Can't trigger on 2 channels simultaneously on oscilloscope =/

  • Thread starter Thread starter Nano-Passion
  • Start date Start date
  • Tags Tags
    Oscilloscope
Click For Summary

Discussion Overview

The discussion revolves around the challenges of triggering on two channels simultaneously using a Tektronix DPO4104 oscilloscope while measuring signals from scintillator detectors. Participants explore various methods to achieve effective triggering in the presence of noise and non-repetitive signals.

Discussion Character

  • Technical explanation
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • One participant notes that oscilloscopes typically allow triggering from only one source at a time, which can complicate simultaneous measurements.
  • Another suggests creating a combination trigger signal using external circuitry, proposing methods such as using resistors, diodes, or comparators to combine the two input signals.
  • A participant describes the characteristics of the signals they are trying to measure, indicating a need for edge triggering with specific voltage levels to filter out noise.
  • Some participants discuss the implications of signal synchronization, noting that asynchronous signals can lead to unstable traces on the oscilloscope.
  • Suggestions are made to use a storage oscilloscope with a 'one-shot' trigger feature to capture and analyze the signals more effectively.
  • There are inquiries about the participant's experimental setup, including the nature of the signals and the desired outcomes from the measurements.
  • One participant raises the possibility of using additional oscilloscopes with trigger outputs to combine trigger signals for the primary oscilloscope.

Areas of Agreement / Disagreement

Participants generally agree that triggering on two channels simultaneously is not standard practice and that external circuitry may be necessary to achieve the desired triggering. However, there are multiple competing views on the best approach to create a combination trigger signal, and the discussion remains unresolved regarding the most effective method.

Contextual Notes

Participants express uncertainty about the specific configurations and limitations of their oscilloscope, as well as the characteristics of the signals being measured. There are also unresolved questions about the practical implementation of suggested solutions.

Nano-Passion
Messages
1,291
Reaction score
0
I'm not sure where to put this so mods feel free to move it as needed.

I'm working on this oscilloscope trying to measure two signals from the detectors. To extract the signficiant signals, I need to use the trigger mechanism. However, the oscilloscope is not letting me (or I don't know how) apply the trigger to two channels at the same time. It has only been working with one trigger at a time.

The oscilloscope type that I am using is Tektronix DPO4104. Help would be very appreciated because I've spent many hours not being able to get this to work.
 
Last edited:
Engineering news on Phys.org
You can't trigger two signals at once (with some exceptions) on oscilloscopes. The trigger can only come from a single source, either a channel or an external source. Once the scope is triggered, everything, on all channels is recorded.
 
You can make a single trigger signal that is the combination of the two input signals -- you just have to do that with a separate circuit. What do the two analog signals look like? What kind of combination trigger signal would work for you?

It might be as simple as using two resistors to make a combination signal, or two diodes to OR the signals, or two comparators to detect separate trigger levels, and then OR the two comparator outputs, etc.
 
Topher925 said:
You can't trigger two signals at once (with some exceptions) on oscilloscopes. The trigger can only come from a single source, either a channel or an external source. Once the scope is triggered, everything, on all channels is recorded.

Oh.. that would explain a lot!

berkeman said:
You can make a single trigger signal that is the combination of the two input signals -- you just have to do that with a separate circuit.

What do the two analog signals look like? What kind of combination trigger signal would work for you?
I'm assuming you are asking what the signals look like on the oscilloscope. It is a negative pulse with sort of a vertical fall followed by a rugged rise on a slope (like a triangle).

Forgive me if I'm interpreting your second question wrong as well. The combination that I would like is straight forward -- just edge triggering on both channels (with noise reduction coupling), with negative slope. I would like the two trigger levels to differ however, one around 100 mv and the other around 28mv.

It might be as simple as using two resistors to make a combination signal, or two diodes to OR the signals, or two comparators to detect separate trigger levels, and then OR the two comparator outputs, etc.

