactions: zambadeos
zambadeos said:
that is quite common to express accuracy in terms of dB,
...and why not? It's only an alternative to percentage for showing the ratio. On a dB scale, it'd really just what you want, I would have thought. You could argue that, in many measurements, the uncertainty is often the same, whatever scale you are using - so it would be more appropriate to use an absolute measure of accuracy. In the end, the choice would depend upon what is being measured and the method being used.
 
Thank you very much for the answers.
Andy Resnick you get the solution to my problem, i didn't thought about a linear relation between the accuracy and the input level but it seems to be the right answer :wink::wink:
 

Similar threads

  • · Replies 2 ·
Replies
2
Views
4K
  • · Replies 10 ·
Replies
10
Views
3K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 6 ·
Replies
6
Views
3K
  • · Replies 0 ·
Replies
0
Views
3K
  • · Replies 46 ·
2
Replies
46
Views
6K
  • · Replies 19 ·
Replies
19
Views
6K
  • · Replies 2 ·
Replies
2
Views
631