Baseline Recovery time of Shaper Amplifier output

AI Thread Summary
The discussion focuses on the baseline recovery time of shaper amplifiers, emphasizing the importance of understanding different definitions based on output response to step inputs. Key points include the significance of time constants, with various thresholds (37%, <10%, and <1%) indicating how quickly the output stabilizes. For audio applications, a recovery time of one time constant may suffice, while classical music may require three to five time constants to avoid distortion. Participants suggest consulting datasheets and relevant literature, such as Knoll's standard text, for deeper insights. Ultimately, measuring the output's return to a specific percentage of its final value is crucial for accurate detector readout applications.
Adeelalishah
Messages
1
Reaction score
0
TL;DR Summary
I have a Bipolar shaper amplifier and I need to measure its Baseline Recovery time. Though I tried to study baseline recovery time concept but could not understand it.
From the picture in attachment (Black Curve) can some body kindly comment on what Baseline Recovery time is and how to measure it?
a.PNG
 
Engineering news on Phys.org
Welcome to the PF. :smile:

Can you post a link to the datasheet for your amp? With a quick Google search I find lots of helpful information on the baseline recovery time of pulse detection amps. Can you post links to some of the reading you have been doing, and ask specific questions about the parts you don't understand? Thanks.
 
There are different, somewhat common, definitions. They are all based on the time it takes from a step input (like the edge of a square wave) for the output to reach its final value.
They are:
[*] Time-Constant: when the output reaches to 37% of its final value
[*] 5 times the Time Constant: the output reaches to <1% of its final value
[*] 10%: the output reaches to <10% of its final value

For audio stuff, I expect that the 10% rule would be adequate but not ideal.

For a Rock Band you may get away with one Time Constant, 37%.

For Classical, or the 1812 Overture, consider 3 to 5 time constants. ( 3 Time constants gets you within 5% of final value. With the huge dynamic range you would want the recovery time to be quite fast to avoid audible distortion.)

Cheers,
Tom

p.s. These numbers are based on somewhat limited experience, so if someone else has different recommendations, believe them!
 
I take it this is for a detector readout application. Baseline recovery and restoration is covered in Knoll (the standard text for this area). Basically you have to define how close to the "final value" you need to get and measure that.
What Tom G. said is exactly right, but for you, a simpler way is to put a bunch of equal, widely spaced pulses in, find the final value (since its a bipolar shaper the final value should be the dc operating point), and then see when the output reaches 1% of that value (1% is typical for the accuracy of the kinds of detectors you would usually be reading out). Depending on what you're using the shaper for, it may be a different number.
 
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I have recently moved into a new (rather ancient) house and had a few trips of my Residual Current breaker. I dug out my old Socket tester which tell me the three pins are correct. But then the Red warning light tells me my socket(s) fail the loop test. I never had this before but my last house had an overhead supply with no Earth from the company. The tester said "get this checked" and the man said the (high but not ridiculous) earth resistance was acceptable. I stuck a new copper earth...
I am not an electrical engineering student, but a lowly apprentice electrician. I learn both on the job and also take classes for my apprenticeship. I recently wired my first transformer and I understand that the neutral and ground are bonded together in the transformer or in the service. What I don't understand is, if the neutral is a current carrying conductor, which is then bonded to the ground conductor, why does current only flow back to its source and not on the ground path...