Hmm, simple in practice of course. I am a complete beginner, I would love if you can send me a source that can guide me in this.

Thanks a lot for your help! :approve:
 
Hi Nano-P. Normally you only trigger on one channel or the other. If the two signals are synchronized then they will both appear as a stable trace on the scope. If however they are asynchronous (different unrelated frequencies for example), then only the trace you trigger can be stable, and the other one will "dance" around.

Sometimes in this situation you can use a storage scope with a 'one-shot" trigger to make more sense of what's going on.
 
Good suggestion by uart -- does your 'scope have a memory? Then you can store with one trigger and one-shot later with the other, and superimpose the two waveforms.

Why do you want to trigger on a combination of the two signals? Are they repetitive, or do they vary? Can you say more about what you are trying to do?

Worst case you can make a couple of amp/comparator circuits to do your own triggering on the two waveforms, and make the overall combined trigger signal as a combination of the two comparator outputs...
 
Uart, I do apologize but I'm not familiar with some of the lingo. And bekerman, I'll expound on what I am doing in the following.

I have a scintillator set up in connection to a photomultiplier, and I am measuring the output voltage from the photomultiplier. I am looking for a significant signal (ie a certain voltage) to tell me that I have measured a muon particle which interacts with the scintillator.

Now what happens is, there is too much noise and I can not extract any meaningful information without setting up trigger criterias. I have two different scintillators set up, each with their own photomultiplier connected with BNC cables onto the oscilloscope.

The signals are not repetitive. There is a lot of noise in the middle and every once in a while (I would estimate once per second) there is a pulse caused by the trigger. I have a attached an image for you to see.

uart said:
Hi Nano-P. Normally you only trigger on one channel or the other. If the two signals are synchronized then they will both appear as a stable trace on the scope. If however they are asynchronous (different unrelated frequencies for example), then only the trace you trigger can be stable, and the other one will "dance" around.

Sometimes in this situation you can use a storage scope with a 'one-shot" trigger to make more sense of what's going on.

berkeman said:
Good suggestion by uart -- does your 'scope have a memory? Then you can store with one trigger and one-shot later with the other, and superimpose the two waveforms.

Why do you want to trigger on a combination of the two signals? Are they repetitive, or do they vary? Can you say more about what you are trying to do?

Worst case you can make a couple of amp/comparator circuits to do your own triggering on the two waveforms, and make the overall combined trigger signal as a combination of the two comparator outputs...
 

Attachments

  • tek00001.png
    tek00001.png
    41.8 KB · Views: 1,004
Wow, that is challenging. Small waveforms and fast timescales. My idea of discrete amps and comparators probably is not practical for this.

Do you have any other oscilloscopes? Some 'scopes and logic analyzer / scope combination units have a "Trigger Out" BNC port. If you had 2 other scopes with Trigger Outupts, you could logically combine those two trigger signals into one to feed into your primary scope's External Trigger input...
 
berkeman said:
Wow, that is challenging. Small waveforms and fast timescales. My idea of discrete amps and comparators probably is not practical for this.

Do you have any other oscilloscopes? Some 'scopes and logic analyzer / scope combination units have a "Trigger Out" BNC port. If you had 2 other scopes with Trigger Outupts, you could logically combine those two trigger signals into one to feed into your primary scope's External Trigger input...

Yes, that sounds like a good idea. Thanks.
 

Similar threads

  • · Replies 10 ·
Replies
10
Views
6K
  • · Replies 12 ·
Replies
12
Views
2K
  • · Replies 3 ·
Replies
3
Views
3K
  • · Replies 9 ·
Replies
9
Views
5K
  • · Replies 6 ·
Replies
6
Views
4K
  • · Replies 11 ·
Replies
11
Views
3K
  • · Replies 41 ·
2
Replies
41
Views
10K
  • · Replies 9 ·
Replies
9
Views
2K
  • · Replies 7 ·
Replies
7
Views
7K
Replies
7
Views
3K