Anyone have any experience with Oxygen meters?

  • Thread starter Thread starter rwooduk
  • Start date Start date
  • Tags Tags
    Experience Oxygen
Click For Summary
SUMMARY

This discussion centers on the use of the HANNA HI98186 Dissolved Oxygen Meter for measuring oxygen levels in degassed potassium iodide and phenol solutions. Users report that the oxygen meter takes a long time to stabilize and question the accuracy of the minimum value obtained. The conversation highlights the need for proper circulation around the probe to ensure accurate readings, as stagnant films can affect measurements. Additionally, the participants explore alternative methods for inferring oxygen levels and the challenges of calibrating measurements between different solutions.

PREREQUISITES
  • Understanding of dissolved oxygen measurement techniques
  • Familiarity with potassium iodide and phenol solutions
  • Knowledge of the HANNA HI98186 Dissolved Oxygen Meter
  • Basic principles of solution degassing and calibration curves
NEXT STEPS
  • Research the effects of solution viscosity on dissolved oxygen measurements
  • Learn about calibration techniques for dissolved oxygen meters
  • Investigate methods for ensuring proper circulation around measurement probes
  • Explore alternative dissolved oxygen meters suitable for specific applications
USEFUL FOR

Researchers, laboratory technicians, and chemists involved in experiments requiring precise oxygen level measurements in liquid solutions, particularly those working with potassium iodide and phenol.

rwooduk
Messages
757
Reaction score
59
If anyone has used or uses an oxygen meter and could offer some advice it would be greatly appreciated.

I degassed a solution of potassium iodide for 10 minutes each time and did experiments. However to get a more accurate measurement of the amount of O2 present I've decided to use an oxgen meter.

At the same time I am degassing phenol solution and need to do experiments at the same oxygen content as the potassium iodide used in previous experiments.

Q1. the oxgen meter takes a LONG time to stabilise when taking a measurement of degassed solution. It decreases rapidly then decreases slowly until it reaches a minimum value. Is this normal? And am I correct in taking the minimum value it gives i.e. the oxygen probe won't cause further degassing so the minimum value it gives must be the closest value to the truest value.

Q2. for the potassium iodide solution, I have dissolved 16g KI in one litre of distilled water. Am I correct in setting the Salinity value of the meter to 16g/L?

Q3. if i degas 400ml potassium iodide for 10 minutes i get an oxygen value of around 13%. If I degas 750ml of phenol solution (not a salt) for 10 minutes I get an oxgen value of around 10%. Shouldnt the potassium iodide solution hold LESS oxygen because its a salt?

Q4. the idea was to test the oxygen content of the potassium iodide solution after 10 minutes degassing (because I have done this in previous experiments) and THEN make a calibration curve of degas time vs oxygen content for the phenol solution. Which I can then use to decide how long to degas the phenol for to achieve the same oxygen content as I had with the KI solution. HOWEVER the curve isn't linear it decreases rapidly with degas time and then flattens out when the solution becomes full degassed. So how would I go about doing this?

If anyone can make any suggestions regarding one or more of the above questions it would really help our project move forward!
 
Engineering news on Phys.org
Can you :

(a) measure some secondary quantity and infer the oxygen levels - maybe pH , density or conductivity ?

(b) collect and analyse the gasses removed ?

(c) reverse the process and start with a fully degassed solution to which you then add metred amounts of oxygen ?
 
What oxygen meter are you currently using ? There are many different ones and there may be something more suitable available .
 
Hi many thanks for the reply,

Nidum said:
Can you :

(a) measure some secondary quantity and infer the oxygen levels - maybe pH , density or conductivity ?

(b) collect and analyse the gasses removed ?

(c) reverse the process and start with a fully degassed solution to which you then add metred amounts of oxygen ?
Nidum said:
What oxygen meter are you currently using ? There are many different ones and there may be something more suitable available .

(a) I could potentially measure the pH, although I am unsure how this would work as phenol turns the water acidic. i.e. I am unfamiliar with this technique, could you perhaps please explain further?

(b) not possible it is degassed via a degas pump

(c) unfortunately we don't have access to oxygen at present as the chemistry department aren't willing to share oxygen due to safety protocols :-/ however to degas and regas samples I think would introduce new problems and we would still need to measure the oxygen content once regassed to be sure of the percentage. But the ideas is very much appreciated, thanks.

Nidum said:
What oxygen meter are you currently using ? There are many different ones and there may be something more suitable available .

I only have access to one and we do not have funds to purchase another type. But to answer your question it is a HANNA HI98186 Dissolved Oxygen Meter...

http://www.hannainstruments.co.uk/advanced-high-range-dissolved-oxygen-and-bod-meter.html

I'm not sure if the above info will be of use to the problem? If you have any insights into problems associated with our O2 meter they would certainly be welcome.

thanks again!
 
One of the problems with immersed probe gauges of all types is that the layer of liquid that the probe is sampling can have different properties to the bulk of liquid .

Main causes are stagnant films and bubble layers . More of a problem with higher viscosity liquids than with lower viscosity ones .

May be worth devising a test procedure which ensures that the liquid around the probe is freely circulating .
 
  • Like
Likes   Reactions: rwooduk
Nidum said:
One of the problems with immersed probe gauges of all types is that the layer of liquid that the probe is sampling can have different properties to the bulk of liquid .

Main causes are stagnant films and bubble layers . More of a problem with higher viscosity liquids than with lower viscosity ones .

May be worth devising a test procedure which ensures that the liquid around the probe is freely circulating .
Thanks and yes I believe circulation is the issue! Thanks! I re-read the manual and there is a small sentence hidden away (typical!) which indicates that for accurate measurements a small circulation flow must be applied, basically stirring using the probe. This is because the probe actually consumes oxygen when left in the solution!

Many thanks again for your replies!
 

Similar threads

Replies
3
Views
2K
  • · Replies 3 ·
Replies
3
Views
11K
Replies
15
Views
3K
  • · Replies 12 ·
Replies
12
Views
3K
Replies
16
Views
3K
Replies
5
Views
4K
  • · Replies 4 ·
Replies
4
Views
4K
Replies
3
Views
850
Replies
6
Views
2K
  • · Replies 5 ·
Replies
5
Views
5K